How does ramping work?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
kdt15
Posts: 39
Joined: Mon Jul 15, 2019 3:36 pm

How does ramping work?

Post by kdt15 » Sat Jan 04, 2020 4:01 pm

For example, 4-20 means it will be at 20 for sure or will it kinda waver between 4-20 after falling asleep? I saw a vid on myair resmed saying that it detects stable breathing pattern as falling asleep, then ramps, but i wasnt sure if it goes to max.

I have been consistently taking the mask off about 50min-1hr (3x now) so im curious if that means pressure is to ohigh and bothering me, if it indeed ramps max

_________________
Mask: Amara Full Face CPAP Mask with Gel & Silicone Cushions
Additional Comments: 30M, 5'9, 165lb, active lifestyle, Diagnosis AHI ~7

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How does ramping work?

Post by palerider » Sat Jan 04, 2020 4:03 pm

https://www.youtube.com/watch?v=4GW97Xk06N8

The min and max pressure set has nothing at all to do with ramp.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13405
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: How does ramping work?

Post by LSAT » Sat Jan 04, 2020 5:16 pm

Do you know for sure that your pressure settings are 4-20? Your set pressures will show on your screen when you turn on your machine. 4-20 are the factory default settings. If your settings are really 4-20 , ramps wont be necessary.

User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: How does ramping work?

Post by kteague » Sun Jan 05, 2020 8:20 am

Terming increasing treatment pressure as "ramping up" can lead to confusion. Yes, the pressure will increase as determined by breathing indicators when set in a range. The confusion comes in because there is a specific "Ramp" feature on most machines, with separate settings from the treatment pressure settings. This optional comfort feature allows one who finds their therapeutic pressure a bit much starting off to ease up to their therapeutic pressure. As was already indicated, anyone with a range of 4-20 has no need for the separate Ramp feature as their pressure is already starting off low. The pressure will only go as high as needed. Do check your data so that you can make decisions based on facts rather than guesses (yours or ours). You are thinking the machine reaching a high pressure is disrupting your sleep. Could be when you get into deeper sleep or REM your machine is unable to address impending events fast enough if your needed pressure is a lot higher than your starting pressure. The pressure increases incrementally over set time frames, so it's best to have the lower end of the pressure range up pretty close to where you expect it to need to go based on history. This shortens the time it takes to reach a therapeutic pressure. Again, your data is your friend.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

kdt15
Posts: 39
Joined: Mon Jul 15, 2019 3:36 pm

Re: How does ramping work?

Post by kdt15 » Sun Jan 05, 2020 11:14 am

kteague wrote:
Sun Jan 05, 2020 8:20 am
Terming increasing treatment pressure as "ramping up" can lead to confusion. Yes, the pressure will increase as determined by breathing indicators when set in a range. The confusion comes in because there is a specific "Ramp" feature on most machines, with separate settings from the treatment pressure settings. This optional comfort feature allows one who finds their therapeutic pressure a bit much starting off to ease up to their therapeutic pressure. As was already indicated, anyone with a range of 4-20 has no need for the separate Ramp feature as their pressure is already starting off low. The pressure will only go as high as needed. Do check your data so that you can make decisions based on facts rather than guesses (yours or ours). You are thinking the machine reaching a high pressure is disrupting your sleep. Could be when you get into deeper sleep or REM your machine is unable to address impending events fast enough if your needed pressure is a lot higher than your starting pressure. The pressure increases incrementally over set time frames, so it's best to have the lower end of the pressure range up pretty close to where you expect it to need to go based on history. This shortens the time it takes to reach a therapeutic pressure. Again, your data is your friend.
i have the data on the sd card inserted in and i check the myair every day, what should i draw from it?

_________________
Mask: Amara Full Face CPAP Mask with Gel & Silicone Cushions
Additional Comments: 30M, 5'9, 165lb, active lifestyle, Diagnosis AHI ~7

User avatar
Dog Slobber
Posts: 4331
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: How does ramping work?

Post by Dog Slobber » Sun Jan 05, 2020 11:40 am

MyAir is a terrible diagnostic too.

We use OSCAR (which has replaced SleepyHead) and import the data from the SD Card.

Read the pinned topics to learn about OSCAR/SleepyHead
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11462
Joined: Fri Jun 19, 2015 4:36 pm

Re: How does ramping work?

Post by zonker » Sun Jan 05, 2020 12:52 pm

kdt15 wrote:
Sun Jan 05, 2020 11:14 am

i have the data on the sd card inserted in and i check the myair every day, what should i draw from it?
to dog pile onto dog slobber(get it? :lol: :lol: :lol: ), please get the oscar software. you'll find a link for downloading in my sig.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: How does ramping work?

Post by chunkyfrog » Sun Jan 05, 2020 2:24 pm

Just get rid of MyAir--it is a complete waste of time.
Everything I need is available on the on-screen sleep report.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: How does ramping work?

Post by kteague » Mon Jan 06, 2020 5:09 am

I've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: How does ramping work?

Post by palerider » Mon Jan 06, 2020 8:44 am

kteague wrote:
Mon Jan 06, 2020 5:09 am
I've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?
It doesn't even give you as much data as the Oscar *SUMMARY* page.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
babydinosnoreless
Posts: 2373
Joined: Fri Nov 02, 2018 2:53 pm

Re: How does ramping work?

Post by babydinosnoreless » Mon Jan 06, 2020 11:35 am

kteague wrote:
Mon Jan 06, 2020 5:09 am
I've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?
This is what you get from my air. Just kind of a "rah rah you've got this" summary it doesn't show you the type of events or anything.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: How does ramping work?

Post by jimbud » Mon Jan 06, 2020 12:02 pm

While you are at it get your clinician manual for your machine.
You will then be ready to ask many more informed questions.

Links to Oscar and your manual below.

https://www.sleepfiles.com/OSCAR/

Clinician manual:

https://www.apneaboard.com/adjust-cpap- ... tup-manual
Scroll down to find your manual.

JPB

Good luck.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: How does ramping work?

Post by chunkyfrog » Mon Jan 06, 2020 1:06 pm

Run, cpap, run.
I remember being completely insulted by the Dick and Jane reading program--at FIVE.
I could read the newspaper in the first grade. But I preferred comics.
And the encyclopedia. So much information--I was addicted.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her