How does ramping work?
How does ramping work?
For example, 4-20 means it will be at 20 for sure or will it kinda waver between 4-20 after falling asleep? I saw a vid on myair resmed saying that it detects stable breathing pattern as falling asleep, then ramps, but i wasnt sure if it goes to max.
I have been consistently taking the mask off about 50min-1hr (3x now) so im curious if that means pressure is to ohigh and bothering me, if it indeed ramps max
I have been consistently taking the mask off about 50min-1hr (3x now) so im curious if that means pressure is to ohigh and bothering me, if it indeed ramps max
_________________
| Mask: Amara Full Face CPAP Mask with Gel & Silicone Cushions |
| Additional Comments: 30M, 5'9, 165lb, active lifestyle, Diagnosis AHI ~7 |
Re: How does ramping work?
https://www.youtube.com/watch?v=4GW97Xk06N8
The min and max pressure set has nothing at all to do with ramp.
The min and max pressure set has nothing at all to do with ramp.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: How does ramping work?
Do you know for sure that your pressure settings are 4-20? Your set pressures will show on your screen when you turn on your machine. 4-20 are the factory default settings. If your settings are really 4-20 , ramps wont be necessary.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: How does ramping work?
Terming increasing treatment pressure as "ramping up" can lead to confusion. Yes, the pressure will increase as determined by breathing indicators when set in a range. The confusion comes in because there is a specific "Ramp" feature on most machines, with separate settings from the treatment pressure settings. This optional comfort feature allows one who finds their therapeutic pressure a bit much starting off to ease up to their therapeutic pressure. As was already indicated, anyone with a range of 4-20 has no need for the separate Ramp feature as their pressure is already starting off low. The pressure will only go as high as needed. Do check your data so that you can make decisions based on facts rather than guesses (yours or ours). You are thinking the machine reaching a high pressure is disrupting your sleep. Could be when you get into deeper sleep or REM your machine is unable to address impending events fast enough if your needed pressure is a lot higher than your starting pressure. The pressure increases incrementally over set time frames, so it's best to have the lower end of the pressure range up pretty close to where you expect it to need to go based on history. This shortens the time it takes to reach a therapeutic pressure. Again, your data is your friend.
_________________
| Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions |
My SleepDancing Video link https://www.youtube.com/watch?v=jE7WA_5c73c
Re: How does ramping work?
i have the data on the sd card inserted in and i check the myair every day, what should i draw from it?kteague wrote: ↑Sun Jan 05, 2020 8:20 amTerming increasing treatment pressure as "ramping up" can lead to confusion. Yes, the pressure will increase as determined by breathing indicators when set in a range. The confusion comes in because there is a specific "Ramp" feature on most machines, with separate settings from the treatment pressure settings. This optional comfort feature allows one who finds their therapeutic pressure a bit much starting off to ease up to their therapeutic pressure. As was already indicated, anyone with a range of 4-20 has no need for the separate Ramp feature as their pressure is already starting off low. The pressure will only go as high as needed. Do check your data so that you can make decisions based on facts rather than guesses (yours or ours). You are thinking the machine reaching a high pressure is disrupting your sleep. Could be when you get into deeper sleep or REM your machine is unable to address impending events fast enough if your needed pressure is a lot higher than your starting pressure. The pressure increases incrementally over set time frames, so it's best to have the lower end of the pressure range up pretty close to where you expect it to need to go based on history. This shortens the time it takes to reach a therapeutic pressure. Again, your data is your friend.
_________________
| Mask: Amara Full Face CPAP Mask with Gel & Silicone Cushions |
| Additional Comments: 30M, 5'9, 165lb, active lifestyle, Diagnosis AHI ~7 |
- Dog Slobber
- Posts: 4333
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: How does ramping work?
MyAir is a terrible diagnostic too.
We use OSCAR (which has replaced SleepyHead) and import the data from the SD Card.
Read the pinned topics to learn about OSCAR/SleepyHead
We use OSCAR (which has replaced SleepyHead) and import the data from the SD Card.
Read the pinned topics to learn about OSCAR/SleepyHead
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: How does ramping work?
to dog pile onto dog slobber(get it?
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: How does ramping work?
Just get rid of MyAir--it is a complete waste of time.
Everything I need is available on the on-screen sleep report.
Everything I need is available on the on-screen sleep report.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: How does ramping work?
I've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?
_________________
| Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions |
My SleepDancing Video link https://www.youtube.com/watch?v=jE7WA_5c73c
Re: How does ramping work?
It doesn't even give you as much data as the Oscar *SUMMARY* page.kteague wrote: ↑Mon Jan 06, 2020 5:09 amI've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- babydinosnoreless
- Posts: 2373
- Joined: Fri Nov 02, 2018 2:53 pm
Re: How does ramping work?
This is what you get from my air. Just kind of a "rah rah you've got this" summary it doesn't show you the type of events or anything.kteague wrote: ↑Mon Jan 06, 2020 5:09 amI've never seen MyAir but people who have say it doesn't give you nearly as much information as Oscar. If you notice some of the other threads where Oscar/Sleepyhead charts have been posted, there's lots of detail that those who know this stuff inside and out can use and be very helpful in getting one's therapy optimized. Without these details there's more guesswork, and who has time for that when you're trying to feel better sooner rather than later, right?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: How does ramping work?
While you are at it get your clinician manual for your machine.
You will then be ready to ask many more informed questions.
Links to Oscar and your manual below.
https://www.sleepfiles.com/OSCAR/
Clinician manual:
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Scroll down to find your manual.
JPB
Good luck.
You will then be ready to ask many more informed questions.
Links to Oscar and your manual below.
https://www.sleepfiles.com/OSCAR/
Clinician manual:
https://www.apneaboard.com/adjust-cpap- ... tup-manual
Scroll down to find your manual.
JPB
Good luck.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: How does ramping work?
Run, cpap, run.
I remember being completely insulted by the Dick and Jane reading program--at FIVE.
I could read the newspaper in the first grade. But I preferred comics.
And the encyclopedia. So much information--I was addicted.
I remember being completely insulted by the Dick and Jane reading program--at FIVE.
I could read the newspaper in the first grade. But I preferred comics.
And the encyclopedia. So much information--I was addicted.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |