All of the sudden intolerable
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
All of the sudden intolerable
Long time lurker first time poster. I’ve searched and have seen many threads about dry mouth issues but I wanted to lay out what I’m struggling with in hopes of finding suggestions or support.
Been doing the CPAP thing for maybe 6 months. No big problems until recently, actually used to enjoy using it.
All of the sudden maybe 2-3 weeks ago it’s like my machine felt like it’s pushing the air harder. I had a SoClean device hooked up and it seemed like it was causing air leaks and extra noise. I ditched the SoClean and the machine became quieter and seems to work better. Problem on that front is I guess the SoClean is a piece of junk and likely I’ll need to throw it out or find a way to fix it. Seems to me like it’s a bad design and their adapter piece causes leaks, or was causing my machine to act differently.
The persistent problem I now can’t fix is dry mouth. I hate using my device now. It’s terrible. I’m probably waking up every few hours with the worst dry mouth I’ve ever experienced. Tried cranking humidity, no help. The chin-strap sounds very unappealing because I’m worried about clenching my teeth all night. It’s like, I was breathing through my mouth for months and never got this extreme dry mouth, now all of the sudden every night. It’s a real downer and I’m exhausted. I also still feel like my machine is punishing me with too much air. I go to bed fine and then an hour later it’s just massive air coming at me and major dry mouth. Several times per night I’m removing the mask out of frustration, then strapping it on and trying again. Waking up with headaches and feeling desperate- any advice out there?
I plan on contacting my sleep doc next week- he’s closed for the weekend.
Been doing the CPAP thing for maybe 6 months. No big problems until recently, actually used to enjoy using it.
All of the sudden maybe 2-3 weeks ago it’s like my machine felt like it’s pushing the air harder. I had a SoClean device hooked up and it seemed like it was causing air leaks and extra noise. I ditched the SoClean and the machine became quieter and seems to work better. Problem on that front is I guess the SoClean is a piece of junk and likely I’ll need to throw it out or find a way to fix it. Seems to me like it’s a bad design and their adapter piece causes leaks, or was causing my machine to act differently.
The persistent problem I now can’t fix is dry mouth. I hate using my device now. It’s terrible. I’m probably waking up every few hours with the worst dry mouth I’ve ever experienced. Tried cranking humidity, no help. The chin-strap sounds very unappealing because I’m worried about clenching my teeth all night. It’s like, I was breathing through my mouth for months and never got this extreme dry mouth, now all of the sudden every night. It’s a real downer and I’m exhausted. I also still feel like my machine is punishing me with too much air. I go to bed fine and then an hour later it’s just massive air coming at me and major dry mouth. Several times per night I’m removing the mask out of frustration, then strapping it on and trying again. Waking up with headaches and feeling desperate- any advice out there?
I plan on contacting my sleep doc next week- he’s closed for the weekend.
Last edited by SmokeScreen on Sat Jan 04, 2020 10:11 am, edited 2 times in total.
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: All of the sudden intolerable
Go with your first thought.SmokeScreen wrote: ↑Sat Jan 04, 2020 8:22 amProblem on that front is I guess the SoClean is a piece of junk and likely I’ll need to throw it out or find a way to fix it.
What kind of mask have you got?The persistent problem I now can’t fix is dry mouth. I hate using my device now. It’s terrible. I’m probably waking up every few hours with the worst dry mouth I’ve ever experienced. Tried cranking humidity, no help. The chin-strap sounds very unappealing because I’m worried about clenching my teeth all night. It’s like, I was breathing through my mouth for months and never got this extreme dry mouth, now all of the sudden every night. It’s a real downer and I’m exhausted. I also still feel like my machine is punishing me with too much air. I go to bed fine and then an hour later it’s just massive air coming at me and major dry mouth. Sever times per night I’m removing the mask out of frustration, then strapping it on and trying again. Waking up with headaches and feeling desperate- any advice out there?
I plan on contacting my sleep doc next week- he’s closed for the weekend.
You might have started mouth breathing, or opening your mouth and a lot of air is rushing out.
Consider taping, or a strap, or maybe a full-face mask.
Load your data into OSCAR and look to see if your leaks have gone up since the dry mouth began.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
Thanks for the response - I’ll check out OSCAR, so far I was just using the App on my IPhone.
My device is a Phillips Dreamstation and the mask is a ResMed AirFit F20. I wear a large mask and seem to get a better mask fit rating than medium size, but I’m almost between sizes. I’ve switched a few times between M and L and lately using the L size.
I also have this feeling there’s a leak in the hose where the mask connects, or something feels a bit off with my equipment. I get my stuff from Apria Health and they haven’t scored high on customer service. Starting to wonder if they sent me inferior replacement parts.
My device is a Phillips Dreamstation and the mask is a ResMed AirFit F20. I wear a large mask and seem to get a better mask fit rating than medium size, but I’m almost between sizes. I’ve switched a few times between M and L and lately using the L size.
I also have this feeling there’s a leak in the hose where the mask connects, or something feels a bit off with my equipment. I get my stuff from Apria Health and they haven’t scored high on customer service. Starting to wonder if they sent me inferior replacement parts.
Re: All of the sudden intolerable
Many sizing guides indicate that if you are between sizes..go for the smaller. Apria is a PITA, but the parts come from the manufacturer.SmokeScreen wrote: ↑Sat Jan 04, 2020 8:41 amThanks for the response - I’ll check out OSCAR, so far I was just using the App on my IPhone.
My device is a Phillips Dreamstation and the mask is a ResMed AirFit F20. I wear a large mask and seem to get a better mask fit rating than medium size, but I’m almost between sizes. I’ve switched a few times between M and L and lately using the L size.
I also have this feeling there’s a leak in the hose where the mask connects, or something feels a bit off with my equipment. I get my stuff from Apria Health and they haven’t scored high on customer service. Starting to wonder if they sent me inferior replacement parts.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: All of the sudden intolerable
The 2 main causes of dry mouth....mouth breathing or medication side effects.
Has anything changed with your medications? Added something new or changed dosage on what you have been using?
Any nasal congestion going on that might be causing more mouth breathing than usual?
Has anything changed with your medications? Added something new or changed dosage on what you have been using?
Any nasal congestion going on that might be causing more mouth breathing than usual?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
No medications so nothing on that front... My left ear has been feeling blocked and perhaps some slight nasal congestion. I did travel a few weeks ago and came home with a cough/cold- the cough hasn't entirely resolved. The weird thing is I've thought I've always been breathing through my mouth, it's just this intense dry-mouth seems to only be a problem very recently. Feeling perplexed, I also run a humidifier in the room and the past few weeks haven't been particularly dry.Pugsy wrote: ↑Sat Jan 04, 2020 8:59 amThe 2 main causes of dry mouth....mouth breathing or medication side effects.
Has anything changed with your medications? Added something new or changed dosage on what you have been using?
Any nasal congestion going on that might be causing more mouth breathing than usual?
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
Here’s something interesting... I still haven’t got the OSCAR software running because no SD card adapter nearby, however, I did start digging into the history on my DreamMapper App.
It seems like the “90% pressure” is higher right around when my problems started.
Typical data from when everything was comfortable:
2.2ahi
95% mask fit
Min/Max pressure: 4/20cmH20
90% pressure: 12cmH20
And more recent when I’m struggling:
0.7ahi
97% mask fit
Min/Max pressure: 4/20cmH20
90% pressure: 17.2cmH20
I’m noticing my problems started when that 90% pressure jumped from ~12 to ~17. It’s the one piece of data on this app that corresponds to when I started having problems.
It seems like the “90% pressure” is higher right around when my problems started.
Typical data from when everything was comfortable:
2.2ahi
95% mask fit
Min/Max pressure: 4/20cmH20
90% pressure: 12cmH20
And more recent when I’m struggling:
0.7ahi
97% mask fit
Min/Max pressure: 4/20cmH20
90% pressure: 17.2cmH20
I’m noticing my problems started when that 90% pressure jumped from ~12 to ~17. It’s the one piece of data on this app that corresponds to when I started having problems.
Re: All of the sudden intolerable
More pressure means more air moving through the mouth (if mouth breathing) and more air movement means more chance of drying out of the mouth.
Now what caused the more pressure.....that is the million dollar question.
Might be sleeping position worsening the OSA causing an increase in pressure needs....might be more REM sleep which can also cause worsening of the OSA....might be the machine being fooled a bit by nasal congestion thinking it is some sort of apnea airway event going on.
Really need to see those detailed reports to see just when the pressure increases and for how long.
90% pressures....all that means is the pressure was AT OR BELOW that number for 90% of the night. People often don't understand the "or below" part of the definition and they assume it means you were at that pressure for 90% of the night.
90% numbers are easily skewed upwards by a relatively short period of time at the higher number so I don't put much stock in just the number until I actually see the detailed data from the software reports.
It takes very little mouth breathing to dry out the mouth so maybe you are just mouth breathing a little more than your usual.
Now what caused the more pressure.....that is the million dollar question.
Might be sleeping position worsening the OSA causing an increase in pressure needs....might be more REM sleep which can also cause worsening of the OSA....might be the machine being fooled a bit by nasal congestion thinking it is some sort of apnea airway event going on.
Really need to see those detailed reports to see just when the pressure increases and for how long.
90% pressures....all that means is the pressure was AT OR BELOW that number for 90% of the night. People often don't understand the "or below" part of the definition and they assume it means you were at that pressure for 90% of the night.
90% numbers are easily skewed upwards by a relatively short period of time at the higher number so I don't put much stock in just the number until I actually see the detailed data from the software reports.
It takes very little mouth breathing to dry out the mouth so maybe you are just mouth breathing a little more than your usual.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: All of the sudden intolerable
Are we REALLY sure?
Unless the parts are packed in original factory packages, we can never be completely sure.
(And maybe not even then.)
If anyone were to handle third world knock-offs, Apria has the resources and character to do so.
Not saying they do, but anything is possible.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
Wanted to update... I still don’t have my hands on an SD Card reader, but I think my problem was a damaged hose.
I felt like the hose might be leaking, swapped it out and noticed the mask-fit test looked a bit better on the DreamStation display screen. With the old hose it still passed my mask-fit, but the bar would go red a few times. Once I switched the hose is was all green the whole test cycle. Past two nights slept fine, no major dry mouth. 90% pressure was back down to 13.
I also changed the filter and started using Flonase spray, but pretty sure something was damaged on my other hose. Anyways, everything is working much better now.
Sent from my iPhone using Tapatalk
I felt like the hose might be leaking, swapped it out and noticed the mask-fit test looked a bit better on the DreamStation display screen. With the old hose it still passed my mask-fit, but the bar would go red a few times. Once I switched the hose is was all green the whole test cycle. Past two nights slept fine, no major dry mouth. 90% pressure was back down to 13.
I also changed the filter and started using Flonase spray, but pretty sure something was damaged on my other hose. Anyways, everything is working much better now.
Sent from my iPhone using Tapatalk
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: All of the sudden intolerable
Good.
Now put the G-D soclean out by the curb, and you're all set.
Now put the G-D soclean out by the curb, and you're all set.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: All of the sudden intolerable
Amazon, less than 10$, delivered right to your door.SmokeScreen wrote: ↑Mon Jan 06, 2020 7:03 pmWanted to update... I still don’t have my hands on an SD Card reader,
Or, bestbuy, walmart, etc if you don't want to use amazon.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
palerider wrote:Amazon, less than 10$, delivered right to your door.SmokeScreen wrote: ↑Mon Jan 06, 2020 7:03 pmWanted to update... I still don’t have my hands on an SD Card reader,
Or, bestbuy, walmart, etc if you don't want to use amazon.
I own one it’s just buried in a drawer and my new MacBook I need to dig out an adapter. Fixing the hardware felt less daunting than digging up stuff to mess with the software

Sent from my iPhone using Tapatalk
-
SmokeScreen
- Posts: 7
- Joined: Sat Jan 04, 2020 7:56 am
Re: All of the sudden intolerable
Yeah seems like I fell for the gimmick... what a terrible product.chunkyfrog wrote:Good.
Now put the G-D soclean out by the curb, and you're all set.
Sent from my iPhone using Tapatalk
Re: All of the sudden intolerable
You could always gift it to someone you hate for their birthday or next Christmas.SmokeScreen wrote: ↑Tue Jan 07, 2020 1:43 pmYeah seems like I fell for the gimmick... what a terrible product.chunkyfrog wrote:Good.
Now put the G-D soclean out by the curb, and you're all set.
Sent from my iPhone using Tapatalk
JPB
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleep on a Buckwheat Hull Pillow. |