False Mask On/Offs

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: False Mask On/Offs

Post by Dog Slobber » Sat Dec 28, 2019 3:59 pm

palerider wrote:
Sat Dec 28, 2019 3:43 pm
Dog Slobber wrote:
Sat Dec 28, 2019 2:34 pm
zonker wrote:
Sat Dec 28, 2019 12:41 pm
Dog Slobber wrote:
Sat Dec 28, 2019 9:21 am
Enjoy your new DreamStation.

And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
i'm going to have to advice you of a spelling error....
if I was advized of a spelling error every time I made one my notifications would look like this:


advize.png
Ok, all you proof readers, let's see if we can get that up to quadruple digits, one notice per error.
Just an FYI:

If you get a link called 'free porn' don't opin it. It is a viras wich deattivates you spelcheck and fcuks up you riting. I also receibed it but lukily I dont uatch porn so I dint opin it. Plaese warm yu frends. Wanks.
Battery Backup: EcoFlow Delta 2

novoski
Posts: 8
Joined: Sun Dec 12, 2010 10:54 am

Re: False Mask On/Offs

Post by novoski » Sun Dec 29, 2019 11:17 am

I did turn off the auto start and set the max pressure in the 12 range because at 20 it was actually pumping out 20 the majority of the time. The other bizarre thing was NO leaks, Not possible.

The Dreamstation will be here Friday and I'll be really surprised if I get similar results.
Annotation 2019-12-29 090726.png
Annotation 2019-12-29 090726.png (53.27 KiB) Viewed 21643 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sun Dec 29, 2019 3:48 pm

novoski wrote:
Sun Dec 29, 2019 11:17 am
I did turn off the auto start and set the max pressure in the 12 range because at 20 it was actually pumping out 20 the majority of the time.
Which would mean you needed 20.
novoski wrote:
Sun Dec 29, 2019 11:17 am
The other bizarre thing was NO leaks, Not possible.
Entirely possible.
novoski wrote:
Sun Dec 29, 2019 11:17 am
The Dreamstation will be here Friday and I'll be really surprised if I get similar results.
So will I.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

novoski
Posts: 8
Joined: Sun Dec 12, 2010 10:54 am

Re: False Mask On/Offs

Post by novoski » Sun Dec 29, 2019 4:06 pm

Palerider. You sound like a Shell answer man. :) What was your explanation for the number of sessions? Do you think that really happened? If it did I can't believe that wouldn't have woken me up.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sun Dec 29, 2019 4:13 pm

novoski wrote:
Sun Dec 29, 2019 4:06 pm
Palerider. You sound like a Shell answer man. :) What was your explanation for the number of sessions? Do you think that really happened? If it did I can't believe that wouldn't have woken me up.
I haven't seen the data, you haven't posted the requested examples, so I have no explanation, and I don't like guessing.

Post a couple of nights data, follow this format:
http://cpaptalk.com/wiki/index.php/Oscar:organize

I don't see any on-off sessions in the last picture you posted, but it doesn't show data that I need to be sure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

novoski
Posts: 8
Joined: Sun Dec 12, 2010 10:54 am

Re: False Mask On/Offs

Post by novoski » Sun Dec 29, 2019 4:31 pm

Palerider,

The Session beginning and ending times are shown on the Left side of my Oscar sleep report.

Dale

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: False Mask On/Offs

Post by Dog Slobber » Sun Dec 29, 2019 4:55 pm

novoski wrote:
Sun Dec 29, 2019 4:06 pm
Palerider. You sound like a Shell answer man. :) What was your explanation for the number of sessions? Do you think that really happened? If it did I can't believe that wouldn't have woken me up.
The fact that the trace continues *through* the session on/off times suggests that the machine did not power down, but your therapy continued.

But as noted previously, trying to get answers from you is a pain in the ass. You've been asked twice now if you've power cycled the machine and you still haven't answered.

You may have bought a new Respironics DreamStation when all you needed was an $8 SD Card, or possibly just a formatting because of a corrupt file system.

There's other things we can learn regarding the leak, by comparing reported detail data against reported summary data but why should we beg you to provide the answer.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sun Dec 29, 2019 5:47 pm

novoski wrote:
Sun Dec 29, 2019 4:31 pm
Palerider,

The Session beginning and ending times are shown on the Left side of my Oscar sleep report.

Dale
I don't care what that says, I want to see a *proper* screenshot, which has things like the SETTING, which you haven't shown.

I've seen machines report false sessions when they *clearly* have not turned off and back on, and yours looks like that's what's happening.

Post a proper screenshot, and I'll show you what I'm talking about.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sun Dec 29, 2019 5:48 pm

Dog Slobber wrote:
Sun Dec 29, 2019 4:55 pm
novoski wrote:
Sun Dec 29, 2019 4:06 pm
Palerider. You sound like a Shell answer man. :) What was your explanation for the number of sessions? Do you think that really happened? If it did I can't believe that wouldn't have woken me up.
The fact that the trace continues *through* the session on/off times suggests that the machine did not power down, but your therapy continued.
YUP! Look at the pressure at the start of the night... compare that to the gaps.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

novoski
Posts: 8
Joined: Sun Dec 12, 2010 10:54 am

Re: False Mask On/Offs

Post by novoski » Sat Jan 04, 2020 11:04 am

I tried a different memory card in the Airsense 10 and had the same results with +30 mask sessions a night and no leaks. I had the mask on for +7 hours and the machine showed that it was operational for +4 hours. The Dreamstation Auto came in yesterday and last night I had 1 session with +8 hours on the machine. I guess the Airsense goes to the dump.

Thanks to those who tried to assist my issues.

Dale

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sat Jan 04, 2020 11:13 am

novoski wrote:
Sat Jan 04, 2020 11:04 am
I guess the Airsense goes to the dump.
I'll pay shipping to send it to me.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: False Mask On/Offs

Post by chunkyfrog » Sat Jan 04, 2020 11:27 am

I wonder if the OP's local electric supplier has been having frequent outages .
"Let's just do the service at night--nobody will notice." :mrgreen:
Call your local utility, and ask if there have been outages in your neighborhood.
Checking the clocks on appliances is not always helpful,
because some have internal batteries for the clock.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: False Mask On/Offs

Post by palerider » Sat Jan 04, 2020 12:43 pm

chunkyfrog wrote:
Sat Jan 04, 2020 11:27 am
I wonder if the OP's local electric supplier has been having frequent outages .
"Let's just do the service at night--nobody will notice." :mrgreen:
Call your local utility, and ask if there have been outages in your neighborhood.
Checking the clocks on appliances is not always helpful,
because some have internal batteries for the clock.
The fact that there's no rampup of pressure tells me it's just data glitches.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: False Mask On/Offs

Post by chunkyfrog » Sat Jan 04, 2020 1:11 pm

IOW, unplug, count to ten, replug and try again.(cold reset)
If the problem persists, warranty service or replacement is indicated.
Or use it anyway, but forget using the data. (And hope for the best)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her