False Mask On/Offs
False Mask On/Offs
About a month ago my Resscan reports for my Airsense 10 autoset began showing no Leak information and more than 10 mask on/offs for the night when I know I only went to bed and woke up so it should show 2 mask on/offs. It has been doing this consistently for over a month. I know it also should show some leaks. Do I need to send this machine in for service?
Thanks,
Dale
Thanks,
Dale
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Re: False Mask On/Offs
is it possible for you to download the oscar software and post some graphs? i'm just wondering if resscan isn't reporting properly. and you may want to double check that using oscar.novoski wrote: ↑Tue Dec 17, 2019 9:55 pmAbout a month ago my Resscan reports for my Airsense 10 autoset began showing no Leak information and more than 10 mask on/offs for the night when I know I only went to bed and woke up so it should show 2 mask on/offs. It has been doing this consistently for over a month. I know it also should show some leaks. Do I need to send this machine in for service?
Thanks,
Dale
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: False Mask On/Offs
Oscar is more user-friendly than Resscan.
Likely because it is developed by and FOR patients.
Likely because it is developed by and FOR patients.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: False Mask On/Offs
How old is the machine?
Have you tried unplugging the machine from the electric plug and then plugging it back in. This forces the machine's internal software to reboot. Very similar to when we reboot a computer to clean out the cobwebs.
To test the leak recording. When you start the machine play with it for about 5 or 10 minutes and force it to leak for sure with a big leak.
See what it records.
If after all that...no joy and it's still not showing you what you know it should show you....it needs to be serviced and you need to alert whomever you got the machine from that there is a problem with what the machine writes to the SD card.
Have you tried unplugging the machine from the electric plug and then plugging it back in. This forces the machine's internal software to reboot. Very similar to when we reboot a computer to clean out the cobwebs.
To test the leak recording. When you start the machine play with it for about 5 or 10 minutes and force it to leak for sure with a big leak.
See what it records.
If after all that...no joy and it's still not showing you what you know it should show you....it needs to be serviced and you need to alert whomever you got the machine from that there is a problem with what the machine writes to the SD card.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: False Mask On/Offs
I downloaded the software and it has similar results to the Resscan regarding false mask on/offs. I have between 20 and 30 sessions showing each night although, to my knowledge, I don't remove my mask until morning, so I would guess the machine must be messing up.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: False Mask On/Offs
Have you a lot of leaks going on ?novoski wrote: ↑Sun Dec 22, 2019 9:44 pmI downloaded the software and it has similar results to the Resscan regarding false mask on/offs. I have between 20 and 30 sessions showing each night although, to my knowledge, I don't remove my mask until morning, so I would guess the machine must be messing up.
Or go into the clinical menu and turn off smart start. It seems your machine thinks youve taken you mask off.
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: False Mask On/Offs
Seeing an example screenshot with pressure, leak and mask pressure might shed some light.novoski wrote: ↑Sun Dec 22, 2019 9:44 pmI downloaded the software and it has similar results to the Resscan regarding false mask on/offs. I have between 20 and 30 sessions showing each night although, to my knowledge, I don't remove my mask until morning, so I would guess the machine must be messing up.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: False Mask On/Offs
A "mask on/off" is really a device on/off.
- Turn machine off, for a minute or so, then turn back on.
- Ensure power cable is properly plugged into device, power brick and wall.
- Turn off "SmartStart".
- Post graphs to see if there is any kind of patterns to power cycling.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: False Mask On/Offs
Attached are some of my my Oscar graphs from last night. I ordered a Dreamstation Auto as a replacement machine. Too bad the Airsense 10 Autoset apparently failed.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
- Attachments
-
- Oscar Graphs.png (70.22 KiB) Viewed 16207 times
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: False Mask On/Offs
That’s not a daily graph that’s a summary graph. Looking at the number of sessions your machine has been turning off for a long time since October. Did you try any of the suggestions like turning off smart start ?
Don’t by a dream station stick with the Resmed Autoset 10 your used too.
Don’t by a dream station stick with the Resmed Autoset 10 your used too.
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: False Mask On/Offs
Enjoy your new DreamStation.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: False Mask On/Offs
Now now, maybe they'll be one of the small percentage of people that prefer a dreamstation, and encore...Dog Slobber wrote: ↑Sat Dec 28, 2019 9:21 amEnjoy your new DreamStation.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: False Mask On/Offs
i'm going to have to advice you of a spelling error....Dog Slobber wrote: ↑Sat Dec 28, 2019 9:21 amEnjoy your new DreamStation.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Dog Slobber
- Posts: 4331
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: False Mask On/Offs
if I was advized of a spelling error every time I made one my notifications would look like this:zonker wrote: ↑Sat Dec 28, 2019 12:41 pmi'm going to have to advice you of a spelling error....Dog Slobber wrote: ↑Sat Dec 28, 2019 9:21 amEnjoy your new DreamStation.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: False Mask On/Offs
Ok, all you proof readers, let's see if we can get that up to quadruple digits, one notice per error.Dog Slobber wrote: ↑Sat Dec 28, 2019 2:34 pmif I was advized of a spelling error every time I made one my notifications would look like this:zonker wrote: ↑Sat Dec 28, 2019 12:41 pmi'm going to have to advice you of a spelling error....Dog Slobber wrote: ↑Sat Dec 28, 2019 9:21 amEnjoy your new DreamStation.
And if you have any questions, please ask. Then proceed to completely ignoring any advise, suggestions or answers and you can just do your own thing anyway.
advize.png
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

