Back up battery for Air Sense 10 with a pressure of 20

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Bern
Posts: 2
Joined: Thu Nov 14, 2019 12:21 pm

Back up battery for Air Sense 10 with a pressure of 20

Post by Bern » Thu Nov 14, 2019 12:31 pm

Having problems finding a back-up power for my CPAP since my pressure is so high. 5 or 6 hours would be acceptable. Medistrom failed me (last 1 hour). MaxOak K5 maybe ok but hard to tell (well for me anyway). Hoping someone can help. I am open to suggestions.

Thank you.

User avatar
zonker
Posts: 11339
Joined: Fri Jun 19, 2015 4:36 pm

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by zonker » Thu Nov 14, 2019 1:04 pm

Bern wrote:
Thu Nov 14, 2019 12:31 pm
Having problems finding a back-up power for my CPAP since my pressure is so high. 5 or 6 hours would be acceptable. Medistrom failed me (last 1 hour). MaxOak K5 maybe ok but hard to tell (well for me anyway). Hoping someone can help. I am open to suggestions.

Thank you.
welcome to the zoo! here's a thread, among many, regarding batteries and cpap-

viewtopic.php?f=1&t=177386&p=1327497&hi ... y#p1327497

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by palerider » Thu Nov 14, 2019 1:51 pm

Bern wrote:
Thu Nov 14, 2019 12:31 pm
Having problems finding a back-up power for my CPAP since my pressure is so high. 5 or 6 hours would be acceptable. Medistrom failed me (last 1 hour). MaxOak K5 maybe ok but hard to tell (well for me anyway). Hoping someone can help. I am open to suggestions.

Thank you.
Well, I you want to save money, buy the 12v cable, a mobility s ioter battery, and a battery tender to keep it charged.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Bern
Posts: 2
Joined: Thu Nov 14, 2019 12:21 pm

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by Bern » Thu Nov 14, 2019 2:07 pm

Thank you so much. A follow-up question if I may. I keep reading 12V (and I saw on the resmed chart). But on the CPAP "storage and travel..." sticker it reads DC24V and 3.75A . I'm pretty much stupid with tech tech stuff. The 12V cable is ok to use?

I apologizes if my question makes no sense. I'm not sure how to properly word it.

Side note: I'm the type of person who is willing to spend more if it means less hassle :-)

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by palerider » Thu Nov 14, 2019 4:01 pm

Bern wrote:
Thu Nov 14, 2019 2:07 pm
Thank you so much. A follow-up question if I may. I keep reading 12V (and I saw on the resmed chart). But on the CPAP "storage and travel..." sticker it reads DC24V and 3.75A . I'm pretty much stupid with tech tech stuff. The 12V cable is ok to use?

I apologizes if my question makes no sense. I'm not sure how to properly word it.

Side note: I'm the type of person who is willing to spend more if it means less hassle :-)
Yes, because it converts 12v to the 24v the unit needs.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by Okie bipap » Thu Nov 14, 2019 4:01 pm

ResMed machines do operate on 24 volt DC. The 12 volt cable for the ResMed machines step the voltage up to 24 volts. Responics machines, on the other hand, operate on 12 volts DC.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by palerider » Thu Nov 14, 2019 6:35 pm

Okie bipap wrote:
Thu Nov 14, 2019 4:01 pm
ResMed machines do operate on 24 volt DC. The 12 volt cable for the ResMed machines step the voltage up to 24 volts. Responics machines, on the other hand, operate on 12 volts DC.
But still require a special cable...

In the end, it doesn't really matter, you've got one end of a cable that hooks to 12v, and the other end that hooks to the machine... magic happens somewhere along the cable.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ragtopcircus
Posts: 220
Joined: Mon Sep 09, 2019 11:55 am

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by ragtopcircus » Fri Nov 15, 2019 11:30 am

I believe the Resmed DC-DC can operate from 12V or 24V in because it is a boost-buck converter.

Using two 12V batteries in series will cut the current in half. Since batteries are generally less efficient at high current, using two in series should more than double the run time, especially at higher pressure.

I don’t know if the 3rd party DC-DC is also rated for 24V in.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: BiPAP VAuto 10-25 PS 4.4, and wherever my tempestuous turbinates take me ....

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Back up battery for Air Sense 10 with a pressure of 20

Post by palerider » Fri Nov 15, 2019 1:35 pm

ragtopcircus wrote:
Fri Nov 15, 2019 11:30 am
because it is a boost-buck converter.
As an aside, I haven't seen any evidence that it's a buck converter, just a boost converter,

If it were a buck converter, it'd be rated for higher than 24 volts, too.

_________________