Humidifier almost empty after 1.5 hours of sleep

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
MapleLeaf84
Posts: 1
Joined: Sat Oct 26, 2019 9:23 pm

Humidifier almost empty after 1.5 hours of sleep

Post by MapleLeaf84 » Sat Oct 26, 2019 9:43 pm

Hi guys, I'm trialing a ResMed AirSense 10 with a full face mask. Last night I took it off in my sleep after 1 hour and 19 minutes. The humidifier was full with distilled water when I went to sleep, but after just that short amount of time, it's almost empty. I keep my room as cold as possible, could this have caused it? Also, I woke up quite congested. I have noticed previously that the depletion rate of the water varies, but usually even after 8 hours there's still a safe amount left.

Please help, I'm scared to use it again because I don't want ruin the demo machine if the humidifier runs totally dry.

Thanks!

Edit: I just remembered another thing: for the last couple of weeks the machine has been stowed away in its case, initially due to being ill and then to prepare for and conduct another home sleep study (the Respiratory Therapist advised me not to use CPAP for a while before the test because he said CPAP might lower the AHI, and I needed as high an AHI as possible for insurance coverage). Could this also have caused it, being unused for a couple of weeks? I now remember there being some sort of a warmup feature on the machine, should I have used it?

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Humidifier almost empty after 1.5 hours of sleep

Post by JDS74 » Sun Oct 27, 2019 3:14 am

Just speculating:
Cold air is dryer than warm air and would cause a higher evaporation rate in the humidifier. It shouldn't be this dramatic though. Are you experiencing rainout (excessive condensation) in your mask?

Check to see if there is a tiny leak in the humidifier tank or signs of leakage at the location of the machine when you use it.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Julie » Sun Oct 27, 2019 5:46 am

I believe you are congested from too much humidity and the cold temps don't help. Try warming the room even a little for a few nights and also try a lower (if possible) setting for the humidifier. You're giving yourself a cold!

On the off chance there is a 'micro'' leak in the tank, go over it carefully, fill it and then watch to see if there is any small leak, as well as be sure it's connected properly and not off kilter leaving a seeping 'seam' somewhere.

Cruiser1234
Posts: 113
Joined: Sat Oct 12, 2019 8:32 pm
Location: Ontario, Canada.

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Cruiser1234 » Sun Oct 27, 2019 5:54 am

As a reference, on both my ASV and autoset, it takes roughly a week of sleeping to empty from the max fill line. I keep my humidity setting on 4 and the indoors is around 50% humidity / 22C.

_________________
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Resmed Aircurve 10 ASV

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Dog Slobber » Sun Oct 27, 2019 7:00 am

  1. You forgot to top it off last night
  2. There is a heat source near your device
  3. There is a small leak
  4. When you took your mask off, the machine was still blowing
Battery Backup: EcoFlow Delta 2

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Humidifier almost empty after 1.5 hours of sleep

Post by chunkyfrog » Sun Oct 27, 2019 8:03 am

You only THOUGHT you refilled the tank--(it can happen)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Humidifier almost empty after 1.5 hours of sleep

Post by zonker » Sun Oct 27, 2019 11:35 am

chunkyfrog wrote: ↑
Sun Oct 27, 2019 8:03 am
You only THOUGHT you refilled the tank--(it can happen)
i would be embarrassed to admit how many times that's happened to me.
:oops:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Humidifier almost empty after 1.5 hours of sleep

Post by chunkyfrog » Sun Oct 27, 2019 11:49 am

zonker wrote: ↑
Sun Oct 27, 2019 11:35 am
chunkyfrog wrote: ↑
Sun Oct 27, 2019 8:03 am
You only THOUGHT you refilled the tank--(it can happen)
i would be embarrassed to admit how many times that's happened to me.
:oops:
Guilty. Several times.
If I wake up at 3 AM, smelling dry tank funk, yeah, I know.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Humidifier almost empty after 1.5 hours of sleep

Post by djams » Sun Oct 27, 2019 12:55 pm

We enjoy a cold room too. Good for sleeping.

What's the humidity in the bedroom? We will be getting out the cold air humidifier when the temperatures drop to the point we have to leave heat running all the time.

Out of curiosity, what are your settings?

I like cool air from the machine. I use manual/4 and hose temp off. When this starts causing rainout in winter I put a hose cozy on and have to turn on the heated hose.

Didclaimer: like mask advice, humility advice borders on being useless because we're all different.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Janknitz » Wed Oct 30, 2019 2:56 pm

Do you mouth breathe? That can empty the chamber really fast, whether you are in a FF mask or not. It just can't keep up with the drying of the mucosa in your mouth.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Legrunt » Wed Oct 30, 2019 6:28 pm

Janknitz wrote: ↑
Wed Oct 30, 2019 2:56 pm
Do you mouth breathe? That can empty the chamber really fast, whether you are in a FF mask or not. It just can't keep up with the drying of the mucosa in your mouth.
How does that affect the rate of evaporation?
How does the machine know if it is preventing the drying of nasal vs oral mucosa?

As far as I know, only the rate of air flowing over the water determines the rate of evaporation (and the temperature of the water itself)
and mouth or nose, your lung capacity remains the same.

The only thing I can think of is, if you are on a nasal mask, or pillows, your mouth falls open, the air rushes out and keeps rushing out in a huge leak for an extended amount of time.

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Humidifier almost empty after 1.5 hours of sleep

Post by palerider » Wed Oct 30, 2019 7:04 pm

Legrunt wrote: ↑
Wed Oct 30, 2019 6:28 pm
Janknitz wrote: ↑
Wed Oct 30, 2019 2:56 pm
Do you mouth breathe? That can empty the chamber really fast, whether you are in a FF mask or not. It just can't keep up with the drying of the mucosa in your mouth.
How does that affect the rate of evaporation?
Because more air flows across the water, thus increasing evaporation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Legrunt
Posts: 31
Joined: Thu Jun 27, 2019 6:56 pm

Re: Humidifier almost empty after 1.5 hours of sleep

Post by Legrunt » Thu Oct 31, 2019 6:19 pm

palerider wrote: ↑
Wed Oct 30, 2019 7:04 pm
Legrunt wrote: ↑
Wed Oct 30, 2019 6:28 pm
Janknitz wrote: ↑
Wed Oct 30, 2019 2:56 pm
Do you mouth breathe? That can empty the chamber really fast, whether you are in a FF mask or not. It just can't keep up with the drying of the mucosa in your mouth.
How does that affect the rate of evaporation?
Because more air flows across the water, thus increasing evaporation.
I'm still unclear. You mean to say I will move more air by breathing through my mouth rather than my nose? I thought rate of respiration was only dependant on lung capacity?
Unless the nose is obstructed... But then thats a different problem altogether no?

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Humidifier: Philips Respironics DreamStation Go Heated Humidifier
Additional Comments: Uses a Dreamwear hybrid full face mask for travel as it packs a lot flatter than the F20. But it's a little noisier.
Derek

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Humidifier almost empty after 1.5 hours of sleep

Post by palerider » Thu Oct 31, 2019 7:12 pm

Legrunt wrote: ↑
Thu Oct 31, 2019 6:19 pm
palerider wrote: ↑
Wed Oct 30, 2019 7:04 pm
Legrunt wrote: ↑
Wed Oct 30, 2019 6:28 pm
Janknitz wrote: ↑
Wed Oct 30, 2019 2:56 pm
Do you mouth breathe? That can empty the chamber really fast, whether you are in a FF mask or not. It just can't keep up with the drying of the mucosa in your mouth.
How does that affect the rate of evaporation?
Because more air flows across the water, thus increasing evaporation.
I'm still unclear. You mean to say I will move more air by breathing through my mouth rather than my nose? I thought rate of respiration was only dependant on lung capacity?
Unless the nose is obstructed... But then thats a different problem altogether no?
I thought that Janknitz was referring to mouth *leaking*... and higher leaks would cause greater evaporation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR