Anyone have suggestions for pillows? I feel like I am getting really close to figuring out the machine and mask, but I can't seem to get comfortable with a pillow. Open to any suggestions.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

User avatar
jimbud
Posts: 1474
Joined: Sat Dec 01, 2018 8:19 pm

Re: Newbie here, just want to some help and encouragement

Post by jimbud » Mon Nov 04, 2019 10:16 am

Would help if you would be more specific.
What is your major problem with the ones you have tried.
What do you think would be the ideal pillow.
Ect.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Mon Nov 04, 2019 10:26 am

jimbud wrote:
Mon Nov 04, 2019 10:16 am
Would help if you would be more specific.
What is your major problem with the ones you have tried.
What do you think would be the ideal pillow.
Ect.

JPB
So I have had a Serta memory foam pillow right now. It has been my go to for several years, but I have started to get neck pain, and I feel like I sink to much into it. I would like a pillow that is has nice softness, but is pretty high. I really like memory foam, so if that pillow exist, that would be awesome. I like to sink a bit into the pillow, but it seems everything I have tried lately, I sink all the way down.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

User avatar
jimbud
Posts: 1474
Joined: Sat Dec 01, 2018 8:19 pm

Re: Newbie here, just want to some help and encouragement

Post by jimbud » Mon Nov 04, 2019 11:00 am

I tried several pillows before I settled on the Buckwheat hull pillow. I liked it so much I bought another.
It is like sleeping on a soft sack of cement. You can form it the way you want and it stays that way (more or less) for the rest of the night.

You are not going to sink in very much though.
It isn't for everyone.
I have very little mask leakage with it.

Pillow are like masks. Very YMMV.

Good luck.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Wed Nov 06, 2019 2:08 pm

jimbud wrote:
Mon Nov 04, 2019 11:00 am
I tried several pillows before I settled on the Buckwheat hull pillow. I liked it so much I bought another.
It is like sleeping on a soft sack of cement. You can form it the way you want and it stays that way (more or less) for the rest of the night.

You are not going to sink in very much though.
It isn't for everyone.
I have very little mask leakage with it.

Pillow are like masks. Very YMMV.

Good luck.

JPB
Thanks! Sorry for the late reply, did not see the notification. I am really not sure what my issue is, but I think my pillow has seen its better days. Maybe I will look into the Buckwheat hull pillow.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Wed Nov 06, 2019 2:16 pm

So I am really trying to get a hold of why I am taking my mask off at night, and what is waking me up. I assume it's events that are causing me to wake up, but I wanted to see what you guys think. This is from last night, and I felt like I woke up about 5 times. Some are listed here with the mask coming off, but some I never took the mask off. I am curious is an adjustment to my auto setting needed. I am running the min at 9 and it seems really good for getting me to sleep, and should that be up to stop events from happening? Also should my max go above 13cm? 13 is what my doc set the CPAP at, and I was getting terrible results from CPAP only. Decided to try auto, and got better results, but they are puzzle. Last week I went on vacation, and had two great nights sleep with the mask, but I came back last Thursday, and it looks like I took so many steps back without changing much of anything.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask
Attachments
screenshot-20191106-151429.png
10/30/19 second best night of sleep
screenshot-20191106-151429.png (735.69 KiB) Viewed 6242 times
screenshot-20191106-151416.png
10/29/19 the best sleep I have gotten
screenshot-20191106-151416.png (779.94 KiB) Viewed 6242 times
screenshot-20191106-151356.png
11/5/19 Not so great sleep. 5 awakenings
screenshot-20191106-151356.png (828.35 KiB) Viewed 6242 times

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Fri Nov 08, 2019 7:25 am

I am really struggling, and I can't seem to find the fix. So last night I decided I would up the pressure min to 11.5 cm and the raise the max to 15cm. I am desperately trying to keep my mask on and get a good nights rest.

It was kind of a mess last night. I woke up every time you see the break in flow, and it appears there are events happening that wake me up. I can't seem to get them under control, and I thought adding the APAP to the puzzle would help. If someone could tell me if my thinking is correct or not (most likely not). From the data I am getting, it appears to me the machine is not ramping up pressure enough to fix the events, and then I am waking up. Isnt that was auto is for? Last night each time before I took it off, there were either 2 or 3 events without pressure being raised, and then a final event that caused me to take the mask off. Am I missing something? Should the min pressure be raised more? I just cant understand what the machine is doing.

Below is the full nights look, and I have zoomed in to see the events that caused the mask to be removed. Any help would be great.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask
Attachments
screenshot-20191108-081636.png
Second event with removal of mask
screenshot-20191108-081636.png (792.21 KiB) Viewed 6217 times
screenshot-20191108-081622.png
First event with removal of mask
screenshot-20191108-081622.png (783.41 KiB) Viewed 6217 times
screenshot-20191108-081549.png
Full view of last night
screenshot-20191108-081549.png (779.46 KiB) Viewed 6217 times

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Fri Nov 08, 2019 7:27 am

This is the last event that caused me to have my mask off for an extended period of time.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask
Attachments
screenshot-20191108-081655.png
Last event that caused the biggest length of time without the mask
screenshot-20191108-081655.png (898.01 KiB) Viewed 6217 times

User avatar
Bertha deBlues
Posts: 317
Joined: Sun Aug 20, 2017 8:36 am

Re: Newbie here, just want to some help and encouragement

Post by Bertha deBlues » Fri Nov 08, 2019 9:09 am

Wish I knew enough to be able to advise. Hopefully some helpful users will respond. Hang in there.
And in the end the love you take is equal to the love you make. - Paul McCartney

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Fri Nov 08, 2019 9:14 am

Bertha deBlues wrote:
Fri Nov 08, 2019 9:09 am
Wish I knew enough to be able to advise. Hopefully some helpful users will respond. Hang in there.
Thanks! I feel very defended right now. Seems like nothing is helping.

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

User avatar
Pugsy
Posts: 65140
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie here, just want to some help and encouragement

Post by Pugsy » Fri Nov 08, 2019 9:43 am

How can you be certain that a lone hyponea 20 minutes prior to a wake up and mask removal was the cause of the wake up?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Fri Nov 08, 2019 9:49 am

Pugsy wrote:
Fri Nov 08, 2019 9:43 am
How can you be certain that a lone hyponea 20 minutes prior to a wake up and mask removal was the cause of the wake up?
Not saying it is, but look at my flat line breathing when I take the mask off. Doesn't that represent an event happening? To me I am seeing this as an event that wakes me up and I remove the mask as its happening.

Also one of my biggest issues before getting diagnosed with sleep apnea was night terrors. Doctor said I was losing oxygen when I was having apnea events and it was causing me to have the terrors. I had one for the first time since I first got the machine. This is telling me something is not working properly. Just my unprofessional/ unexperienced opinion. If I am wrong, what is a more experienced persons opinion on what is going on?

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

User avatar
Pugsy
Posts: 65140
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie here, just want to some help and encouragement

Post by Pugsy » Fri Nov 08, 2019 9:56 am

Okay, you are talking about the flat line immediate to the mask removal.
See that big breath immediately before the flat line? That's an arousal or awake breath and then the flat line is either awake/arousal evidence of the machine not sensing breathing either because the mask is off and/or leak is happening.
You weren't asleep when that flat line shows up...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Brounmoney
Posts: 83
Joined: Tue Oct 22, 2019 2:19 pm

Re: Newbie here, just want to some help and encouragement

Post by Brounmoney » Fri Nov 08, 2019 9:59 am

Pugsy wrote:
Fri Nov 08, 2019 9:56 am
Okay, you are talking about the flat line immediate to the mask removal.
See that big breath immediately before the flat line? That's an arousal or awake breath and then the flat line is either awake/arousal evidence of the machine not sensing breathing either because the mask is off and/or leak is happening.
You weren't asleep when that flat line shows up...
So then you are saying something else woke me up before I took the mask off correct?

_________________
Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Also trying the Dreamwear full face mask

User avatar
Pugsy
Posts: 65140
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie here, just want to some help and encouragement

Post by Pugsy » Fri Nov 08, 2019 10:00 am

You need to zoom in much closer on the time where you take the mask off...maybe 3 to 5 minute segments.

Zoom in much closer on the area where you take the mask off.
See the spike immediately prior to the flat line....that's arousal/awake breathing.