Airsense 10 question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
xUnderCoverHoBo
Posts: 15
Joined: Sat Oct 12, 2019 8:35 am

Airsense 10 question

Post by xUnderCoverHoBo » Sun Oct 13, 2019 8:41 am

So I know I have been posting tons of questions but I found this forum the other day and I’m absolutely loving the feedback I have received.

One more question in regards to the humidifier in my airsense 10. It has lines on it showing the max amount of water and then 2 other lines. Since I have been drooling a lot can I put water in that is below that bottom line or should that be the minimal amount of water?


Thank you all again so much for your help as I transition into cpap therapy! I am also in the process of updating my signature to show my equipment

_________________
MachineMask

User avatar
LSAT
Posts: 13395
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Airsense 10 question

Post by LSAT » Sun Oct 13, 2019 10:04 am

Doing that will cause your humidifier to run out of water more quickly...It's possible that it may run out during the night and you will get the smell of burning plastic and could damage the humidifier when it runs dry. doing what you suggest will have no effect on your drooling.

User avatar
babydinosnoreless
Posts: 2372
Joined: Fri Nov 02, 2018 2:53 pm

Re: Airsense 10 question

Post by babydinosnoreless » Sun Oct 13, 2019 10:31 am

Its helpful to put your questions in one thread so that everyone can follow along. Starting a new thread each time gets confusing. :)