so clean nueltrilizer

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 10:15 am

how often should you use the so clean neutralizer and should you soak equipment in it for long time or just a quick soak and clean..Thank you

User avatar
LSAT
Posts: 13411
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: so clean nueltrilizer

Post by LSAT » Sat Oct 12, 2019 10:28 am

Aunkster wrote:
Sat Oct 12, 2019 10:15 am
how often should you use the so clean neutralizer and should you soak equipment in it for long time or just a quick soak and clean..Thank you
Never! Manufacturers recommend using soap and water only.

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 10:42 am

no for the SoClean machine you have to use the neutralizer..

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: so clean nueltrilizer

Post by Dog Slobber » Sat Oct 12, 2019 10:56 am

Aunkster wrote:
Sat Oct 12, 2019 10:42 am
no for the SoClean machine you have to use the neutralizer..
Many of us don't believe the SoClean is necessary, and even potentially harmful to the user and equipment.

CPAP manufacturers don't recommend SoClean or any sanitizing device.
Battery Backup: EcoFlow Delta 2

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 11:24 am

I do know there are many of you here , but there are many who do believe and I am one of them.. sorry if my posts offends you , I am not trying , I’m just looking for answers from the not so negative

User avatar
Pugsy
Posts: 65388
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: so clean nueltrilizer

Post by Pugsy » Sat Oct 12, 2019 11:33 am

I take it that SoClean calls their cleaning solution "neutralizer". It's use is to get the grime, oil, snot, slobber off your mask.
So how long to use it...as long as it takes to get rid of the grime, oil, snot, slobber, etc is probably sufficient. That's all it does.

Most of the forum members here don't use the SoClean...much less the SoClean cleaning product because soap and water works just as well.
So....maybe your question would be best answered by the device manufacturer and not by a bunch of people here who don't believe in using the device for any number of reasons.

What are the directions on the bottle of the "neutralizer" you got?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 11:35 am

I get it I really do but there are people who do use ,,so why comment just to be negative..

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: so clean nueltrilizer

Post by Dog Slobber » Sat Oct 12, 2019 11:48 am

Aunkster wrote:
Sat Oct 12, 2019 11:35 am
I get it I really do but there are people who do use ,,so why comment just to be negative..
What part of these two responses are negative?
LSAT wrote:
Sat Oct 12, 2019 10:28 am
Never! Manufacturers recommend using soap and water only.
Dog Slobber wrote:
Sat Oct 12, 2019 10:56 am
Many of us don't believe the SoClean is necessary, and even potentially harmful to the user and equipment.

CPAP manufacturers don't recommend SoClean or any sanitizing device.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11486
Joined: Fri Jun 19, 2015 4:36 pm

Re: so clean nueltrilizer

Post by zonker » Sat Oct 12, 2019 11:49 am

Aunkster wrote:
Sat Oct 12, 2019 11:35 am
I get it I really do but there are people who do use ,,so why comment just to be negative..
on this forum, we have a strong bias towards facts. we're NOT trying to be negative. that's your interpretation. but if you bring up the topic here, expect the "negativity" to commence and continue.

there ARE a very few here who believe so clean helps them. and more power to them. perhaps they will answer your question.

if the attitude regarding this topic gets on your nerves, i suggest you try apneaboard.

they will give you all kinds of sweetness and light, just not as much facts.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 11:50 am

Thank you ,thats a reply post I can use...

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: so clean nueltrilizer

Post by chunkyfrog » Sat Oct 12, 2019 11:54 am

I am sorry you got "taken to the cleaners", when ordinary dish soap/ hot water,
or even just an unscented baby wipe would have done the job just fine--way cheaper.
Some rabid proponents of soclean repeatedly troll this forum.
Sorry you were mistaken for one of THEM.
Welcome to the lifetime hosers club.
Sleep well, and blessings on you.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 11:55 am

Dog Slobber wrote:
Sat Oct 12, 2019 11:48 am
Aunkster wrote:
Sat Oct 12, 2019 11:35 am
I get it I really do but there are people who do use ,,so why comment just to be negative..
What part of these two responses are negative?
LSAT wrote:
Sat Oct 12, 2019 10:28 am
Never! Manufacturers recommend using soap and water only.
Dog Slobber wrote:
Sat Oct 12, 2019 10:56 am
Many of us don't believe the SoClean is necessary, and even potentially harmful to the user and equipment.

CPAP manufacturers don't recommend SoClean or any sanitizing device.
to me the part about being harmful ,and the neutralizer not needed..I know everyone has opinions but before I started using the machine I was sick all the time..my pulmonologist said lets give the machine a try ,they gave me one at no cost and I have been using it for about 8 months and feel like a new person..thats all..
im sorry for offending anyone and as a matter of fact I have received a few private messages from people who say they are glad I spoke up..im a good positive person and I am kind to everyone..I guess thats my story

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: so clean nueltrilizer

Post by Wulfman... » Sat Oct 12, 2019 11:58 am

Aunkster wrote:
Sat Oct 12, 2019 11:35 am
I get it I really do but there are people who do use ,,so why comment just to be negative..
We aren't "negative", just trying to warn the gullible not to be hornswoggled or bamboozled into buying these money-sucking, worthless products.
It's just amazing how people would be taken in by these charlatans. They're "snake oil" salesmen who have no consciences.
They don't "clean" anything but the wallets and bank accounts.

search.php?keywords=%22so+clean%22


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: so clean nueltrilizer

Post by chunkyfrog » Sat Oct 12, 2019 11:59 am

"Neutralizer" is nothing more than ordinary cieaner,
basically to wipe off the schmutz that the $300 device cannot touch.
--And another cash cow.
Cute, but not very honest.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her
Last edited by chunkyfrog on Sat Oct 12, 2019 12:02 pm, edited 1 time in total.

Aunkster
Posts: 37
Joined: Sat Jan 11, 2014 7:06 pm

Re: so clean nueltrilizer

Post by Aunkster » Sat Oct 12, 2019 12:01 pm

well then I guess I should be happy that I didn't pay a cent and it was given to me by my pulmonologist