Question about MyAir readings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Blmkso
Posts: 3
Joined: Mon Oct 07, 2019 1:08 pm

Question about MyAir readings

Post by Blmkso » Mon Oct 07, 2019 2:17 pm

I’ve been using ResMed 10 for several months now with a nasal mask. My average AHI numbers are around 1.5 -3.0 but leakage is averaging 25-30. It seems like when the AHI is lowest , the leakage is highest. I have tried a full face mask but AHI’s are higher and of course leakage is real low. Is the AHI reading the important one and does anyone know if there is a relation between high leakage and low AHI’s. Thanks for any info.

User avatar
LSAT
Posts: 13405
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Question about MyAir readings

Post by LSAT » Mon Oct 07, 2019 3:02 pm

Leakage is the mask...If you are sleeping well and your AHI is low, I would ignore the leaks. Your machine compensates for most leaks.... No relationship between low leaks and higher AHI. 1.5-3.0 is not significant in my opinion.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question about MyAir readings

Post by palerider » Mon Oct 07, 2019 3:06 pm

Blmkso wrote:
Mon Oct 07, 2019 2:17 pm
I’ve been using ResMed 10 for several months now with a nasal mask. My average AHI numbers are around 1.5 -3.0 but leakage is averaging 25-30. It seems like when the AHI is lowest , the leakage is highest. I have tried a full face mask but AHI’s are higher and of course leakage is real low. Is the AHI reading the important one and does anyone know if there is a relation between high leakage and low AHI’s. Thanks for any info.
Myair is useless when trying to solve problems, get Oscar.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65338
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question about MyAir readings

Post by Pugsy » Mon Oct 07, 2019 3:11 pm

MyAir reports what is called a 95% leak number. It is NOT an overall average leak number.
95% numbers just mean where a person was AT or BELOW for 95% of the time. People often don't understand the "or below" part of the definition.

For almost 95% of the night night you were most likely below the 24 L/min red line where large leak territory starts. The relatively short periods of time you might have been above 24 L/min aren't enough to really impact the overall AHI that much.
The machine won't start missing flagging of events until somewhere around 35 L/min excess leak.

As long as the leaks didn't wake me up.....I wouldn't worry about them at this point if this is the worst they get.

95% numbers are easily elevated making things look worse than they really might be. Get the available software and see the detailed report and you will understand what I mean. Example and more details explained in this thread.
viewtopic.php?f=1&t=108724&p=1036669&hi ... t#p1036669

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.