Clusters of hypopneas with no flow limitation

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mannsch
Posts: 25
Joined: Sat Aug 17, 2019 4:56 am

Clusters of hypopneas with no flow limitation

Post by mannsch » Wed Sep 18, 2019 3:55 am

I am getting big clusters of hypopneas throughout the night where the breathing pattern on the flow rate gets much smaller, then followed by a big breath. Then repeat several times.

See here: https://prnt.sc/p7o5le
And here: https://prnt.sc/p7o32i

There is no flow limitation shown on the charts that correlates with these patterns.

Anyone else get these? Any tips to deal with them?

I am using the Resmed Aircurve 10 ASV.

_________________
Mask

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Clusters of hypopneas with no flow limitation

Post by Pugsy » Wed Sep 18, 2019 7:10 am

Are you for sure solidly asleep when things seem to go off?.
You do extremely well for a little over half the sleep session and then things get wacky. That makes me wonder what changed?
Awake tossing and turning? Sleep position change?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mannsch
Posts: 25
Joined: Sat Aug 17, 2019 4:56 am

Re: Clusters of hypopneas with no flow limitation

Post by mannsch » Wed Sep 18, 2019 9:20 am

I can't say for sure if I was awake or not. I feel like I did manage to have a long stretch of uninterrupted sleep last night and a lot of these events are within that I think.

Maybe I should try to record myself on video for a few nights and see if I can correlate these events with any sleep positions or disturbances.

_________________
Mask

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Clusters of hypopneas with no flow limitation

Post by Pugsy » Wed Sep 18, 2019 9:30 am

The two variables that I can think of
not quite asleep or a change in sleep position.
Your breathing pattern during the bulk of the ugly stuff that you show doesn't look like asleep breathing to me.
Looks like tossing and turning breathing to me.

Learn how to zoom in on your flow rate to get an idea of awake breathing vs asleep breathing and you can see what I mean.
The vidoes here explain how to do it.
http://freecpapadvice.com/sleepyhead-free-software

If you were asleep....
It is of course entirely possible that those hyponeas aren't obstructive in nature and if that's the case the absence of any FLs doesn't mean much because we wouldn't expect to see them with central in nature events. Hyponeas aren't always obstructive in nature.

If you can video yourself to rule in or out the sleeping on your back thing...that would help with your detective work figuring out what is happening here if you are indeed asleep.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Clusters of hypopneas with no flow limitation

Post by Jay Aitchsee » Wed Sep 18, 2019 11:36 am

Hypopneas and Obstructive apneas are not necessarily accompanied by Flow Limitations.
Hypopneas and Apneas are a reductions in flow while Flow Limitations generally refer to the shape of the flow wave. Flow LImitations generally result in a flattened inspiration wave top. Flows can be diminished without being flattened resulting in a hypopnea without a corresponding Flow Limitation.
Hypopneas can be "central" in nature where breathing becomes shallow resulting in reduced flow amplitude without exhibiting the characteristic flattening of Flow Limitations.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

pouetance
Posts: 3
Joined: Mon Jul 15, 2019 6:46 pm

Re: Clusters of hypopneas with no flow limitation

Post by pouetance » Wed Sep 18, 2019 1:08 pm

That's REM sleep hypopnea I believe (look like central hypopneas actually if such things exist). Looks at the timing of the clusters. They are all almost ninety minutes apart and as the night progress, you get more events.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Clusters of hypopneas with no flow limitation

Post by palerider » Wed Sep 18, 2019 1:54 pm

mannsch wrote:
Wed Sep 18, 2019 3:55 am
I am getting big clusters of hypopneas throughout the night where the breathing pattern on the flow rate gets much smaller, then followed by a big breath. Then repeat several times.

See here: https://prnt.sc/p7o5le
And here: https://prnt.sc/p7o32i

There is no flow limitation shown on the charts that correlates with these patterns.

Anyone else get these? Any tips to deal with them?

I am using the Resmed Aircurve 10 ASV.
Flow limitation and hypopnea are not related, you can easily have one without the other.

https://www.youtube.com/watch?v=-gie2dhqP2c

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Clusters of hypopneas with no flow limitation

Post by Jay Aitchsee » Wed Sep 18, 2019 2:00 pm

pouetance wrote:
Wed Sep 18, 2019 1:08 pm
(look like central hypopneas actually if such things exist)
http://jcsm.aasm.org/ViewAbstract.aspx?pid=31260
The AASM Manual for the Scoring of Sleep and Associated Events: Rules, Terminology and Technical Specifications (AASM Scoring Manual) provides the option of scoring hypopneas as obstructive or central.1,2 Obstructive hypopneas are those with evidence of increased airflow flattening, chest/ abdominal effort belt paradox, or snoring during the event. If none of these are present the hypopnea is classified as central.
Jay Aitchsee wrote:
Wed Sep 18, 2019 11:36 am
Hypopneas can be "central" in nature where breathing becomes shallow resulting in reduced flow amplitude without exhibiting the characteristic flattening of Flow Limitations.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Clusters of hypopneas with no flow limitation

Post by raisedfist » Wed Sep 18, 2019 4:09 pm

The clinical guide for ResMed under the ASV section says to raise the EPAP:

"Any obstructive apneas, hypopneas or RERAs? Increase EPAP by 1 cm H2O every 20 minutes until obstructive events are eliminated"


Not all cases are textbook, but worth a shot to raise the EPAP min and see what happens to the hypopneas. It won't cause any harm either way.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Clusters of hypopneas with no flow limitation

Post by Jas_williams » Thu Sep 19, 2019 12:07 am

They don’t look obstructive to me but centrally based as the amplitude goes down although the ASV is blowing too it’s max settings to encourage the small breath that’s there. I would raise the max PS to its max setting.


Beyond that not sure what else you can do.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

mannsch
Posts: 25
Joined: Sat Aug 17, 2019 4:56 am

Re: Clusters of hypopneas with no flow limitation

Post by mannsch » Fri Sep 20, 2019 5:21 am

Thanks for the help, everyone.

I tried raising the PS Max to the maximum setting of 19, so the pressure was going up to 25 during these shallow breaths.

Here's last night, full chart: https://prnt.sc/p8nk5a

Here's a zoom-in one of the hypopnea clusters: https://prnt.sc/p8nlgn

The machine clearly ramps up to the maximum pressure of 25, but it still doesn't suffice to force breaths.

I'll try raising the min EPAP next, although I'm not sure how that would help if these hypopneas are central in nature.

I do feel like I am sleeping better with the ASV, though. I'm not waking up nearly as much during the night and I am having clear, vivid dreams every night.

_________________
Mask

idlewire
Posts: 35
Joined: Sun May 26, 2019 6:07 pm

Re: Clusters of hypopneas with no flow limitation

Post by idlewire » Fri Sep 20, 2019 1:58 pm

I would lower your PS Min to the default of 3. A min PS of 6 is enough to actually induce centrals in someone who doesn't normally have them. If your machine is over-ventilating you at baseline, you'll be more likely to have central events.

mannsch
Posts: 25
Joined: Sat Aug 17, 2019 4:56 am

Re: Clusters of hypopneas with no flow limitation

Post by mannsch » Sat Sep 21, 2019 2:02 am

idlewire wrote:
Fri Sep 20, 2019 1:58 pm
I would lower your PS Min to the default of 3. A min PS of 6 is enough to actually induce centrals in someone who doesn't normally have them. If your machine is over-ventilating you at baseline, you'll be more likely to have central events.
That's interesting, thanks!

I tried lowering the min EPAP and min PS settings:

Min EPAP: 4
Max EPAP: 15
Min PS: 3
Max PS: 20

Last night, I had an AHI of 1.59 which is the lowest since I started the therapy! See here: https://prnt.sc/p906il

Maybe it was excessive ventilation that was causing these central events. Looks like the ASV easily handles the obstructive apneas as needed as I have zero of them now.

_________________
Mask

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Clusters of hypopneas with no flow limitation

Post by Pugsy » Sat Sep 21, 2019 7:17 am

ResMed ASV machines won't ever flag an OA because they don't differentiate between OA and central.
Instead they dump them in the UA basket.
It's just what they do.

You've got a handful of UAs....probably obstructive. Not enough to worry about.

Give yourself more nights at these new settings and see what happens.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mannsch
Posts: 25
Joined: Sat Aug 17, 2019 4:56 am

Re: Clusters of hypopneas with no flow limitation

Post by mannsch » Mon Sep 23, 2019 7:33 am

Still doing really well with the lower pressures -- AHI is consistently much lower than before: https://prnt.sc/p9tzw4

The pressure support now seems to be sufficiently powerful to eliminate most of the hypopneas from before.

However, I am wondering if the big pressure spikes themselves could also be causing me to wake up.

See here: https://prnt.sc/p9u1ee
And here: https://prnt.sc/p9u1se

Anyone know how to deal with this?

_________________
Mask