First night on CPAP, coming up in a few hours

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
HuffDog
Posts: 23
Joined: Fri Aug 02, 2019 7:41 pm

Re: First night on CPAP, coming up in a few hours

Post by HuffDog » Sat Aug 17, 2019 9:50 am

Julie wrote:
Sat Aug 17, 2019 8:19 am
I would try (ramp off) EPR of 2 - you may not really need it and 3 may interfere with other features without your realizing it. I would bump min. setting to 8, and max to at least 15 if not 20. And remember the machine should adapt to your breathing pattern, not the other way around (people unconsciously try to match breathing to machine and never get comfortable).
I may try adjusting the EPR and pressures to your recommendations then. But even at EPR 3, it feels slightly uncomfortable to exhale. I did try 2, 1 and off this morning just to see how it feels. Even more uncomfortable at the lower numbers and off.

I presume that most of my initial issues regarding the sensation of not being able to breathe comfortably are due to me being new to this. I realize there is an adjustment period here. Both mentally and getting the machine dialed in.

I'm not deterred, but I'm impatient and wish everything worked perfectly right now! Kidding :)

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure min/max 7/13. Flex=3

User avatar
Grace~~~
Posts: 662
Joined: Tue May 10, 2016 3:27 pm
Location: Florida

Re: First night on CPAP, coming up in a few hours

Post by Grace~~~ » Sat Aug 17, 2019 9:57 am

I wore my mask A LOT in the very beginning when getting used to it. I got very desensitized to the feeling, and the dreamwear is so comfortable to me.

I added cotton balls by the air intake valve of different scented oils I enjoy. (Like spearmint) and that was a wonderful treat and turned in to long sessions of meditative breathing.

All the extra wear also helped me dial in the most comfortable position and tightness of the straps.

I love my mask now. Even the thought of putting it on makes me sleepy and feels like a luxury of great sleep.

Keep at it. It takes some work and some tweaking but it is SOOOOOO worth it.
Began XPAP May 2016. Autoset Pressure min. 8 / max 15. Ramp off. ERP set at 2. No humidity. Sleepyhead software installed and being looked at daily, though only beginning to understand the data.

User avatar
HuffDog
Posts: 23
Joined: Fri Aug 02, 2019 7:41 pm

Re: First night on CPAP, coming up in a few hours

Post by HuffDog » Sat Aug 17, 2019 10:18 am

Grace~~~ wrote:
Sat Aug 17, 2019 9:57 am
I added cotton balls by the air intake valve of different scented oils I enjoy. (Like spearmint) and that was a wonderful treat and turned in to long sessions of meditative breathing.
Ooh good idea. I have a few essential oils that I use in a vaporizer. Will try this too. Thanks!

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure min/max 7/13. Flex=3

User avatar
HuffDog
Posts: 23
Joined: Fri Aug 02, 2019 7:41 pm

Re: First night on CPAP, coming up in a few hours

Post by HuffDog » Sat Aug 17, 2019 10:21 am

Also, any thoughts on why the pressure goes higher and higher even when I'm not asleep? It went as high as 11 last night. Is it possible I'm just fighting the machine instead of just relaxing and letting it do its job?

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure min/max 7/13. Flex=3

User avatar
babydinosnoreless
Posts: 2363
Joined: Fri Nov 02, 2018 2:53 pm

Re: First night on CPAP, coming up in a few hours

Post by babydinosnoreless » Sat Aug 17, 2019 10:54 am

HuffDog wrote:
Sat Aug 17, 2019 9:39 am
babydinosnoreless wrote:
Sat Aug 17, 2019 9:28 am
Have you worn it during the day when you are not trying to
sleep ? Sit and watch tv, read a book, get used to the machine while you are awake before trying to sleep with it.
Kinda. I just got the machine yesterday afternoon. I wore it for about 30 minutes after I got the machine set up, the mask adjusted, etc. I think doing more of this will be helpful though. Thanks!
You're welcome. :) I see someone recommended the scented cottonballs. I wash my mask in Dr Bronners peppermint soap. This gives it a nice minty scent.

User avatar
Pugsy
Posts: 65127
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: First night on CPAP, coming up in a few hours

Post by Pugsy » Sat Aug 17, 2019 11:37 am

HuffDog wrote:
Sat Aug 17, 2019 10:21 am
any thoughts on why the pressure goes higher and higher even when I'm not asleep?
The machine doesn't know if you are asleep or not. It only measures air flow.
Awake breathing is very irregular when compared to asleep breathing but that awake breathing being irregular can cause the machine to respond like it would if your breathing was irregular while asleep because of apnea events.
You were probably tossing and turning and getting frustrated and maybe sighing a lot...lots of irregular breathing and the machine thought you were asleep having apnea events.
Probably holding your breath some and not realizing it...frustrated most likely.
You got the machine confused. :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
HuffDog
Posts: 23
Joined: Fri Aug 02, 2019 7:41 pm

Re: First night on CPAP, coming up in a few hours

Post by HuffDog » Sat Aug 17, 2019 11:39 am

Pugsy wrote:
Sat Aug 17, 2019 11:37 am
HuffDog wrote:
Sat Aug 17, 2019 10:21 am
any thoughts on why the pressure goes higher and higher even when I'm not asleep?
The machine doesn't know if you are asleep or not. It only measures air flow.
Awake breathing is very irregular when compared to asleep breathing but that awake breathing being irregular can cause the machine to respond like it would if your breathing was irregular while asleep because of apnea events.
You were probably tossing and turning and getting frustrated and maybe sighing a lot...lots of irregular breathing and the machine thought you were asleep having apnea events.
Probably holding your breath some and not realizing it...frustrated most likely.
You got the machine confused. :lol:
I was definitely tossing, turning and sighing with frustration!

I think "relax and be patient" are two things I'll have to learn. :)

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure min/max 7/13. Flex=3

User avatar
zonker
Posts: 11340
Joined: Fri Jun 19, 2015 4:36 pm

Re: First night on CPAP, coming up in a few hours

Post by zonker » Sat Aug 17, 2019 12:58 pm

HuffDog wrote:
Sat Aug 17, 2019 9:50 am


I presume that most of my initial issues regarding the sensation of not being able to breathe comfortably are due to me being new to this. I realize there is an adjustment period here. Both mentally and getting the machine dialed in.

I'm not deterred, but I'm impatient and wish everything worked perfectly right now! Kidding :)
this is SPOT ON! we have pretty much all of us been in this position on night one. and for many following nights too, i'm afraid. i wish i could say to you "'dog, on your 32nd night at 2am, everything will click together". man think of the money i could make with predictions like that, huh? :lol:

but alas this trip is slightly different for each of us. you WILL get used to it, but at your own pace.

but you have an excellent attitude and i'm confident you'll get the hang of it.

continued good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Dog Slobber
Posts: 4255
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: First night on CPAP, coming up in a few hours

Post by Dog Slobber » Sat Aug 17, 2019 1:26 pm

zonker wrote:
Sat Aug 17, 2019 12:58 pm

i wish i could say to you "'dog, on your 32nd night at 2am, everything will click together". man think of the money i could make with predictions like that, huh? :lol:
Holy Crap Zonk, that was bang on. Well, except the time was two hours off, but I suspect that's the time difference between Arizona and Ontario.

Zonk_the_Greek.png
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11340
Joined: Fri Jun 19, 2015 4:36 pm

Re: First night on CPAP, coming up in a few hours

Post by zonker » Sat Aug 17, 2019 3:04 pm

Dog Slobber wrote:
Sat Aug 17, 2019 1:26 pm
zonker wrote:
Sat Aug 17, 2019 12:58 pm

i wish i could say to you "'dog, on your 32nd night at 2am, everything will click together". man think of the money i could make with predictions like that, huh? :lol:
Holy Crap Zonk, that was bang on. Well, except the time was two hours off, but I suspect that's the time difference between Arizona and Ontario.


Zonk_the_Greek.png

i may not know computers, but i know your soul.

mwahahahahaha!!!!!
Last edited by zonker on Tue Aug 20, 2019 5:01 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
HuffDog
Posts: 23
Joined: Fri Aug 02, 2019 7:41 pm

Re: First night on CPAP, coming up in a few hours

Post by HuffDog » Tue Aug 20, 2019 10:37 am

So i had a few rough nights on my Resmed machine that I purchased out of pocket. I could never get comfortable with exhale on that machine.

Insurance finally got around to sending me a machine, but it's a Dreamstation. Exhale seems more comfortable for me. Current equipment now in my signature.

I actually got a few hours of sleep. Still feel fatigued, maybe a bit more than usual, but not necessarily sleepy. The sleep "seemed" better, but I just didn't get enough hours of it, I believe.

Can anybody take a look at this and let me know your thoughts? Not real good at looking at these yet (I'm learning through the tutorial though), so any advice from the more knowledgeable would be greatly appreciated.

The breaks at 22:45 to 23:20 are me readjusting the mask a few times. I woke up at about 3:20 and was awake until 4:40 or so. Took the mask off and slept for another 2 hours.

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure min/max 7/13. Flex=3
Attachments
screenshot-20190820-112824.png
screenshot-20190820-112824.png (248.42 KiB) Viewed 512 times