Water collecting in hose with higher humidity settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Charlie6067
Posts: 5
Joined: Mon Jul 22, 2019 9:55 pm

Water collecting in hose with higher humidity settings

Post by Charlie6067 » Mon Aug 05, 2019 2:50 pm

ResMed 10. Humidity setting at 6 but mouth still dry. A 7 or 8 setting results in gurgling of moisture in the hose. Any ideas how to correct?
Thanks
Charlie

_________________
MachineMask

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Water collecting in hose with higher humidity settings

Post by Pugsy » Mon Aug 05, 2019 3:06 pm

Read about options for dealing with condensation in the mask or hose here....posts number 2 and 3

viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Water collecting in hose with higher humidity settings

Post by D.H. » Mon Aug 05, 2019 3:35 pm

Try one or more of these:

1) Make sure that the unit is lower than your head.
2) Use a heated hose if you're not already doing so.
3) Lower the humidity setting on your unit.
4) Raise the ambient temperature in the room.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
LSAT
Posts: 13329
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Water collecting in hose with higher humidity settings

Post by LSAT » Mon Aug 05, 2019 3:39 pm

Dry mouth could be from mouth breathing. If your mouth was closed you would not need as much humidity...lower the setting and it may eliminate the water in the hose.

TropicalDiver
Posts: 350
Joined: Sun Nov 04, 2018 11:29 pm

Re: Water collecting in hose with higher humidity settings

Post by TropicalDiver » Mon Aug 05, 2019 3:42 pm

D.H. wrote:
Mon Aug 05, 2019 3:35 pm
Try one or more of these:

1) Make sure that the unit is lower than your head.
2) Use a heated hose if you're not already doing so.
3) Lower the humidity setting on your unit.
4) Raise the ambient temperature in the room.
5). Add an insulating wrap to your house
Machine: Aircurve 10 Vauto (Prior S9 VPAP)
Mask: Quattro Air FFM and AirTouch F20 FFM

User avatar
jimbud
Posts: 1465
Joined: Sat Dec 01, 2018 8:19 pm

Re: Water collecting in hose with higher humidity settings

Post by jimbud » Mon Aug 05, 2019 3:48 pm

TropicalDiver wrote:
Mon Aug 05, 2019 3:42 pm
D.H. wrote:
Mon Aug 05, 2019 3:35 pm
Try one or more of these:

1) Make sure that the unit is lower than your head.
2) Use a heated hose if you're not already doing so.
3) Lower the humidity setting on your unit.
4) Raise the ambient temperature in the room.
5). Add an insulating wrap to your house
:lol: :!:

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
LSAT
Posts: 13329
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Water collecting in hose with higher humidity settings

Post by LSAT » Mon Aug 05, 2019 4:01 pm

jimbud wrote:
Mon Aug 05, 2019 3:48 pm
TropicalDiver wrote:
Mon Aug 05, 2019 3:42 pm
D.H. wrote:
Mon Aug 05, 2019 3:35 pm
Try one or more of these:

1) Make sure that the unit is lower than your head.
2) Use a heated hose if you're not already doing so.
3) Lower the humidity setting on your unit.
4) Raise the ambient temperature in the room.
5). Add an insulating wrap to your house
:lol: :!:
We did that when we changed siding.. :P

User avatar
Pugsy
Posts: 65075
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Water collecting in hose with higher humidity settings

Post by Pugsy » Mon Aug 05, 2019 4:03 pm

The typo police at work again. :lol: :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TropicalDiver
Posts: 350
Joined: Sun Nov 04, 2018 11:29 pm

Re: Water collecting in hose with higher humidity settings

Post by TropicalDiver » Mon Aug 05, 2019 4:11 pm

LSAT wrote:
Mon Aug 05, 2019 4:01 pm
jimbud wrote:
Mon Aug 05, 2019 3:48 pm
TropicalDiver wrote:
Mon Aug 05, 2019 3:42 pm
D.H. wrote:
Mon Aug 05, 2019 3:35 pm
Try one or more of these:

1) Make sure that the unit is lower than your head.
2) Use a heated hose if you're not already doing so.
3) Lower the humidity setting on your unit.
4) Raise the ambient temperature in the room.
5). Add an insulating wrap to your house
:lol: :!:
We did that when we changed siding.. :P
Adding it to the hose would be even better!
Machine: Aircurve 10 Vauto (Prior S9 VPAP)
Mask: Quattro Air FFM and AirTouch F20 FFM

jsmit86
Posts: 264
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Water collecting in hose with higher humidity settings

Post by jsmit86 » Mon Aug 05, 2019 4:44 pm

I'd suggest insulating the house and the hose... just to be sure! :mrgreen:

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
zonker
Posts: 11294
Joined: Fri Jun 19, 2015 4:36 pm

Re: Water collecting in hose with higher humidity settings

Post by zonker » Mon Aug 05, 2019 5:03 pm

Charlie6067 wrote:
Mon Aug 05, 2019 2:50 pm
ResMed 10. Humidity setting at 6 but mouth still dry. A 7 or 8 setting results in gurgling of moisture in the hose. Any ideas how to correct?
Thanks
Charlie
plenty of suggestions for you to try, so far. bare in mind that humidity is one of those things where what works for me may not work for thee. which is why i usually keep quiet about it.

but as someone mentioned, dry mouth might not be down to humidity at all.

i use xylimelts for dry mouth. it's a small lozenge that fits between cheek and teeth, with a bit of tacky stuff that helps it adhere to the teeth. it promotes saliva as the lozenge melts. lasts all night long.

you can find them in the oral care section of drugstore or grocery.

it doesn't get to the cause of dry mouth but does help alleviate it.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: Water collecting in hose with higher humidity settings

Post by D.H. » Mon Aug 05, 2019 5:13 pm

Actually, if you're using a heated hose, I don't think that wrapping it will do anything (except to save a marginal amount of electricity). The hose is heated to a certain temperature, and that same temperature will be reached, wrapping or not.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Water collecting in hose with higher humidity settings

Post by palerider » Mon Aug 05, 2019 5:47 pm

TropicalDiver wrote:
Mon Aug 05, 2019 3:42 pm
5). Add an insulating wrap to your house
How would that work? I mean, insulating your house is a good idea, but what will it so for the rainout?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Water collecting in hose with higher humidity settings

Post by palerider » Mon Aug 05, 2019 5:48 pm

Pugsy wrote:
Mon Aug 05, 2019 4:03 pm
The typo police at work again. :lol: :lol:
C'mon, that one was funny!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11294
Joined: Fri Jun 19, 2015 4:36 pm

Re: Water collecting in hose with higher humidity settings

Post by zonker » Mon Aug 05, 2019 6:04 pm

palerider wrote:
Mon Aug 05, 2019 5:47 pm
TropicalDiver wrote:
Mon Aug 05, 2019 3:42 pm
5). Add an insulating wrap to your house
How would that work? I mean, insulating your house is a good idea, but what will it so for the rainout?
it would help keep the rain.....out.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg