Suplimental oxygen question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Woody
Posts: 591
Joined: Sat Mar 10, 2007 7:59 pm

Suplimental oxygen question

Post by Woody » Fri Aug 02, 2019 10:21 am

I need to use suplimental oxygen with my cpap. Acording to the
directions I am suposed to turn the cpap on before I turn the oxygen
concentrator on. I heard there is a valve I could buy to just turn
the oxygen off at the bleed in adapter but Cpap.com doesn't have
one any idea who might have such a thing.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: SleapyHead software

User avatar
zonker
Posts: 11393
Joined: Fri Jun 19, 2015 4:36 pm

Re: Suplimental oxygen question

Post by zonker » Fri Aug 02, 2019 11:09 pm

bump
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Suplimental oxygen question

Post by palerider » Fri Aug 02, 2019 11:21 pm

Woody wrote:
Fri Aug 02, 2019 10:21 am
but Cpap.com doesn't have
one any idea who might have such a thing.
Because cpap.com doesn't sell oxygen equipment, concentrators, tubing, etc. Try someone who does.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suplimental oxygen question

Post by Pugsy » Sat Aug 03, 2019 7:11 am

I think you need to be checking with whomever is supplying your oxygen and see if they know of an adapter that has an on/off valve.....I couldn't find one.
Admittedly I just did a quick 10 minute search but it's not something that is apparently common or maybe it is called something entirely different than what I was searching for.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

D.H.
Posts: 3533
Joined: Sun Sep 04, 2016 7:07 pm

Re: Suplimental oxygen question

Post by D.H. » Sat Aug 03, 2019 8:35 am

You can google an for "CPAP Oxygen Port."

Image

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Pugsy
Posts: 65190
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Suplimental oxygen question

Post by Pugsy » Sat Aug 03, 2019 8:47 am

D.H. wrote:
Sat Aug 03, 2019 8:35 am
You can google an for "CPAP Oxygen Port."
That has no ability to shut the O2 off or turn it back on.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Suplimental oxygen question

Post by Goofproof » Sat Aug 03, 2019 10:15 am

D.H. wrote:
Sat Aug 03, 2019 8:35 am
You can google an for "CPAP Oxygen Port."

Image
You might be able to, with help but, that won't make it a valve. I have and use a port in mine that has a ball check valve built into it that vents the O2 into the room when the XPAP isn't running, when the XPAP is on it closes off the vent port and lets the XPAP pressure in. I works like the one pictured in your post, except it's not Dumb.

I don't know where you can buy one , mine came from Lincare 12 years ago, the no longer supply one like it. No doubt the company was put out of business. when the CHEAP ones you pictured came on the market. The new ones cost less that $ .15 to make and sell for $ 5 to $ 20, I hope the markup is large enough. I'd hate to see them all out of business too. :mrgreen: Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
zonker
Posts: 11393
Joined: Fri Jun 19, 2015 4:36 pm

Re: Suplimental oxygen question

Post by zonker » Sat Aug 03, 2019 10:47 am

Pugsy wrote:
Sat Aug 03, 2019 8:47 am
D.H. wrote:
Sat Aug 03, 2019 8:35 am
You can google an for "CPAP Oxygen Port."
That has no ability to shut the O2 off or turn it back on.
never let facts interfere with a good post.
:roll:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ClayL
Posts: 369
Joined: Mon Feb 12, 2007 8:45 am
Location: Palisade, CO now, was Full time in a motor home for 12 years- Anywhere USA

Re: Suplimental oxygen question

Post by ClayL » Sat Aug 03, 2019 11:02 am

I wonder why they think it is necessary to turn the oxygen on after the CPAP? Neither my wife's sleep doctor or our DME said it was needed. For years I have turned her concentrator on when I go to bed ( a couple of hours before her) because she had a habit of forgetting to do it. The oxygen enters at the base of her hose and if the CPAP is off exits the mask. I don't see the problem with that. I think some oxygen exits the mask on exhale in normal operation anyway.
Respironics DreamStation
Pressure = 11.5 min 14.5 max
C-Flex = 1

Dreamwear FF

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Suplimental oxygen question

Post by Goofproof » Sat Aug 03, 2019 11:43 am

ClayL wrote:
Sat Aug 03, 2019 11:02 am
I wonder why they think it is necessary to turn the oxygen on after the CPAP? Neither my wife's sleep doctor or our DME said it was needed. For years I have turned her concentrator on when I go to bed ( a couple of hours before her) because she had a habit of forgetting to do it. The oxygen enters at the base of her hose and if the CPAP is off exits the mask. I don't see the problem with that. I think some oxygen exits the mask on exhale in normal operation anyway.
The problem is O2 is a oxidizer, it supports combustion, the bearings are lubed, lube + O2, are a fire hazzard, when the XPAP is running the O2 is forced away from the XPAP, when the XPAP is off the O2 can get into the XPAP! My thoughts, unlikely, but again I didn't put the "Warning Signs" on my bedroom windows "Oxygen in Use"! I'm not a Lawyer, I take chances based of odd's not $$$.

With a valve like mine it protects the machine, me too, if the machine shuts off and I don't awaken, the O2 still can come down the hose to the mask, little good but every bit helps! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Suplimental oxygen question

Post by palerider » Sat Aug 03, 2019 1:45 pm

D.H. wrote:
Sat Aug 03, 2019 8:35 am
You can google an for "CPAP Oxygen Port."

Image
Once again, answering questions without comprehending what the question WAS.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Suplimental oxygen question

Post by palerider » Sat Aug 03, 2019 2:07 pm

ClayL wrote:
Sat Aug 03, 2019 11:02 am
I wonder why they think it is necessary to turn the oxygen on after the CPAP? Neither my wife's sleep doctor or our DME said it was needed. For years I have turned her concentrator on when I go to bed ( a couple of hours before her) because she had a habit of forgetting to do it. The oxygen enters at the base of her hose and if the CPAP is off exits the mask. I don't see the problem with that. I think some oxygen exits the mask on exhale in normal operation anyway.
It also backs up and fills the cpap with pure oxygen, which, in the (hopefully unlikely) event that there's a spark, will burn vigorously.
https://www.youtube.com/results?search_ ... nvironment

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
bwexler
Posts: 1586
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: Suplimental oxygen question

Post by bwexler » Sat Aug 03, 2019 8:54 pm

I have only been using supplemental oxygen since 2010. Have only used a simple adapter like the one pictured until I got my Aircurve ASV with heated hose and oxygen port.
I have always turned on my O2 before I get into bed, usually just a few minutes never more than an hour before I mask up.
Maybe tonight will be the night I go up in a ball of flames.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
zonker
Posts: 11393
Joined: Fri Jun 19, 2015 4:36 pm

Re: Suplimental oxygen question

Post by zonker » Sat Aug 03, 2019 9:01 pm

bwexler wrote:
Sat Aug 03, 2019 8:54 pm
I have only been using supplemental oxygen since 2010. Have only used a simple adapter like the one pictured until I got my Aircurve ASV with heated hose and oxygen port.
I have always turned on my O2 before I get into bed, usually just a few minutes never more than an hour before I mask up.
Maybe tonight will be the night I go up in a ball of flames.
well, now....you've got to post something here every morning.

the day we don't hear from you....
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

USMCVet
Posts: 837
Joined: Sun Nov 12, 2017 6:46 pm
Contact:

Re: Suplimental oxygen question

Post by USMCVet » Sun Aug 04, 2019 11:48 am

What exactly are you trying to do? Just be able to turn on oxygen concentrator first without CPAP turned on?

First question is what happens when you turn it on with valve closed? Does the oxygen concentrator have a automatic pressure release valve or would you need something like this that they sell for fish tanks https://www.kensfish.com/cgi-kensfish/s ... fd224662db . You could have one valve open to the room when you turn on concentrator then close it when you open the valve going to CPAP.
That site has a ton of different types of valves if it helps you find what you need https://www.kensfish.com/cgi-kensfish/s ... e084af26f6

_________________
MachineMask
10 to 15 cm h20