Sorry - Another Leak Question - Where's it coming from?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Gator5000e
Posts: 134
Joined: Mon May 07, 2018 1:07 pm

Re: Sorry - Another Leak Question - Where's it coming from?

Post by Gator5000e » Thu Jul 18, 2019 7:43 am

Thanks for the help. I changed mask size and hose and still had a significant leak, according to the machine, of 99.6L/min. I had the mask strapped on pretty tight last night and didn't hear any major leaks. Wouldn't a leak of that size be fairly audible if it was the mask, tubing or one of the joints? Could it be the machine, humidifier tub or the machine to ResMed tube elbow ? I'm kind of grasping at straws here. Thanks for any thoughts. Again.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sorry - Another Leak Question - Where's it coming from?

Post by Pugsy » Thu Jul 18, 2019 7:45 am

Can I see the report from last night?

Do you have facial hair?

Still using the mask liner?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Gator5000e
Posts: 134
Joined: Mon May 07, 2018 1:07 pm

Re: Sorry - Another Leak Question - Where's it coming from?

Post by Gator5000e » Thu Jul 18, 2019 7:51 am

I didn't bring my SD card into work today. Sorry. I can post it tonight. No facial hair to speak of although I've been told I need to shave more often. LOL! Yes, I am using the liner. Those seem to help. But again, maybe not. I still have my original Dreamstation machine so maybe i will give that a shot tonight and see if it makes any difference. Just thought of doing that. But I will post a SH pic of last night's session.

BTW, should I now be using that other software, Oscar, instead of SH? I tried to see if I should switch. I also see that there is a version of SH 1.1. I'm still using v1.0.0-bets-2.

Thanks again.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sorry - Another Leak Question - Where's it coming from?

Post by Pugsy » Thu Jul 18, 2019 8:10 am

If your machine was acting up you wouldn't have times with no large leak showing on the leak graph.
Your other reports showed no large leak at times and then things went to hell. That tells me it is mask movement that is the issue and not the machine having a problem with the mask type per se. If it never showed leaks under control that's one thing which would maybe point to machine having a problem with your mask but that's not what I saw on your other reports.

So that makes me wonder what else might be causing mask movement and poor fit.
Hence the liner and facial hair question.
Liners can make leaks worse in terms of amounts...but they do buffer the sound so less likely to wake us up and that's why most people use liners. Leaks are minor but don't wake us up.

Facial hair....even if it's a short stiff stubble which could actually be more of a problem than a full soft beard...is the next thing to look at.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.