SmartStart Feature autostop not working

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
macmermaid
Posts: 36
Joined: Wed Dec 02, 2009 6:21 pm

SmartStart Feature autostop not working

Post by macmermaid » Thu Jul 04, 2019 11:30 am

Hi All,

So I have had my Airsense Autoset 10 For Her for almost a year now. The SmartStart feature is enabled and I've had no problem with it up until a couple of days ago. I checked the settings and it is enabled.- I actually turned it off and back on again. I have not changed my mask or tubing and there are no holes. I have Dreamwear with a nasal cushion. The pressure is set fro 4 to 20 and has not ever been changed.

The past couple of days, when I get up in the middle of the night to go to the bathroom and when I get up and take it off the mask in the morning, it keeps blowing really hard until I press stop. I look at the panel and it shows it is still running and the pressure level, etc until I press the stop button. It is definitely not just the light air that blows when you first take it off to dry the condensation in the hose.

Also, last night, maybe unrelated, I was awakened because the pressure had gone way up and I couldn't exhale. I took the mask off and put it back on and it was still blowing too hard. So, I turned the machine off and back on and then it seemed to go back to normal. All in the middle of the night!

I do not have a DME because I bought the machine from CPAP.com. Does anybody know if CPAP.com will help me with the warranty or do l need to call ResMed?
Does anybody know offhand how long the warranty lasts? I am a few days away from having the machine for a year.

Thanks!

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: SmartStart Feature autostop not working

Post by jimbud » Thu Jul 04, 2019 11:39 am

Turn off smartstart. Just turn the machine on and off manually.
I and a few others have had problems with it.
A few of the machine appear to have various glitches.
My personal glitch (with my machine) is that it would turn itself off(while I was asleep) on its own. Not fun.
I turned off smartstart and have had no problems since.
After all how much trouble is it to go back to the old days and turn it off and on manually? :)
Good luck, JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
LSAT
Posts: 13393
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: SmartStart Feature autostop not working

Post by LSAT » Thu Jul 04, 2019 12:48 pm

Many people have complained about the auto off feature. It is very sensitive. Just as Jimbud indicated...turn it off manually. It can't be too difficult unless your machine is in a difficult location. Warranty is 2 years...are you prepared to be without your machine for several weeks?

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: SmartStart Feature autostop not working

Post by Goofproof » Thu Jul 04, 2019 1:23 pm

jimbud wrote:
Thu Jul 04, 2019 11:39 am
Turn off smartstart. Just turn the machine on and off manually.
I and a few others have had problems with it.
A few of the machine appear to have various glitches.
My personal glitch (with my machine) is that it would turn itself off(while I was asleep) on its own. Not fun.
I turned off smartstart and have had no problems since.
After all how much trouble is it to go back to the old days and turn it off and on manually? :)
Good luck, JPB
100% As a Mechanic I'd have to train heavy equiptment operators to operate Mining Trucks, they would jump into the Cab, look at the controls and ask me how do you start this thing. I'd tell them that was the last question, what they needed to get down was How Do I Stop This Thing! Horse Power, 1200, Speed
M.P.H., Loads of 150 Ton, much harder to stop than start.

Odds.are your XPAP if fine use the buttons, you could let it set unplugged so it will reset.Cpap.Com gave you the manuals to set your machine (The Secret Handshake book) and in it is the warranty info, setting up info, I am guessing two years... I'd lre my fingers do the walking, use the buttons. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SmartStart Feature autostop not working

Post by palerider » Thu Jul 04, 2019 1:32 pm

macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: SmartStart Feature autostop not working

Post by zonker » Thu Jul 04, 2019 1:38 pm

palerider wrote:
Thu Jul 04, 2019 1:32 pm
macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:

to the OP: yes, there has been some talk about this on the forum here before. so some machines develop this "feature".

the only way to know if cpap.com stands behind the warranty is to contact them.

i'm guessing they will, but that's just a guess.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: SmartStart Feature autostop not working

Post by Wulfman... » Thu Jul 04, 2019 1:42 pm

macmermaid wrote:
Thu Jul 04, 2019 11:30 am
Hi All,

So I have had my Airsense Autoset 10 For Her for almost a year now. The SmartStart feature is enabled and I've had no problem with it up until a couple of days ago. I checked the settings and it is enabled.- I actually turned it off and back on again. I have not changed my mask or tubing and there are no holes. I have Dreamwear with a nasal cushion. The pressure is set fro 4 to 20 and has not ever been changed.

The past couple of days, when I get up in the middle of the night to go to the bathroom and when I get up and take it off the mask in the morning, it keeps blowing really hard until I press stop. I look at the panel and it shows it is still running and the pressure level, etc until I press the stop button. It is definitely not just the light air that blows when you first take it off to dry the condensation in the hose.

Also, last night, maybe unrelated, I was awakened because the pressure had gone way up and I couldn't exhale. I took the mask off and put it back on and it was still blowing too hard. So, I turned the machine off and back on and then it seemed to go back to normal. All in the middle of the night!

I do not have a DME because I bought the machine from CPAP.com. Does anybody know if CPAP.com will help me with the warranty or do l need to call ResMed?
Does anybody know offhand how long the warranty lasts? I am a few days away from having the machine for a year.

Thanks!
Warranties are generally for two years (unless stipulated otherwise). CPAP.COM is your first line for warranty service. They're GREAT.
I'm with the others.......just turn it off manually. It's probably not the machine, but mask or other factors like leaks.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SmartStart Feature autostop not working

Post by palerider » Thu Jul 04, 2019 3:17 pm

zonker wrote:
Thu Jul 04, 2019 1:38 pm
palerider wrote:
Thu Jul 04, 2019 1:32 pm
macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:
You'll have to put the machine outside the room, and tape... *extensively* to maintain a good seal in the room.
Report back.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: SmartStart Feature autostop not working

Post by Goofproof » Thu Jul 04, 2019 3:25 pm

palerider wrote:
Thu Jul 04, 2019 3:17 pm
zonker wrote:
Thu Jul 04, 2019 1:38 pm
palerider wrote:
Thu Jul 04, 2019 1:32 pm
macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:
You'll have to put the machine outside the room, and tape... *extensively* to maintain a good seal in the room.
Report back.
While a stupid solution, it still will not work even if you use a 60 inch barn fan. IXPAP works by keeping the pressure inside your airway at a higher CM than the room air surrounding your body.

Just get a iron lung, it's the answer to your dreams, or not! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SmartStart Feature autostop not working

Post by palerider » Thu Jul 04, 2019 4:11 pm

Goofproof wrote:
Thu Jul 04, 2019 3:25 pm
palerider wrote:
Thu Jul 04, 2019 3:17 pm
zonker wrote:
Thu Jul 04, 2019 1:38 pm
palerider wrote:
Thu Jul 04, 2019 1:32 pm
macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:
You'll have to put the machine outside the room, and tape... *extensively* to maintain a good seal in the room.
Report back.
While a stupid solution, it still will not work even if you use a 60 inch barn fan. IXPAP works by keeping the pressure inside your airway at a higher CM than the room air surrounding your body.

Just get a iron lung, it's the answer to your dreams, or not! Jim
Go back to sleep, you don't know what Zonker and I are talking about.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: SmartStart Feature autostop not working

Post by Goofproof » Thu Jul 04, 2019 4:33 pm

palerider wrote:
Thu Jul 04, 2019 4:11 pm
Goofproof wrote:
Thu Jul 04, 2019 3:25 pm
palerider wrote:
Thu Jul 04, 2019 3:17 pm
zonker wrote:
Thu Jul 04, 2019 1:38 pm
palerider wrote:
Thu Jul 04, 2019 1:32 pm


It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:
You'll have to put the machine outside the room, and tape... *extensively* to maintain a good seal in the room.
Report back.
While a stupid solution, it still will not work even if you use a 60 inch barn fan. XPAP works by keeping the pressure inside your airway at a higher CM than the room air surrounding your body.

Just get a iron lung, it's the answer to your dreams, or not! Jim
Go back to sleep, you don't know what Zonker and I are talking about.
You are correct I seldom know why or what you are talking about, but again I seldom care. Only Facts I Do! Jim


Ho Hum, 6 more hours until Bedtime.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
zonker
Posts: 11421
Joined: Fri Jun 19, 2015 4:36 pm

Re: SmartStart Feature autostop not working

Post by zonker » Thu Jul 04, 2019 5:06 pm

palerider wrote:
Thu Jul 04, 2019 3:17 pm
zonker wrote:
Thu Jul 04, 2019 1:38 pm
palerider wrote:
Thu Jul 04, 2019 1:32 pm
macmermaid wrote:
Thu Jul 04, 2019 11:30 am
I took the mask off and put it back on and it was still blowing too hard
It only "blows hard" when the mask is off, because the poor thing is trying to pressurize your entire room, instead of the small space between the mask and your face.

As soon as you put the mask on, it *immediately* goes back to normal operation.
i'm trying to get the ahi in my bedroom to a reasonable level.

if i post some charts, can you help?
:lol: :lol: :lol:
You'll have to put the machine outside the room, and tape... *extensively* to maintain a good seal in the room.
Report back.
<studiously measures room to see if he can get a big enough scunci>
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: SmartStart Feature autostop not working

Post by chunkyfrog » Thu Jul 04, 2019 5:21 pm

In an earlier thread, I posited that making auto-off harder to engage might be a safety feature.
Auto on: good, but only when it works.
Auto off: nice when it works--otherwise, no biggie.
(Unless you are missing an arm.)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

jsmit86
Posts: 265
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: SmartStart Feature autostop not working

Post by jsmit86 » Thu Jul 04, 2019 7:13 pm

I had a similar issue with my unit. (Same)
The AutoStop quit working. You'll see replies above that suggest just using the on/off button, but you also mentioned other issues.
I had the same problem. First autostop quit, but then I had some nights where I was struggling for air.
Bottom line? My unit has been out for repair for one month. It's on the way back to my DME from the warranty repair.
I hope to get an update of what they found. My advice, if auto stop is has suddenly stopped working, and you don't have any mask or tube problems, leaks, etc., you probably want to have it checked out. When my DME contacted Resmed, their folks did on online check, and said no problem found. Thankfully the DME had checked the unit and confirmed that it was NOT working properly. Resmed authorized the return, and I hope to get the unit back in the next week.

If you are having issues, insist on a proper repair.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
LSAT
Posts: 13393
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: SmartStart Feature autostop not working

Post by LSAT » Thu Jul 04, 2019 8:05 pm

jsmit86 wrote:
Thu Jul 04, 2019 7:13 pm
I had a similar issue with my unit. (Same)
The AutoStop quit working. You'll see replies above that suggest just using the on/off button, but you also mentioned other issues.
I had the same problem. First autostop quit, but then I had some nights where I was struggling for air.
Bottom line? My unit has been out for repair for one month. It's on the way back to my DME from the warranty repair.
I hope to get an update of what they found. My advice, if auto stop is has suddenly stopped working, and you don't have any mask or tube problems, leaks, etc., you probably want to have it checked out. When my DME contacted Resmed, their folks did on online check, and said no problem found. Thankfully the DME had checked the unit and confirmed that it was NOT working properly. Resmed authorized the return, and I hope to get the unit back in the next week.

If you are having issues, insist on a proper repair.
Unless there is a back up machine, having a machine out for a month could be a disaster for some people.