Resmed on full pressure always?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Delta_Dave
Posts: 6
Joined: Wed Jul 03, 2019 11:12 pm

Resmed on full pressure always?

Post by Delta_Dave » Wed Jul 03, 2019 11:33 pm

I'm about 2 weeks in as a new CPAP user and still adjusting. I have a Resmed airsense 10 autoset with full face mask. Last night I noticed pressure in my ears when I put on my mask to go to sleep. Not just ear popping, but constant pressure in my ears. I had pressure in my ears all night and woke up with a terrible sinus/ear headache. Tonight before bed I tried to adjust the pressures in the Clinician menu. It seems like air pressure is on full and the mask is constantly "pillowed" against my face. It wasn't like this before. When I started with this machine, it started at 4. I felt like I was suffocating. I adjusted the pressure to 6 and have been there for several prior nights. EPR is on and is on 3, the original setting. Now there does not seem to be any difference when I adjust from 4-20. It just feels like it's on full. It is set to Auto not CPAP. Is it broken? Is the Doctor's office able to change settings remotely and they changed it? I can't sleep with it on tonight with this high pressure and the pressure in my ears.

User avatar
Dog Slobber
Posts: 4325
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed on full pressure always?

Post by Dog Slobber » Thu Jul 04, 2019 6:21 am

You can't go by what it feels like, especially if you're new to CPAP.
  • Read the pinned topics,
  • Download OSCAR,
  • Post some charts,
You'll probably find that the pressure wasn't near full.
Battery Backup: EcoFlow Delta 2

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Resmed on full pressure always?

Post by Julie » Thu Jul 04, 2019 6:43 am

Do you actually NEED the high setting of EPR?

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Resmed on full pressure always?

Post by raisedfist » Thu Jul 04, 2019 7:28 am

Is the correct mask type chosen in the menu? Undo everything and plug it all back together. Make sure the AB filter is turned off as well if you're not using one.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed on full pressure always?

Post by LSAT » Thu Jul 04, 2019 9:14 am

Delta_Dave wrote:
Wed Jul 03, 2019 11:33 pm
I'm about 2 weeks in as a new CPAP user and still adjusting. I have a Resmed airsense 10 autoset with full face mask. Last night I noticed pressure in my ears when I put on my mask to go to sleep. Not just ear popping, but constant pressure in my ears. I had pressure in my ears all night and woke up with a terrible sinus/ear headache. Tonight before bed I tried to adjust the pressures in the Clinician menu. It seems like air pressure is on full and the mask is constantly "pillowed" against my face. It wasn't like this before. When I started with this machine, it started at 4. I felt like I was suffocating. I adjusted the pressure to 6 and have been there for several prior nights. EPR is on and is on 3, the original setting. Now there does not seem to be any difference when I adjust from 4-20. It just feels like it's on full. It is set to Auto not CPAP. Is it broken? Is the Doctor's office able to change settings remotely and they changed it? I can't sleep with it on tonight with this high pressure and the pressure in my ears.
The mask should be "pillowed"against your face. It fills with air and makes a seal. It should NOT be worn tight. Because you adjusted the pressure to 6 (Assume minimum) you have no idea what the average pressure was all night. 6 is just the starting pressure if you are set on auto,

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Resmed on full pressure always?

Post by jimbud » Thu Jul 04, 2019 9:41 am

Dog Slobber wrote:
Thu Jul 04, 2019 6:21 am
You can't go by what it feels like, especially if you're new to CPAP.
  • Read the pinned topics,
  • Download OSCAR,
  • Post some charts,
You'll probably find that the pressure wasn't near full.
See post above.
Quit panicking.
Post charts.
Get informed help. :)
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed on full pressure always?

Post by zonker » Thu Jul 04, 2019 11:03 am

jimbud wrote:
Thu Jul 04, 2019 9:41 am
Dog Slobber wrote:
Thu Jul 04, 2019 6:21 am
You can't go by what it feels like, especially if you're new to CPAP.
  • Read the pinned topics,
  • Download OSCAR,
  • Post some charts,
You'll probably find that the pressure wasn't near full.
See post above.
Quit panicking.
Post charts.
Get informed help. :)
JPB
:lol: :lol: why are you telling dogslobber this? :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: Resmed on full pressure always?

Post by jimbud » Thu Jul 04, 2019 11:11 am

zonker wrote:
Thu Jul 04, 2019 11:03 am
jimbud wrote:
Thu Jul 04, 2019 9:41 am
Dog Slobber wrote:
Thu Jul 04, 2019 6:21 am
You can't go by what it feels like, especially if you're new to CPAP.
  • Read the pinned topics,
  • Download OSCAR,
  • Post some charts,
You'll probably find that the pressure wasn't near full.
See post above.
Quit panicking.
Post charts.
Get informed help. :)
JPB
:lol: :lol: why are you telling dogslobber this? :lol: :lol:
Something good is always worth repeating.
Something good is always worth repeating.
Something good is always worth repeating.

Also I am plagiarizing Dogslobbers post and feel I should notify him. :roll:
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.
Last edited by jimbud on Thu Jul 04, 2019 12:09 pm, edited 1 time in total.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Resmed on full pressure always?

Post by Goofproof » Thu Jul 04, 2019 11:22 am

Julie wrote:
Thu Jul 04, 2019 6:43 am
Do you actually NEED the high setting of EPR?
It's a Comfort Setting, If you use it or not, the relief you like is totally up to you. It's a setting you can play with is you are so inclined, its not a live or die thing. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed on full pressure always?

Post by zonker » Thu Jul 04, 2019 12:20 pm

jimbud wrote:
Thu Jul 04, 2019 11:11 am

Something good is always worth repeating.
Something good is always worth repeating.
Something good is always worth repeating.

Also I am plagiarizing Dogslobbers post and feel I should notify him. :roll:
JPB
:lol: :lol: i agree;dogslobber is always worth repeating.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed on full pressure always?

Post by palerider » Thu Jul 04, 2019 1:36 pm

zonker wrote:
Thu Jul 04, 2019 12:20 pm
jimbud wrote:
Thu Jul 04, 2019 11:11 am

Something good is always worth repeating.
Something good is always worth repeating.
Something good is always worth repeating.

Also I am plagiarizing Dogslobbers post and feel I should notify him. :roll:
JPB
:lol: :lol: i agree;dogslobber is always worth repeating.
Apparently, we can all use a little more slobber in our lives, as long as it's doggie sourced.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed on full pressure always?

Post by zonker » Thu Jul 04, 2019 1:39 pm

palerider wrote:
Thu Jul 04, 2019 1:36 pm
zonker wrote:
Thu Jul 04, 2019 12:20 pm
jimbud wrote:
Thu Jul 04, 2019 11:11 am

Something good is always worth repeating.
Something good is always worth repeating.
Something good is always worth repeating.

Also I am plagiarizing Dogslobbers post and feel I should notify him. :roll:
JPB
:lol: :lol: i agree;dogslobber is always worth repeating.
Apparently, we can all use a little more slobber in our lives, as long as it's doggie sourced.
ew!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Muse-Inc
Posts: 4382
Joined: Fri Jun 12, 2009 8:44 pm
Location: Atlanta, GA

Re: Resmed on full pressure always?

Post by Muse-Inc » Fri Jul 05, 2019 5:28 am

Delta_Dave wrote:
Wed Jul 03, 2019 11:33 pm
...pressure in my ears all night and woke up with a terrible sinus/ear headache...
To help with the ear pressure, chew some gum or suck on mints. Obviously can't do that with the mask on, so do it when you take the mask off. Make sure the device is set to APAP and not CPAP.
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.