Do asv machines eliminate centrals?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zonker
Posts: 11495
Joined: Fri Jun 19, 2015 4:36 pm

Re: Do asv machines eliminate centrals?

Post by zonker » Sat Jul 06, 2019 6:02 pm

Chapalauser wrote: ↑
Sat Jul 06, 2019 5:31 pm

I have attached a sleepyhead screenshot
I am having a problem attaching a screenshot
okay. at the bottom of your text, there is a tab that says ATTACHMENT. click that.

next click the box that reads "Add files"

pick the screenshot you want to attach.

double click it.

you'll see the name of your file and next to it "place inline". click that.
screenshot-20190629-114643.png
viola! your screenshot should appear as mine has above.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Chapalauser
Posts: 33
Joined: Thu Jan 17, 2019 7:14 am
Location: Cda500 ft/Mex5000ft

Re: Do asv machines eliminate centrals?

Post by Chapalauser » Sat Jul 06, 2019 6:40 pm

Here is a recent screenshot. any comments would be appreciated
screenshot-20190706-174536.png
screenshot-20190706-174536.png (240.2 KiB) Viewed 2939 times

_________________
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: AIRCURVE 10 ASV min epap 5, min ps=4,max ps=10

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Do asv machines eliminate centrals?

Post by palerider » Sat Jul 06, 2019 6:56 pm

Chapalauser wrote: ↑
Sat Jul 06, 2019 6:40 pm
Here is a recent screenshot. any comments would be appreciatedscreenshot-20190706-174536.png
I would lower your PS, try bumping the minEPAP to 6 and drop PS to 2, and see what happens.