I posted a little while ago about my issues with the Dreamstation not seeming to react to my disturbances. I got very helpful advice from Palerider, zonker and others. I set my base pressure from 6 to 8 and dropped about 0.5AHI. But looking at my charts, the Dreamstation simply seems to be not reacting fast enough. Resmed Airsense 10 is universally agreed to have faster reaction time.
My insurance will not pay for a new machine since it has not yet been even three years. So I'll buy it myself. The local hospital and vendors are all 50-100% above what I see online. But then online you never know what you get. Any good recommendations?
I apologize should I infringe on board policy by asking this question, just thought I got some awesome advice before...
Online Shop recommendations?
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Online Shop recommendations?
This board is provided by cpap.com. you can toss 'em some of your money if you want.SleepySwiss wrote: ↑Tue Jul 02, 2019 12:25 amI posted a little while ago about my issues with the Dreamstation not seeming to react to my disturbances. I got very helpful advice from Palerider, zonker and others. I set my base pressure from 6 to 8 and dropped about 0.5AHI. But looking at my charts, the Dreamstation simply seems to be not reacting fast enough. Resmed Airsense 10 is universally agreed to have faster reaction time.
My insurance will not pay for a new machine since it has not yet been even three years. So I'll buy it myself. The local hospital and vendors are all 50-100% above what I see online. But then online you never know what you get. Any good recommendations?
I apologize should I infringe on board policy by asking this question, just thought I got some awesome advice before...
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Sheriff Buford
- Posts: 4111
- Joined: Mon Aug 09, 2010 8:01 am
- Location: Kingwood, Texas
Re: Online Shop recommendations?
The Resmed machines have a 9 second delay in their response. This response delay is for coughs, yawns, ect… So... I assume you are referring to the reaction time the machine takes once it decides it's an event occurring.
Sheriff
Sheriff
_________________
| Machine: AirSense 11 Autoset |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
Re: Online Shop recommendations?
I bought a used, low hour machine from LSAT here on the board. And I have bought several items from cpap.com. Id give either a high recommendation.SleepySwiss wrote: ↑Tue Jul 02, 2019 12:25 amI posted a little while ago about my issues with the Dreamstation not seeming to react to my disturbances. I got very helpful advice from Palerider, zonker and others. I set my base pressure from 6 to 8 and dropped about 0.5AHI. But looking at my charts, the Dreamstation simply seems to be not reacting fast enough. Resmed Airsense 10 is universally agreed to have faster reaction time.
My insurance will not pay for a new machine since it has not yet been even three years. So I'll buy it myself. The local hospital and vendors are all 50-100% above what I see online. But then online you never know what you get. Any good recommendations?
I apologize should I infringe on board policy by asking this question, just thought I got some awesome advice before...
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Online Shop recommendations?
Thanks, Sheriff. I think this is exactly what I am referring to. Below is a pic from last night. It seems to me that dreamstation ignores many events, but does react if there is a double. I understand (maybe mistakenly?) that resmed would catch these and change pressure more frequently.Sheriff Buford wrote: ↑Tue Jul 02, 2019 6:21 amThe Resmed machines have a 9 second delay in their response. This response delay is for coughs, yawns, ect… So... I assume you are referring to the reaction time the machine takes once it decides it's an event occurring.
Sheriff

_________________
| Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
- Attachments
-
- 2019-07-02_7-57-52.jpg (756.84 KiB) Viewed 2330 times
Re: Online Shop recommendations?
SleepySwiss wrote: ↑Tue Jul 02, 2019 12:25 amI posted a little while ago about my issues with the Dreamstation not seeming to react to my disturbances. I got very helpful advice from Palerider, zonker and others.
palerider said it, but i'll repeat it, as it helps the bots pick up on it-
https://www.cpap.com/
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
