Minimum Pressure vs pressure spike.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Fri Jun 28, 2019 10:09 am

I had a sleep trial done with an Dreamstation APAP. The machine was NOT responsive enough so my sleep clinician had set it on CPAP mode on 11 for the remainder of the trial

However now I have a S10 Autoset and have dialled in a APAP setting of 6-20 for the first night. Now I am getting pressure spikes for extremely brief periods before dropping below.

attached are a few reports from the Dreamstation trial and a report from sleepyhead from the S10 I have now

https://imgur.com/a/y1QJTqV

What should I set my minimum pressure to be?

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum Pressure vs pressure spike.

Post by palerider » Fri Jun 28, 2019 10:57 am

DreamInSleepNow wrote:
Fri Jun 28, 2019 10:09 am
I had a sleep trial done with an Dreamstation APAP. The machine was NOT responsive enough so my sleep clinician had set it on CPAP mode on 11 for the remainder of the trial

However now I have a S10 Autoset and have dialled in a APAP setting of 6-20 for the first night. Now I am getting pressure spikes for extremely brief periods before dropping below.

attached are a few reports from the Dreamstation trial and a report from sleepyhead from the S10 I have now

https://imgur.com/a/y1QJTqV

What should I set my minimum pressure to be?
First, it's not a S10, they dropped the "s" with the air10 line for some unknown reason...

Second, let's try a minimum pressure of 8, that'll help a lot.

Look at the pressure curve, see where it jumps up after the pressure drops? you want the min to be a bit above those spots where it jumps up... so as to prevent those events from happening in the first place.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Re: Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Fri Jun 28, 2019 7:47 pm

I bumped it up to 9 just after I made the post.

There was practically no OA however I did wake up in the middle of the night when the pressure jumped to 15


Here is the sleep report.
Image

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum Pressure vs pressure spike.

Post by palerider » Fri Jun 28, 2019 9:30 pm

DreamInSleepNow wrote:
Fri Jun 28, 2019 7:47 pm
I bumped it up to 9 just after I made the post.

There was practically no OA however I did wake up in the middle of the night when the pressure jumped to 15


Here is the sleep report.
Image
It looks better, except for that bad cluster. Let's let it ride for a bit.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Re: Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Fri Jun 28, 2019 11:34 pm

Sounds good to me. I was thinking if more of these spikes occur every few days or once a week or so I should cap the max pressure and ride through the episodes rather than disturb sleep??

I got a larger mask in the mail so hopefully that tightens up leak and comfort even further. Maybe I can tolerate the higher pressures then!

SDBud
Posts: 266
Joined: Thu Jan 03, 2019 3:24 pm
Location: Sunny SoCal

Re: Minimum Pressure vs pressure spike.

Post by SDBud » Sat Jun 29, 2019 1:18 am

DreamInSleepNow wrote:
Fri Jun 28, 2019 10:09 am
I had a sleep trial done with an Dreamstation APAP. The machine was NOT responsive enough so my sleep clinician had set it on CPAP mode on 11 for the remainder of the trial

However now I have a S10 Autoset and have dialled in a APAP setting of 6-20 for the first night. Now I am getting pressure spikes for extremely brief periods before dropping below.

attached are a few reports from the Dreamstation trial and a report from sleepyhead from the S10 I have now

https://imgur.com/a/y1QJTqV

What should I set my minimum pressure to be?
If you needed a pressure of 11 during your study, WHY would you think 6 would be OK at home??

I'd suggest at LEAST 10 for the minimum to start, and you will probably end up a bit higher than that.
Airsense 10 Autoset
AirTouch and AirFit F10 masks
:roll:

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum Pressure vs pressure spike.

Post by palerider » Sat Jun 29, 2019 1:58 am

DreamInSleepNow wrote:
Fri Jun 28, 2019 11:34 pm
Sounds good to me. I was thinking if more of these spikes occur every few days or once a week or so I should cap the max pressure and ride through the episodes rather than disturb sleep??
Most likely, it's the *episodes* that are disturbing your sleep, after all, that's pretty much the whole point of getting a cpap, because not breathing results in disturbed sleep.

Only, with the added stimulus of the cpap, when the events disturbed your sleep, you woke up enough to notice the mask and higher pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Re: Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Sat Jun 29, 2019 9:41 am

SDBud wrote:
Sat Jun 29, 2019 1:18 am

If you needed a pressure of 11 during your study, WHY would you think 6 would be OK at home??

I'd suggest at LEAST 10 for the minimum to start, and you will probably end up a bit higher than that.
Fom my past readings there has been reports of the Autoset being more "aggressive" or being able to treat sleep apneas at a "lower pressure". Me wanting to see it first hand was the reason why I had it wide from 6-20.

I'll see how 9 for now as I don't want to move it around too much without properly gathering data. At a minimum of 6 it seemed clear that it was too low.

palerider wrote:
Sat Jun 29, 2019 1:58 am
Most likely, it's the *episodes* that are disturbing your sleep, after all, that's pretty much the whole point of getting a cpap, because not breathing results in disturbed sleep.

Only, with the added stimulus of the cpap, when the events disturbed your sleep, you woke up enough to notice the mask and higher pressure.

I wouldn't think it would've been the one waking me up as I've always slept deeply! But I'll definitely keep that in mind.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Minimum Pressure vs pressure spike.

Post by palerider » Sat Jun 29, 2019 1:27 pm

DreamInSleepNow wrote:
Sat Jun 29, 2019 9:41 am
I wouldn't think it would've been the one waking me up as I've always slept deeply! But I'll definitely keep that in mind.
That is a fallacy that people tell themselves. Before cpap, you were passed out from exhaustion, spending the night with the equivalent of someone poking you with a sharp stick every couple minutes (what was your pre-treatment AHI?)

The fact that you don't *REMEMBER* waking up dozens, if not hundreds of times a night doesn't mean it wasn't happening.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

SDBud
Posts: 266
Joined: Thu Jan 03, 2019 3:24 pm
Location: Sunny SoCal

Re: Minimum Pressure vs pressure spike.

Post by SDBud » Sat Jun 29, 2019 5:33 pm

palerider wrote:
Sat Jun 29, 2019 1:27 pm
DreamInSleepNow wrote:
Sat Jun 29, 2019 9:41 am
I wouldn't think it would've been the one waking me up as I've always slept deeply! But I'll definitely keep that in mind.
That is a fallacy that people tell themselves. Before cpap, you were passed out from exhaustion, spending the night with the equivalent of someone poking you with a sharp stick every couple minutes (what was your pre-treatment AHI?)

The fact that you don't *REMEMBER* waking up dozens, if not hundreds of times a night doesn't mean it wasn't happening.
+1
Airsense 10 Autoset
AirTouch and AirFit F10 masks
:roll:

DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Re: Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Sun Jun 30, 2019 7:48 am

Pretreatment AHI was 15 on sides - 90 on my back!

That would explain why for me it seems the pressure would spike instead of gradual changes for others?

Last night was good! I'm going to give it some time then bumping it up to 9.6-10 cmh20

Image

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Minimum Pressure vs pressure spike.

Post by raisedfist » Sun Jun 30, 2019 11:14 am

I would leave things as they are for now and try to work on mask fit. Make sure the cushion is the right size, and usually for a FFM it is beneficial to be clean shaven to get a good seal.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

DreamInSleepNow
Posts: 9
Joined: Fri Jun 28, 2019 9:40 am

Re: Minimum Pressure vs pressure spike.

Post by DreamInSleepNow » Sun Jun 30, 2019 1:01 pm

The mask I have now is fitting fairly well but I got a larger one in the mail. I can’t grow much hopefully a handful of strands don’t effect it too much!