apnea on exhale

Post by tyrinryan » Wed Jun 26, 2019 2:34 pm

I read on the forum somewhere that there are more apneas on exhale than inhale. What are the stats? proof?

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: apnea on exhale

Post by JDS74 » Wed Jun 26, 2019 5:04 pm

Interesting thought.
My understanding of an apnea was a cessation of breathing in, OSA's are inhale, Hypopneas are inhale, and CSA's are no air movement in or out.
Is there an RT who can clarify if this is correct?
Edit: a couple of typos.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Last edited by JDS74 on Fri Jun 28, 2019 7:03 pm, edited 1 time in total.
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: apnea on exhale

Post by palerider » Wed Jun 26, 2019 5:17 pm

tyrinryan wrote:
Wed Jun 26, 2019 2:34 pm
I read on the forum somewhere that there are more apneas on exhale than inhale. What are the stats? proof?
And, where did you read that?

You can find all sorts of garbage, but was it refuted? Was it credible?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Patriot Partner
Posts: 44
Joined: Wed Jun 26, 2019 9:56 am

Re: apnea on exhale

Post by Patriot Partner » Thu Jun 27, 2019 5:15 am

My 2 cents

This is why a lot of cpap people should be on bipa
Even if they're showing compliant

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: apnea on exhale

Post by palerider » Thu Jun 27, 2019 10:39 am

Patriot Partner wrote:
Thu Jun 27, 2019 5:15 am
My 2 cents

This is why a lot of cpap people should be on bipa
Even if they're showing compliant
And why is that?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11398
Joined: Fri Jun 19, 2015 4:36 pm

Re: apnea on exhale

Post by zonker » Thu Jun 27, 2019 10:39 am

Patriot Partner wrote:
Thu Jun 27, 2019 5:15 am
My 2 cents

This is why a lot of cpap people should be on bipa
Even if they're showing compliant
please define "a lot".
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: apnea on exhale

Post by ChicagoGranny » Thu Jun 27, 2019 11:32 am

JDS74 wrote:
Wed Jun 26, 2019 5:04 pm
My understanding of an apnea was a cessation of breathing in
An apnea is a cessation of breathing, period.

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: apnea on exhale

Post by ChicagoGranny » Thu Jun 27, 2019 11:34 am

Patriot Partner wrote:
Thu Jun 27, 2019 5:15 am
My 2 cents

This is why a lot of cpap people should be on bipa
Even if they're showing compliant
If there were more apneas on exhale, then some should use a "reverse bilevel" - more pressure on exhale/less pressure on inhale.

User avatar
ChicagoGranny
Posts: 15455
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: apnea on exhale

Post by ChicagoGranny » Thu Jun 27, 2019 11:39 am

tyrinryan wrote:
Wed Jun 26, 2019 2:34 pm
I read on the forum somewhere that there are more apneas on exhale than inhale. What are the stats? proof?
JDS74 wrote:
Wed Jun 26, 2019 5:04 pm
Interesting thought.
My understanding of an apnea was a cessation of breathing in, OSA's are inhale, Hypopneas are inhale, and CSA's are not air movement in or out.
Is there and RT who can clarify if this is correct?
Are you guys thinking that an obstructive apnea just begins "boom" at a single point in time? All the trace flows of apneas show a somewhat gradual decline of stability in the airway leading to a complete, or nearly complete, cessation of breathing.

Janknitz
Posts: 8514
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: apnea on exhale

Post by Janknitz » Thu Jun 27, 2019 2:32 pm

Exhalation is a passive relaxation of the diaphragm muscle, and generally has sufficient pressure to push through any obstruction in the airway. That's why exhalation pressure relief works--you don't have to hold the airway open much if at all for exhalation in most cases.

I've seen a few people here say that their apneas are on exhalation and they perceive them even while awake. I don't really understand this enough to comment. However there is an interesting article about expiratory apnea here: https://doctorstevenpark.com/expiratory ... ral-apneas

Inspiration requires respiratory effort--the flattening of the diaphragm to pull air in, and may not be strong enough to push through an obstruction because the airway is more relaxed during sleep.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

tyrinryan
Posts: 133
Joined: Wed Jun 26, 2019 1:46 pm

Re: apnea on exhale

Post by tyrinryan » Tue Jul 23, 2019 3:21 pm

I thought that this was a known generalization (all apneas are on exhale) and I was sure I read it straight-up on this forum or somewhere else. It seemed a little counterintuitive. I was looking for confirmation and proof from the Forum members. However, without prompting, my DME sleep tech said it to me too, just the other day-"all apneas are on exhale". I am a newbie and I of course nodded with agreement! When I looked at an Oscar chart for one night, I could not see any clear or dramatic evidence of this concept. Could the question could be better phrased: do all apneas start with an airway closure (or even almost closure) during expiration? If this were so, it would seem to me to important when choosing a CPAP machine with one of the EPR/C Flex algorithms. I read somewhere, that without positive pressure, the "regular" pressure is lowest at the end-expiration point. Would that not mean that EPR should stop and regular PAP pressure should begin at the end of expiration instead of the start of inspiration. I read somewhere that that was the difference in the algorithms of Phillips and Resmed. Would this then be a reason to choose Phillips over Resmed? It is hard to believe that Resmed overlooked this. I look forward to your comments.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: apnea on exhale

Post by palerider » Tue Jul 23, 2019 10:24 pm

tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
I thought that this was a known generalization (all apneas are on exhale)
You're seem to be quite confused about this.
Apneas don't happen *on exhale* typically, the throat closes *after exhale* and *before inhale*... preventing the inhalation from happening. Now, there is a thing called 'palatial prolapse' where your soft palate can get in the way of exhaling, but it's very unusual.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
and I was sure I read it straight-up on this forum or somewhere else. It seemed a little counterintuitive. I was looking for confirmation and proof from the Forum members. However, without prompting, my DME sleep tech said it to me too, just the other day-"all apneas are on exhale".
Either they don't know what they're talking about, or they just poorly phrased things.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
Could the question could be better phrased: do all apneas start with an airway closure (or even almost closure) during expiration?
No. If they did, you'd see a sudden cessation of exhalation flow, as opposed to the normal tapering off you see in almost every instance.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
If this were so, it would seem to me to important when choosing a CPAP machine with one of the EPR/C Flex algorithms.
I'm not aware of any modern machines that don't have some form of exhale relief, so that's a non-sequitur.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
I read somewhere, that without positive pressure, the "regular" pressure is lowest at the end-expiration point.
I have no idea what you're reading, but CPAPs *always* have positive pressure, that's what the CONTINUOUS means.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
Would that not mean that EPR should stop and regular PAP pressure should begin at the end of expiration instead of the start of inspiration.
Then it would be like *flex, which does exactly that, and confuses many people into thinking that the machine is trying to rush their breathing.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
I read somewhere that that was the difference in the algorithms of Phillips and Resmed.
Yes.
pressure-relief.png
It seems a majority of people prefer the way Resmed does it.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
Would this then be a reason to choose Phillips over Resmed?
Quite the opposite.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
It is hard to believe that Resmed overlooked this.
"overlooked" that they have a superior system? pretty sure they didn't.
tyrinryan wrote:
Tue Jul 23, 2019 3:21 pm
I look forward to your comments.
Well, you have 'em.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

tyrinryan
Posts: 133
Joined: Wed Jun 26, 2019 1:46 pm

Re: apnea on exhale

Post by tyrinryan » Tue Jul 23, 2019 11:55 pm

Thanks pale rider That chart is terrific

tyrinryan
Posts: 133
Joined: Wed Jun 26, 2019 1:46 pm

Re: apnea on exhale

Post by tyrinryan » Fri Jul 26, 2019 11:36 am

What does the grey line shown in the graphs shown by Palerider represent?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: apnea on exhale

Post by palerider » Fri Jul 26, 2019 12:25 pm

tyrinryan wrote:
Fri Jul 26, 2019 11:36 am
What does the grey line shown in the graphs shown by Palerider represent?
Respiratory flow.