Resemd Airsense 10 Autoset fully reset itself.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
exhale
Posts: 5
Joined: Wed Jun 26, 2019 8:56 am

Resemd Airsense 10 Autoset fully reset itself.

Post by exhale » Wed Jun 26, 2019 9:18 am

Hello everyone,

I've been on CPAP therapy for close to 2 months now and i've been using the Airsense 10 Autoset for this whole period of time. Today i went to sleep in another room, since the AC unit is in my bedroom and i wanted to avoid sleeping in the same room with it, while carrying the machine with me. I've left the humidifier in the bedroom, used the cover, set it up, put on the mask and went to sleep. After waking up, i unplugged the mask for the usual cleaning and brought the machine back in the bedroom. When i plugged it in, that's when i noticed that every bit of therapy info it had, it was reset to 0/null/--. No info about today's sleep session or any of the past ones, no info about usage hours (be it in the Patient or Clinical menu)... it's like the machine has been master factory reset (even the usage hours from the clinical menu, the one which i shouldn't be able to reset, have been reset). Has this happened to anyone else? What concerns me is that during this sleep session, the device recorded a very high amount of Central events (~3) which i rarely see jumping over 0.5. When i tried to import this session data, alongside other 5-6 days since the last import, into Sleepyhead, it refused to import anything. I had to download OSCAR and finally got it to work.

User avatar
Dog Slobber
Posts: 4325
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by Dog Slobber » Wed Jun 26, 2019 10:37 am

Hopefully you have an SD card in it.

Load the SD Card into S.H./OSCAR and see if the information is there.

The 0's could just be on the last session, change the the reporting period to a week or month and see if they are 0 as well.
Battery Backup: EcoFlow Delta 2

User avatar
LSAT
Posts: 13400
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by LSAT » Wed Jun 26, 2019 10:50 am

Usage hours CAN be reset..RUN hours cannot. Have you looked at the Sleep report on your screen? What time was it when you looked for the data...if it was after Noon, the machine may have reset to the next day with all zeros.

exhale
Posts: 5
Joined: Wed Jun 26, 2019 8:56 am

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by exhale » Wed Jun 26, 2019 11:34 am

Dog Slobber wrote:
Wed Jun 26, 2019 10:37 am
Hopefully you have an SD card in it.

Load the SD Card into S.H./OSCAR and see if the information is there.

The 0's could just be on the last session, change the the reporting period to a week or month and see if they are 0 as well.
Every session i had is on the SD card... so fortunately, the card has everything on it, including this last session (which i could only import into OSCAR but not SH). The device itself shows 0 (zero) on everything, including usage hours (which are not meant to be reset, from what i've read so far). I took some pictures but i can't seem to be able to upload them (probably because i'm a new user with a new account).
LSAT wrote:
Wed Jun 26, 2019 10:50 am
Usage hours CAN be reset..RUN hours cannot. Have you looked at the Sleep report on your screen? What time was it when you looked for the data...if it was after Noon, the machine may have reset to the next day with all zeros.
That's why i'm so stunned by what's happening... there is no DME or physician access, i own the device and bought it as new. But even so, the RUN hours are 0 (zero). All the settings have been preserved (mask type, min/max pressure, climate/tube temp, etc.).

User avatar
zonker
Posts: 11438
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by zonker » Wed Jun 26, 2019 11:37 am

exhale wrote:
Wed Jun 26, 2019 11:34 am
Dog Slobber wrote:
Wed Jun 26, 2019 10:37 am
Hopefully you have an SD card in it.

Load the SD Card into S.H./OSCAR and see if the information is there.

The 0's could just be on the last session, change the the reporting period to a week or month and see if they are 0 as well.
Every session i had is on the SD card... so fortunately, the card has everything on it, including this last session (which i could only import into OSCAR but not SH). The device itself shows 0 (zero) on everything, including usage hours (which are not meant to be reset, from what i've read so far). I took some pictures but i can't seem to be able to upload them (probably because i'm a new user with a new account).

i am an idiot in such things. it could be that you are too new to upload.

having said that, do you see the attachments tab below? have you followed that link?

if yes, your pics may need to be downsized in order to post.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

exhale
Posts: 5
Joined: Wed Jun 26, 2019 8:56 am

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by exhale » Wed Jun 26, 2019 11:47 am

zonker wrote:
Wed Jun 26, 2019 11:37 am
exhale wrote:
Wed Jun 26, 2019 11:34 am
Dog Slobber wrote:
Wed Jun 26, 2019 10:37 am
Hopefully you have an SD card in it.

Load the SD Card into S.H./OSCAR and see if the information is there.

The 0's could just be on the last session, change the the reporting period to a week or month and see if they are 0 as well.
Every session i had is on the SD card... so fortunately, the card has everything on it, including this last session (which i could only import into OSCAR but not SH). The device itself shows 0 (zero) on everything, including usage hours (which are not meant to be reset, from what i've read so far). I took some pictures but i can't seem to be able to upload them (probably because i'm a new user with a new account).

i am an idiot in such things. it could be that you are too new to upload.

having said that, do you see the attachments tab below? have you followed that link?

if yes, your pics may need to be downsized in order to post.

good luck!
It seems the size of the pictures was too big, downsizing them did the trick! So here's what my machine displays.. i have close to 2 months of therapy, 400 hours of usage and never, ever even entered the "Reset data" menu... as i said, i just woke up, moved the machine back and this is what it started displaying. Should i be concerned? Should i contact my DME, ask him to look at this?
Attachments
Poza3.jpg
Poza3.jpg (229.76 KiB) Viewed 2421 times
Poza2.jpg
Poza2.jpg (176.64 KiB) Viewed 2421 times
Poza1.jpg
Poza1.jpg (228.3 KiB) Viewed 2421 times

User avatar
LSAT
Posts: 13400
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by LSAT » Wed Jun 26, 2019 11:57 am

Go back to the DME...bet they will have no answer for you....to me it looks like a brand new machine. :?:

exhale
Posts: 5
Joined: Wed Jun 26, 2019 8:56 am

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by exhale » Wed Jun 26, 2019 12:10 pm

LSAT wrote:
Wed Jun 26, 2019 11:57 am
Go back to the DME...bet they will have no answer for you....to me it looks like a brand new machine. :?:
I was afraid this would be my only choice :(. It isn't new and i can prove that (if that's needed) with screenshots from OSCAR / SH (which contain the device S/N) and the one from the device's display. Thank you!

jsmit86
Posts: 265
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by jsmit86 » Wed Jun 26, 2019 6:30 pm

I have a question... I have a Resmed Airsense 10 Autoset, that is currently out for repair.
In looking at the PIX above, I am seeing detailed sleep data with LEAK, AHI. Central etc....
I have never seen that display on my unit, and as far as I know I had detailed or extended info set in the menu.

Can anyone reply with how to get to that screen, including any necessary options in the menu as a pre-requisite?
I'd like to see of I can get that once my machine returns.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
Dog Slobber
Posts: 4325
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by Dog Slobber » Wed Jun 26, 2019 8:25 pm

exhale wrote:
Wed Jun 26, 2019 11:47 am

It seems the size of the pictures was too big, downsizing them did the trick! So here's what my machine displays.. i have close to 2 months of therapy, 400 hours of usage and never, ever even entered the "Reset data" menu... as i said, i just woke up, moved the machine back and this is what it started displaying. Should i be concerned? Should i contact my DME, ask him to look at this?
Wow.

Check to make sure your therapy settings are as expected.


Tomorrow morning, make a not of your results, download your data. Wait an hour or so if you use MyAir or need data sent for compliancy purposes. The deliberately unplug, leave it unplugged for a few minutes, then plug it back in again and see if it's reset itself again.

If it has, then definitely have it replaced.

If it hasn't, keep your eye on it, then in a few years, if it does it again sell it as a new machine. :)
Battery Backup: EcoFlow Delta 2

User avatar
LSAT
Posts: 13400
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by LSAT » Wed Jun 26, 2019 8:33 pm

jsmit86 wrote:
Wed Jun 26, 2019 6:30 pm
I have a question... I have a Resmed Airsense 10 Autoset, that is currently out for repair.
In looking at the PIX above, I am seeing detailed sleep data with LEAK, AHI. Central etc....
I have never seen that display on my unit, and as far as I know I had detailed or extended info set in the menu.

Can anyone reply with how to get to that screen, including any necessary options in the menu as a pre-requisite?
I'd like to see of I can get that once my machine returns.
Have you ever looked at your Sleep Report Screen? It's right below OPTIONS on your opening screen. If some things are missing, go into your SETTINGS menu...find ESSENTIALS and change the setting to PLUS

exhale
Posts: 5
Joined: Wed Jun 26, 2019 8:56 am

Re: Resemd Airsense 10 Autoset fully reset itself.

Post by exhale » Thu Jun 27, 2019 12:43 am

Dog Slobber wrote:
Wed Jun 26, 2019 8:25 pm
exhale wrote:
Wed Jun 26, 2019 11:47 am

It seems the size of the pictures was too big, downsizing them did the trick! So here's what my machine displays.. i have close to 2 months of therapy, 400 hours of usage and never, ever even entered the "Reset data" menu... as i said, i just woke up, moved the machine back and this is what it started displaying. Should i be concerned? Should i contact my DME, ask him to look at this?
Wow.

Check to make sure your therapy settings are as expected.


Tomorrow morning, make a not of your results, download your data. Wait an hour or so if you use MyAir or need data sent for compliancy purposes. The deliberately unplug, leave it unplugged for a few minutes, then plug it back in again and see if it's reset itself again.

If it has, then definitely have it replaced.

If it hasn't, keep your eye on it, then in a few years, if it does it again sell it as a new machine. :)
If only these were sought after, back here.. Yeah, i think the machine had a brain fart, after tonight's sleep session (of 6 hours), it showed, in the morning, 9h30min of usage with a 0.1 AHI... but before, after yesterday's afternoon sleep, it showed 2+ AHI (mainly Clear Airways)... just reset it and gonna give it another chance during this weekend, maybe it decides it loves me after all. If not, back to the DME it goes!
Attachments