I got a second machine for another place where I sleep . I normally used a AutoPAP machine set to approximately 7-13 cmh2o because my prescribed CPAP pressure is 12.
I though this time I want to try BIPAP and got a Respironics Dreamstation AutoBIPAP. This machine apparently only three modes : Fixed CPAP, Fixed BiPAP and Auto BIPAP.
Can a BIPAP be configured similarly to AutoPAP or do I need a new sleep study to get the right numbers?
New BiPAP machine.
- Dog Slobber
- Posts: 4334
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: New BiPAP machine.
Pick a machine, I'd start with the APAP.stevej1234 wrote: ↑Tue Jun 25, 2019 10:21 amI got a second machine for another place where I sleep . I normally used a AutoPAP machine set to approximately 7-13 cmh2o because my prescribed CPAP pressure is 12.
I though this time I want to try BIPAP and got a Respironics Dreamstation AutoBIPAP. This machine apparently only three modes : Fixed CPAP, Fixed BiPAP and Auto BIPAP.
Can a BIPAP be configured similarly to AutoPAP or do I need a new sleep study to get the right numbers?
Post some charts, have the experts optimize your settings. Then when you're dialed-in with optimal settings. Switch over to the BiLevel, and then repeat.
Then whatever machine you're getting better results with, keep at the place you sleep more often at.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- ChicagoGranny
- Posts: 15523
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New BiPAP machine.
It can be figured identically. Select APAP mode and choose the same settings you currently use. Do you know how to get into the clinical menu to change the pressure settings?
Which model machine are you currently using?
Re: New BiPAP machine.
Auto Bipap will give you functionality of apap with a range of pressures if you choose PS minimum and max of 0.0.
So if you wanted to mimic current apap settings...(you didn't mention any exhale relief or use of Flex)
Auto Bilevel mode
Minimum EPAP 7
PS minimum 0 and PS maximum of 0
Maximum IPAP 13
It will function like your APAP. You can add in BiFlex exhale relief if you wish.
So if you wanted to mimic current apap settings...(you didn't mention any exhale relief or use of Flex)
Auto Bilevel mode
Minimum EPAP 7
PS minimum 0 and PS maximum of 0
Maximum IPAP 13
It will function like your APAP. You can add in BiFlex exhale relief if you wish.