Airsense 10 Red Frowny Face

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
el344
Posts: 27
Joined: Fri Dec 29, 2017 1:59 pm
Location: Tennessee

Airsense 10 Red Frowny Face

Post by el344 » Tue Jun 18, 2019 9:54 am

I have both an Airsense 10 Elite and an Airsense 10 Autoset, both set to a fixed pressure of 9. I have been using the Elite for most of the past 6 months. I have been using a Dreamwear nasal pillow mask for over a year. I have almost never gotten a red frowny face for mask fit until the last two weeks. Now I get it every night. My AHI is still around 0.5. When I run the mask fit option it tells me it's good.

I've changed back to my Swift FX. No change. I've replaced my pillows, headgear, and frame for both masks. No change. Last night I changed from my Elite to my Autoset machine and surprise -- no more red frowny face.

So it seems there is something wrong with my Elite. Does anyone have any suggestions as to how to fix this?

Thanks!

_________________
Machine
Additional Comments: Airfit N30 Mask
me:
ResMed AirSense10 Elite
AirFit N30 Mask

him:
ResMed AirSense 10 Autoset
AirFit F20 Mask

rcobourn
Posts: 207
Joined: Mon Apr 03, 2017 8:25 am

Re: Airsense 10 Red Frowny Face

Post by rcobourn » Tue Jun 18, 2019 10:22 am

Sounds like you still need to eliminate a problem in the hose or humidifier, unless you are using the same ones for both machines.

User avatar
el344
Posts: 27
Joined: Fri Dec 29, 2017 1:59 pm
Location: Tennessee

Re: Airsense 10 Red Frowny Face

Post by el344 » Tue Jun 18, 2019 10:50 am

I am using the same hose and humidifier. All settings are the same.

_________________
Machine
Additional Comments: Airfit N30 Mask
me:
ResMed AirSense10 Elite
AirFit N30 Mask

him:
ResMed AirSense 10 Autoset
AirFit F20 Mask

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 10 Red Frowny Face

Post by Pugsy » Tue Jun 18, 2019 11:26 am

Have you done an awake experiment with the potentially ailing Elite to see if your leak is ever within acceptable limits or zero when you know for a fact it isn't leaking?
That's what I would do first...awake...make sure no leaks...use the machine for about 30 minutes while awake...then look at the leak line graphs and see if they are showing consistent large leak reporting or intermittent variations in the leak being reported.
If you are awake and you know you aren't leaking anyway...there should be no excess leaks at all and most certainly no variance in the leak line.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.