23 days on cpap and still nothing?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
UkuleleDez
Posts: 7
Joined: Wed Jun 12, 2019 8:50 pm
Location: Australia, Perth

23 days on cpap and still nothing?

Post by UkuleleDez » Thu Jun 13, 2019 11:50 pm

I've read so many stories of people feeling better after just one night or one week on cpap, but after 23 days, I feel no different. I'm still as tired and sleepy as I have always been.
My sleep study at the clinic showed I had 33 events an hour and stopped breathing for about 30-40 seconds and longest time was 71 seconds.
Im doing the 4 week trial cpap which cost $300
Here's my readings so far:
23 days
7.4 average sleep
10.8 pressure
11L a min average leak
AHI: 0.5
Total AI: 0.3
Central AI: 0.2
Face mask seal good

I use the Air Touch F20 mask with the memory foam, instructions say air passes/leaks through the foam to create a more comfortable experience on the skin. I had tried 2 masks previous, but the rubber irritates my skin, this Air Touch is very comfortable. The mask fits good and I dont feel any air leaks, its a good seal.

Are these readings acceptable including the 11L a min air leak?

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 23 days on cpap and still nothing?

Post by Pugsy » Fri Jun 14, 2019 6:56 am

Welcome to the forum.

Where are you getting the numbers you reported? Off the machine LCD screen???
If so that 11 L/min is not an average leak number. It is what is called a 95 % number and all that means is you were AT OR BELOW that number for 95 % of the night. The or below part is very important and often people don't understand it or they forget that part.
24 L/min is the red line in the sand and even then it's not set in concrete. It's not like things go totally in the toilet if you have 27 L/min.

Leaks are not impacting therapy quality. Now they might still be impacting sleep quality if they wake you up.

Do leaks ever wake you up?

Do you take any medications of any kind? If so, what?

Do you wake up often during the night for any reason?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: 23 days on cpap and still nothing?

Post by raisedfist » Fri Jun 14, 2019 7:27 am

Well, if you're doing a 30 day trial of an APAP, and it's set at default settings in order for them to gather data and determine an appropriate pressure or pressure range for the future, you should not expect to feel any better than before until you get the pressure range honed in - the minimum pressure being the most important one.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: 23 days on cpap and still nothing?

Post by Julie » Fri Jun 14, 2019 8:08 am

It looks like a plain Cpap, not Apap...

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: 23 days on cpap and still nothing?

Post by NoOnesPerfect » Fri Jun 14, 2019 8:23 am

Julie wrote:
Fri Jun 14, 2019 8:08 am
It looks like a plain Cpap, not Apap...
While that IS what is shown in the OPs profile, a lot of people don’t realize the Autosense version is shown further down on the drop down menu, under “R” for ResMed, rather than “A” for autosense, so it is possible this is about an APAP machine.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 23 days on cpap and still nothing?

Post by Pugsy » Fri Jun 14, 2019 8:27 am

Need to see the software detailed reports to really know what machine and what settings and what all those numbers really point to.

That 10.8 pressure might be a fixed cpap pressure for all we know. Unlikely of course but not impossible.
Someone might have used their crystal ball and pulled 10.8 out to use.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

UkuleleDez
Posts: 7
Joined: Wed Jun 12, 2019 8:50 pm
Location: Australia, Perth

Re: 23 days on cpap and still nothing?

Post by UkuleleDez » Fri Jun 14, 2019 8:28 am

Yes you are right, the ResMed one was further down the list, I have changed it

UkuleleDez
Posts: 7
Joined: Wed Jun 12, 2019 8:50 pm
Location: Australia, Perth

Re: 23 days on cpap and still nothing?

Post by UkuleleDez » Fri Jun 14, 2019 8:29 am

OK, would that info be on the SD card? If so I should look at it?

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: 23 days on cpap and still nothing?

Post by NoOnesPerfect » Fri Jun 14, 2019 8:33 am

UkuleleDez wrote:
Fri Jun 14, 2019 8:29 am
OK, would that info be on the SD card? If so I should look at it?
Detailed information on both the machine settings and the daily graphs of therapy “results” are on the SD card. You absolutely should download the OSCAR software and use it to look at the data. That will also make it easy for you to post screenshots of the pertinent info for people here to review and comment on.

User avatar
Pugsy
Posts: 65137
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 23 days on cpap and still nothing?

Post by Pugsy » Fri Jun 14, 2019 8:37 am

See my sticky at the top of the Announcement section of the forum....Newbies please read
in it you will find information on how to get and use the software to see the detailed data your machine gathers.

Answers to my other questions would also maybe shed some light on why you aren't feeling the good numbers yet....

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

UkuleleDez
Posts: 7
Joined: Wed Jun 12, 2019 8:50 pm
Location: Australia, Perth

Re: 23 days on cpap and still nothing?

Post by UkuleleDez » Fri Jun 14, 2019 8:42 am

ok found the software thanks :D

UkuleleDez
Posts: 7
Joined: Wed Jun 12, 2019 8:50 pm
Location: Australia, Perth

Re: 23 days on cpap and still nothing?

Post by UkuleleDez » Fri Jun 14, 2019 9:03 am

Many thanks for your help.

I have attached my read out, if someone can explain it, that would be great :D
Attachments
Readout.jpg
Readout.jpg (529.91 KiB) Viewed 1571 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: 23 days on cpap and still nothing?

Post by Julie » Fri Jun 14, 2019 10:48 am

Need to see the graphs as per Pugsy's sticky.

User avatar
zonker
Posts: 11358
Joined: Fri Jun 19, 2015 4:36 pm

Re: 23 days on cpap and still nothing?

Post by zonker » Fri Jun 14, 2019 10:54 am

UkuleleDez wrote:
Thu Jun 13, 2019 11:50 pm
I've read so many stories of people feeling better after just one night or one week on cpap, but after 23 days, I feel no different. I'm still as tired and sleepy as I have always been.
ace mask seal good
welcome to the zoo! you are already in the capable hands of other forum members. i've nothing to add at this point EXCEPT-

yes, there are a tiny minority of folks who can turn on their cpap machines and get relief almost at once! but we don't see them here. and why should we? we can't help them! :lol:

but for the lot of us over here in this forum, it's taken a bit of time to see good results. now that you've found us, you are going to get all kinds of good suggestions.

it's going to take time, even with your therapy optimized.

but stick with it. you can do this.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: 23 days on cpap and still nothing?

Post by Okie bipap » Fri Jun 14, 2019 12:04 pm

The numbers actually look good. You are using the machine over 7 hours every night and your AHI is averaging around 0.6 which is very good. Your leak rates look good. You have had a small amount of high leakage, but if it is not disturbing your sleep, I wouldn't worry about it.