Something I Hate

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
i_am_jim
Posts: 38
Joined: Wed Oct 26, 2005 2:14 pm

Something I Hate

Post by i_am_jim » Mon Jun 10, 2019 3:00 pm

I admit I have little patience, but . . .

One of the things that simply infuriate me is when I explain a problem I am having with a device they made or sold me and they tell me it's not doing what I just explained it was doing. I use it -- they don't, "please don't tell me not to believe my eyes," or in this case what's happening around me.

I got a new CPAP machine (resmed airsense 10) about 3 weeks ago. Previously I had an Autoset S9 so I know how they should work and the S9 worked close to perfect in this regard. I use the automatic start and stop mode. Since I began using this one it has only started once by me sucking on it -- I have to press the start button and then suck on it for it to start. About 80% of the time it doesn't turn off when I take it off. I have to pull the hose out of the unit to make it stop. One day I forgot to do that and it blew all day. I only noticed it when I was about to put it back on. The S9 always started and almost always stopped properly.

My supplier says they looked at it on the cellular network and it's working perfectly.

NO! It's not.

<sometimes I just have to vent>
ResMed airsense10 w/integrated Humidifier
ResMed Mirage Swift Lt Pillow System
Image

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: Something I Hate

Post by NoOnesPerfect » Mon Jun 10, 2019 3:13 pm

I have found that I need to exhale strongly to get the SmartStart to work. The stop feature did not seem to be working for me, but I have never waited more than a second or two before hitting the off button.

To your main point though, I agree completely. Is your DME close enough that it would be easy to go to their office and demo the problem?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Something I Hate

Post by chunkyfrog » Mon Jun 10, 2019 3:59 pm

I have a similar machine.
I use the smart start to turn it on, but choose to shut it off manually,
rather than wait however long it finally takes to shut off. (Not patient, I am.)
No biggie; but if I had a disability, this would be a huge issue.
Resmed has been told--but maybe they cannot hear . . .
This problem is quite common.---maybe on the next model?

User avatar
Dog Slobber
Posts: 4340
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Something I Hate

Post by Dog Slobber » Mon Jun 10, 2019 4:41 pm

chunkyfrog wrote: ↑
Mon Jun 10, 2019 3:59 pm
I have a similar machine.
I use the smart start to turn it on, but choose to shut it off manually,
rather than wait however long it finally takes to shut off. (Not patient, I am.)
No biggie; but if I had a disability, this would be a huge issue.
Resmed has been told--but maybe they cannot hear . . .
This problem is quite common.---maybe on the next model?
I to the same thing, turn it on with a couple strong exhales, but always turn it off via the off button.
Battery Backup: EcoFlow Delta 2

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Something I Hate

Post by Janknitz » Mon Jun 10, 2019 5:20 pm

My trick to trigger smart start is to cover the vent with my finger--then it starts right up regardless of the type of mask I'm using.

Automatic stop never seems to work. I do it manually. I do find that annoying because for some reason I always wake up on the very edge of our king-sized bed and rolling over or reaching for that damned button can land me on the floor if I'm not careful, but I hate the rushing air when I take off the mask while still running.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Something I Hate

Post by chunkyfrog » Mon Jun 10, 2019 5:35 pm

I suspect the auto-off has been programmed to NOT be triggered by an apnea.
A fail-safe as it were.
If it is hazardous to reach over and shut it off---let it run.
With the mask off, it WILL stop eventually.
Alternatively, one could stuff the mask under a pillow.

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Something I Hate

Post by Okie bipap » Mon Jun 10, 2019 6:25 pm

If I simply disconnect the elbow from the front of my mast, the auto off works fine. With nasal pillows, the amount of air flow may not be enough to trigger the auto off function.
Growing old is mandatory, but growing up is optional.

User avatar
i_am_jim
Posts: 38
Joined: Wed Oct 26, 2005 2:14 pm

Re: Something I Hate

Post by i_am_jim » Mon Jun 10, 2019 6:47 pm

chunkyfrog wrote: ↑
Mon Jun 10, 2019 5:35 pm
With the mask off, it WILL stop eventually.
One morning I forgot to turn it off and it ran until I was ready to use it that night. In a way that made it convenient since the start doesn't work. Actually I became aware it was still running when I removed the humidity chamber to add water. Even with that large blast of air, it didn't turn itself off, I had to turn it off manually.
ResMed airsense10 w/integrated Humidifier
ResMed Mirage Swift Lt Pillow System
Image

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Something I Hate

Post by palerider » Mon Jun 10, 2019 7:30 pm

chunkyfrog wrote: ↑
Mon Jun 10, 2019 5:35 pm
I suspect the auto-off has been programmed to NOT be triggered by an apnea.
A fail-safe as it were.
If it is hazardous to reach over and shut it off---let it run.
With the mask off, it WILL stop eventually.
Alternatively, one could stuff the mask under a pillow.
It's programmed to do an auto-off when the pressure drops far enough that the only reason can be that the mask has been taken off.

When I first started using the P10, I forgot to change the mask type from 'full face', and my machine wouldn't turn off, because the p10 created enough back pressure that the machine thought it was just a really huge leak.

As soon as I set the mask type to 'pillows', auto-off started working every time.

Having low pressure makes it iffy, though, because the pressure drop isn't as dramatic.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Scluvsrain
Posts: 69
Joined: Wed Apr 04, 2018 9:01 am

Something I Hate

Post by Scluvsrain » Mon Jun 10, 2019 8:12 pm

Okie bipap wrote:If I simply disconnect the elbow from the front of my mast, the auto off works fine. With nasal pillows, the amount of air flow may not be enough to trigger the auto off function.
When I first got my Airsense 10 I had a nasal mask and it shut off automatically within a few seconds of removing my mask. But once I started using the pillows it no longer shut off, so I just do it manually now. I consider the reaching over and stretching my arm to turn off the machine just part of my morning aerobics......


Sent from my iPhone using Tapatalk

User avatar
Pugsy
Posts: 65445
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Something I Hate

Post by Pugsy » Mon Jun 10, 2019 9:09 pm

What pressure are you using?

This is in the troubleshooting section.
smartstarttroubleshooting.JPG
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Something I Hate

Post by chunkyfrog » Mon Jun 10, 2019 9:42 pm

Pugsy wrote: ↑
Mon Jun 10, 2019 9:09 pm
What pressure are you using?

This is in the troubleshooting section.

smartstarttroubleshooting.JPG
Mine is set for 9.2-13, and it turns on, but not off.
My P10 pillows are size small. Pillow size might be relevant.

User avatar
Pugsy
Posts: 65445
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Something I Hate

Post by Pugsy » Mon Jun 10, 2019 9:49 pm

chunkyfrog wrote: ↑
Mon Jun 10, 2019 9:42 pm
Pugsy wrote: ↑
Mon Jun 10, 2019 9:09 pm
What pressure are you using?

This is in the troubleshooting section.

smartstarttroubleshooting.JPG
Mine is set for 9.2-13, and it turns on, but not off.
My P10 pillows are size small. Pillow size might be relevant.
I think it is a combination of pillow size and pressure and the type of pillow maybe (like FX vs P10).
And 7 cm is NOT necessarily the cut off either. I know I tried 10 cm on my P10 and Auto Off still wouldn't work. I used the Small size pillow...I never did think to maybe try a larger pillow though.

What people can do to see if it is just the pillow resistance not allowing for the machine to sense if the human is still using it or bad sensors....just take the mask off of the long hose and see if the machine will then turn off like it should.
If it doesn't then there is a problem with the sensors..
If it does turn off then it's the mask that is being used that seems to be the culprit...and we just have to manually push the button to turn it off.
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Something I Hate

Post by chunkyfrog » Mon Jun 10, 2019 9:54 pm

Taking the mask off has no effect.
Unplugging the mask from the hose works IMMEDIATELY.
Go figure.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Something I Hate

Post by palerider » Mon Jun 10, 2019 10:19 pm

chunkyfrog wrote: ↑
Mon Jun 10, 2019 9:42 pm
Pugsy wrote: ↑
Mon Jun 10, 2019 9:09 pm
What pressure are you using?

This is in the troubleshooting section.

smartstarttroubleshooting.JPG
Mine is set for 9.2-13, and it turns on, but not off.
My P10 pillows are size small. Pillow size might be relevant.
Could be... I use the large pillows, so bigger holes, less back pressure.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.