What to do when all methods of sleep apnea treatment are exhausted?
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: What to do when all methods of sleep apnea treatment are exhausted?
I would get a nasal mask again and then post data from a typical night here.
It is possible that any leaks you had were minor or could be resolved. If you weren't having major leaks during that first year, then leaks may not be a big problem. Many doctors do not know how to get the best results. For example, many believe that an ahi under 5 is great, but that is the line where insurance will pay. Most people need to get under 2 to feel a noticeable improvement.
Since you are not able to sleep well with a full face mask, give a nasal mask another try. And this time, let a bunch of experienced (been there, done that) people help you to deal in success. Most people who come here, post graphs and try the advice do have pretty good improvements. I know it feels like you have tried everything already, but it is possible that your previous coaches didn't quite have the right experience to help you succeed. Most doctors have not slept with a cpap themselves, so they don't really understand how to make it work at its best.
It is possible that any leaks you had were minor or could be resolved. If you weren't having major leaks during that first year, then leaks may not be a big problem. Many doctors do not know how to get the best results. For example, many believe that an ahi under 5 is great, but that is the line where insurance will pay. Most people need to get under 2 to feel a noticeable improvement.
Since you are not able to sleep well with a full face mask, give a nasal mask another try. And this time, let a bunch of experienced (been there, done that) people help you to deal in success. Most people who come here, post graphs and try the advice do have pretty good improvements. I know it feels like you have tried everything already, but it is possible that your previous coaches didn't quite have the right experience to help you succeed. Most doctors have not slept with a cpap themselves, so they don't really understand how to make it work at its best.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
-
reader2580
- Posts: 123
- Joined: Tue May 14, 2013 6:57 pm
Re: What to do when all methods of sleep apnea treatment are exhausted?
To be clear, I have always had an APAP. I might need to get a new machine because it seems fairly noisy compared to when new.
I suspect the reason I wake up has nothing to do with my mask or CPAP. It is just that it seems to take a lot longer to fall back asleep with the mask on so that I end up taking it off.
I suspect the reason I wake up has nothing to do with my mask or CPAP. It is just that it seems to take a lot longer to fall back asleep with the mask on so that I end up taking it off.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
- ChicagoGranny
- Posts: 15392
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: What to do when all methods of sleep apnea treatment are exhausted?
Here's a big one: The doc has not pored over your SH Daily Details reports.reader2580 wrote: ↑Wed May 08, 2019 7:05 pmI have no reason not to believe the doctor.ChicagoGranny wrote: ↑Wed May 08, 2019 4:57 pmThat might be the problem. You took the doctor's advice.
I see you have 41 posts here so far. Have you posted SH charts for review by members?
-
reader2580
- Posts: 123
- Joined: Tue May 14, 2013 6:57 pm
Re: What to do when all methods of sleep apnea treatment are exhausted?
The doctor uses the Philips Respironics software to look at the data from my machine. This doctor has longer appointments than any other doctor I have ever seen. My last appointment with him was almost an hour.ChicagoGranny wrote: ↑Thu May 09, 2019 6:49 amHere's a big one: The doc has not pored over your SH Daily Details reports.
If one can't trust a medical professional then why see them? The first sleep center I went to was a joke. They only had a part time MD who only seemed interested in selling CPAP machines and supplies. The second appointment I had there was with a PA because the MD had moved on. After that I found my current doctor who I like very much. I can look at the data with SH, but I am not trained to analyze the data like a sleep medicine MD.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
- Miss Emerita
- Posts: 3765
- Joined: Sun Nov 04, 2018 8:07 pm
Re: What to do when all methods of sleep apnea treatment are exhausted?
As zoocrewphoto suggests, one method to try is to post a chart or two so the resident experts can take a look. I glanced back through your previous posts, and I might have missed it, but I think that’s something you haven’t ever done. There are lots of people on the forum ready to help if possible!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: What to do when all methods of sleep apnea treatment are exhausted?
I am still trying to figure out how you expect us to help if your apparently good doctor hasn't been able to sort out your problems and he has a lot more information about you than we do.
Did you really only want to know about alternatives to cpap therapy for sleep apnea or are you wanting to try to fix whatever problems you are having?
I don't mean to sound rude or uncaring but I can't figure out if you are just wanting to vent your frustrations or only wanted an answer to the "other alternative" question or get some ideas on how to make you happy with the mask.
Until you can sleep all night with a mask...you don't stand a snowball's chance in hell of ever feeling rested. It's hard enough when we can sleep all night with our masks to feel better. If we can't or won't keep the mask on all night...we can't expect to feel better. Just isn't going to happen.
Mask issues are critical to successful therapy though...gotta get that sleep with the mask on all night to have any chance of ever feeling better.
Did you really only want to know about alternatives to cpap therapy for sleep apnea or are you wanting to try to fix whatever problems you are having?
I don't mean to sound rude or uncaring but I can't figure out if you are just wanting to vent your frustrations or only wanted an answer to the "other alternative" question or get some ideas on how to make you happy with the mask.
Until you can sleep all night with a mask...you don't stand a snowball's chance in hell of ever feeling rested. It's hard enough when we can sleep all night with our masks to feel better. If we can't or won't keep the mask on all night...we can't expect to feel better. Just isn't going to happen.
Mask issues are critical to successful therapy though...gotta get that sleep with the mask on all night to have any chance of ever feeling better.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- ChicagoGranny
- Posts: 15392
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: What to do when all methods of sleep apnea treatment are exhausted?
What does your own testimony say? ------------------->
reader2580 wrote: ↑Thu May 09, 2019 10:02 amThe first sleep center I went to was a joke. They only had a part time MD who only seemed interested in selling CPAP machines and supplies. The second appointment I had there was with a PA because the MD had moved on.
- ChicagoGranny
- Posts: 15392
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: What to do when all methods of sleep apnea treatment are exhausted?
That's exactly what he has missed so far. He needs to post a Daily Details chart of a recent night.Miss Emerita wrote: ↑Thu May 09, 2019 10:11 amAs zoocrewphoto suggests, one method to try is to post a chart or two so the resident experts can take a look. I glanced back through your previous posts, and I might have missed it, but I think that’s something you haven’t ever done. There are lots of people on the forum ready to help if possible!
- Dog Slobber
- Posts: 4268
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: What to do when all methods of sleep apnea treatment are exhausted?
He also needs to answer questions asked of him.ChicagoGranny wrote: ↑Thu May 09, 2019 11:24 amThat's exactly what he has missed so far. He needs to post a Daily Details chart of a recent night.
For example, he was asked if he had tried a nasal mask and taped his mouth. He didn't mention tape.
reader2580 wrote: ↑Wed May 08, 2019 6:51 pmI no longer have a nasal mask as it has been over five years since I switched to full face. I used the nasal mask for at least a year without any relief.babydinosnoreless wrote: ↑Wed May 08, 2019 5:57 pmSo you have taped your mouth shut and worn the nasal mask and looked at the data to see if your still leaking air ?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: What to do when all methods of sleep apnea treatment are exhausted?
We are happy to help.
We are happy to hear your frustrated vent.
How can we help?
We are happy to hear your frustrated vent.
How can we help?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
Re: What to do when all methods of sleep apnea treatment are exhausted?
If everyone could trust their medical professionals to properly manage their treatment,reader2580 wrote: ↑Thu May 09, 2019 10:02 amThe doctor uses the Philips Respironics software to look at the data from my machine. This doctor has longer appointments than any other doctor I have ever seen. My last appointment with him was almost an hour.ChicagoGranny wrote: ↑Thu May 09, 2019 6:49 amHere's a big one: The doc has not pored over your SH Daily Details reports.
If one can't trust a medical professional then why see them?
THIS FORUM WOULD NOT EXIST.
We'll be here whenever you decide you actually want to do something to help yourself.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
reader2580
- Posts: 123
- Joined: Tue May 14, 2013 6:57 pm
Re: What to do when all methods of sleep apnea treatment are exhausted?
I wrote a big long reply earlier, but I apparently I closed the tab in my browser too quickly and it never finished posting. I hope to post some stuff from SleepyHead tonight.
I don't recall if I ever tried taping my mouth with a nasal mask. If I did do it previously it was maybe one or two nights. I don't know how well I would tolerate tape over my mouth, but I suppose I could try it for a few weeks to see if it helps. It wouldn't be a big deal to order a new nasal mask.
I am still wondering if there are other treatments for Sleep Apnea that aren't CPAP/APAP, mandibular advancement, or Inspire? I have an appointment for a follow-up with the ENT surgeon on Tuesday and I'll see if there are other surgical interventions that make sense. I've read here that a lot of the other surgical options are not great and there is at least one surgery that has a side affect of a chronic sore throat.
I don't recall if I ever tried taping my mouth with a nasal mask. If I did do it previously it was maybe one or two nights. I don't know how well I would tolerate tape over my mouth, but I suppose I could try it for a few weeks to see if it helps. It wouldn't be a big deal to order a new nasal mask.
I am still wondering if there are other treatments for Sleep Apnea that aren't CPAP/APAP, mandibular advancement, or Inspire? I have an appointment for a follow-up with the ENT surgeon on Tuesday and I'll see if there are other surgical interventions that make sense. I've read here that a lot of the other surgical options are not great and there is at least one surgery that has a side affect of a chronic sore throat.
_________________
| Machine: PR System One REMStar 60 Series Auto CPAP Machine |
| Mask: SleepWeaver Anew™ Full Face Mask with Headgear |
- ChicagoGranny
- Posts: 15392
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: What to do when all methods of sleep apnea treatment are exhausted?
Don't worry about a long reply. Just post one night of SH using the instructions in the thread at the top of the forum index page - "Pugsy's Pointers...SleepyHead Tutorial and TX Hints".reader2580 wrote: ↑Thu May 09, 2019 2:53 pmI wrote a big long reply earlier, but I apparently I closed the tab in my browser too quickly and it never finished posting. I hope to post some stuff from SleepyHead tonight.
The first thing people here want to see is the SH chart. Long discussions, if at all necessary, come later.
- babydinosnoreless
- Posts: 2365
- Joined: Fri Nov 02, 2018 2:53 pm
Re: What to do when all methods of sleep apnea treatment are exhausted?
I have not heard of any other treatments. I occasionally see those little anti snoring things advertised on facebook but I don't think anyone here has actually made something like that work. We all wouldn't be here if we had. Finding the right mask and getting my settings optimized made a word of difference for me.reader2580 wrote: ↑Thu May 09, 2019 2:53 pmI wrote a big long reply earlier, but I apparently I closed the tab in my browser too quickly and it never finished posting. I hope to post some stuff from SleepyHead tonight.
I don't recall if I ever tried taping my mouth with a nasal mask. If I did do it previously it was maybe one or two nights. I don't know how well I would tolerate tape over my mouth, but I suppose I could try it for a few weeks to see if it helps. It wouldn't be a big deal to order a new nasal mask.
I am still wondering if there are other treatments for Sleep Apnea that aren't CPAP/APAP, mandibular advancement, or Inspire? I have an appointment for a follow-up with the ENT surgeon on Tuesday and I'll see if there are other surgical interventions that make sense. I've read here that a lot of the other surgical options are not great and there is at least one surgery that has a side affect of a chronic sore throat.
Mouth taping doesn't sound very fun but a lot of people do it to avoid the full face masks. Others use chin straps, cervical collars or just scrunchies to keep their mouth closed at night.
I also have to wonder if your doctor thinks your apnea is being treated correctly by the machine if there are any other issues causing your sleep to be fragmented ? Medicines you might take, limb movements while sleeping, or other things keeping you from getting a full nights sleep?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
- Dog Slobber
- Posts: 4268
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: What to do when all methods of sleep apnea treatment are exhausted?
You may want to try a set of ResMed P10 Nasal Pillows.reader2580 wrote: ↑Thu May 09, 2019 2:53 pm
I don't recall if I ever tried taping my mouth with a nasal mask. If I did do it previously it was maybe one or two nights. I don't know how well I would tolerate tape over my mouth, but I suppose I could try it for a few weeks to see if it helps. It wouldn't be a big deal to order a new nasal mask.
Very minimal and a very popular mask. You probably will have to tape.
https://www.cpap.com/en/ca/productpage/ ... k-headgear
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2