Electrical Problems - PR 460P

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Sport303
Posts: 7
Joined: Sun Apr 21, 2019 10:29 am

Electrical Problems - PR 460P

Post by Sport303 » Sun Apr 21, 2019 2:40 pm

I have a Respironics SystemOne 460P. I’ve had two issues recently and was looking for some advice.

1) When connected to the humidifier, the unit will power off and give and error message after about 1-2 hours of use. I tried removing the cover and put some foam pieces in to provide a better seal on the pressure sensors, but that didn’t help. If I run it without the humidifier, it works fine.

2) I travel somewhat regularly overseas and have never had problems on 220V/230V power supplies. But this last trip, when in Vietnam and Cambodia, the unit would turn on for 1-2 seconds and then the power would just cut out. I thought the unit itself had died but then on the plane and when I got back to the US, it worked fine. I use the original power supply rated for 100V-240V.

Is a new power supply a good troubleshooting step? Anyone else with similar experiences or other ideas?

D.H.
Posts: 3534
Joined: Sun Sep 04, 2016 7:07 pm

Re: Electrical Problems - PR 460P

Post by D.H. » Sun Apr 21, 2019 3:07 pm

If the power supply is in proper working condition, any voltage from 100 to 240 (and probably even slightly outside that range) should work. Also, either 50 hz or 60 hz should be good.

My best guess (especially if this happened more than one accommodation) is that you power supply is not working properly, but you're "getting away with it" at 110 volts.

However, if this one happened at one accommodation, that accommodation is suspect!

BTW, did you every use it (or attempt to use it) on an aircraft? Sometimes, the power is 400 hz, and I'm not sure that that's compatible.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Electrical Problems - PR 460P

Post by Wulfman... » Sun Apr 21, 2019 7:06 pm

Sport303 wrote:
Sun Apr 21, 2019 2:40 pm
I have a Respironics SystemOne 460P. I’ve had two issues recently and was looking for some advice.

1) When connected to the humidifier, the unit will power off and give and error message after about 1-2 hours of use. I tried removing the cover and put some foam pieces in to provide a better seal on the pressure sensors, but that didn’t help. If I run it without the humidifier, it works fine.

2) I travel somewhat regularly overseas and have never had problems on 220V/230V power supplies. But this last trip, when in Vietnam and Cambodia, the unit would turn on for 1-2 seconds and then the power would just cut out. I thought the unit itself had died but then on the plane and when I got back to the US, it worked fine. I use the original power supply rated for 100V-240V.

Is a new power supply a good troubleshooting step? Anyone else with similar experiences or other ideas?
Contact http://www.acbio.com/

memberlist.php?mode=viewprofile&u=1002

They're a long-time member of this forum and an authorized repair station for ResMed, Philips/Respironics and others.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Sport303
Posts: 7
Joined: Sun Apr 21, 2019 10:29 am

Re: Electrical Problems - PR 460P

Post by Sport303 » Sun Apr 21, 2019 7:15 pm

D.H. wrote:
Sun Apr 21, 2019 3:07 pm
If the power supply is in proper working condition, any voltage from 100 to 240 (and probably even slightly outside that range) should work. Also, either 50 hz or 60 hz should be good.

My best guess (especially if this happened more than one accommodation) is that you power supply is not working properly, but you're "getting away with it" at 110 volts.

However, if this one happened at one accommodation, that accommodation is suspect!

BTW, did you every use it (or attempt to use it) on an aircraft? Sometimes, the power is 400 hz, and I'm not sure that that's compatible.
It was at 3 different accommodations, so the “getting away with it” possibility seems probable. It did work on the aircraft, although I didn’t realize the difference in Hz. I’m wondering if using it on board may have had a negative impact on the power supply. Either way it seems like swapping it out is a good bet.

Thanks!

Sport303
Posts: 7
Joined: Sun Apr 21, 2019 10:29 am

Re: Electrical Problems - PR 460P

Post by Sport303 » Mon Apr 22, 2019 7:38 pm

New power supply and now the humidifier works again. I’ll be able to test overseas on 220V in about a month. Looks like that was the issue.

D.H.
Posts: 3534
Joined: Sun Sep 04, 2016 7:07 pm

Re: Electrical Problems - PR 460P

Post by D.H. » Mon Apr 22, 2019 8:20 pm

Sport303 wrote:
Mon Apr 22, 2019 7:38 pm
New power supply and now the humidifier works again. I’ll be able to test overseas on 220V in about a month. Looks like that was the issue.
So the most likely explanation is that you tried to use it on an aircraft and damaged it. Please don't do this again unless you're sure that the electricity is 50hz or 60hz, or that the new power supply is clearly compatible with 400hz.

BTW, I'm not aware of any country on Earth that uses other than 50hz or 60hz (on Mars all bets are off :lol: ).

_________________