Waking up with sinus headache

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zonker
Posts: 11371
Joined: Fri Jun 19, 2015 4:36 pm

Re: Waking up with sinus headache

Post by zonker » Mon Apr 22, 2019 5:03 pm

Par301 wrote:
Mon Apr 22, 2019 4:52 pm
screenshot-20190422-183106.png
others will be by to tell you how they want your chart arranged. right now, i have a question. why is your machine set at a fixed pressure? in other words, it's in cpap mode. it will need to be in APAP mode, so that the pressure can rise to take care of your events.

oh, and yes, you want to have it posted right here. and keep posting any follow ups here that the experts will no doubt ask for.

good luck!
Last edited by zonker on Mon Apr 22, 2019 9:04 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Waking up with sinus headache

Post by Jas_williams » Mon Apr 22, 2019 8:43 pm

My suggestion on your graphs is turn off the pie chart

Also you will need more pressure I guess your on fixed pressure to try and prevent the CA’s there’s nothing your machine can do about those so let’s fix the obstructive and hypopnoea events with say 1 cm more of min pressure

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask