Pressure high at start

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zonker
Posts: 11499
Joined: Fri Jun 19, 2015 4:36 pm

Re: Pressure high at start

Post by zonker » Mon Apr 22, 2019 5:22 pm

Usman wrote: ↑
Mon Apr 22, 2019 8:08 am


do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.

the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: Pressure high at start

Post by Chris8243 » Mon Apr 22, 2019 7:24 pm

zonker wrote: ↑
Mon Apr 22, 2019 5:22 pm
Usman wrote: ↑
Mon Apr 22, 2019 8:08 am


do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.

the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.

good luck!
Zonker,
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?

User avatar
zonker
Posts: 11499
Joined: Fri Jun 19, 2015 4:36 pm

Re: Pressure high at start

Post by zonker » Mon Apr 22, 2019 7:38 pm

Chris8243 wrote: ↑
Mon Apr 22, 2019 7:24 pm
zonker wrote: ↑
Mon Apr 22, 2019 5:22 pm
Usman wrote: ↑
Mon Apr 22, 2019 8:08 am


do i have to install any software on my laptop for this purpose ?
and what is the use of bluetooth on dreamstation
please click on the last link at the bottom of this post to download the sleepyhead software.

the bluetooth is there to get the data generated by the machine to the doctor remotely. most of us turn it off because we don't want any doctors making changes to our data remotely.

good luck!
Zonker,
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
<sigh> it's been a long day and i'm not sure that's what i meant to type. i think i was thinking of some of the past conspiracy theories brought up and that's what i typed.

you are right, it's for apps.

at about the same time, i just pmed pugsy and palerider to look at some charts, but then linked to the wtong post.

i may have to go lay down for awhile.
:roll:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure high at start

Post by palerider » Mon Apr 22, 2019 7:42 pm

Chris8243 wrote: ↑
Mon Apr 22, 2019 7:24 pm
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
And you're going to trust them to tell you what's really going on?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: Pressure high at start

Post by Chris8243 » Mon Apr 22, 2019 8:48 pm

palerider wrote: ↑
Mon Apr 22, 2019 7:42 pm
Chris8243 wrote: ↑
Mon Apr 22, 2019 7:24 pm
I thought the modem was for the DME or doctor to see results & the bluetooth was just for the app. Am I wrong?
And you're going to trust them to tell you what's really going on?
Ahhhh, I do like my doctor. Last time I saw him, he said "I see you've changed your pressure. Good job. I would have done the same." I asked if he'd heard of Sleepyhead, he did, and I told him how I learned of it on this forum, blah blah, blah.

I turned the Bluetooth off on my machine long ago, the app is meaningless. The modem can only be turned off for 3 days at a time (it's set to come back on after 3 days.) I would remove it or turn it off permanently but apparently not an option.

User avatar
Usman
Posts: 15
Joined: Sun Apr 21, 2019 5:54 am

Re: Pressure high at start

Post by Usman » Tue Apr 23, 2019 12:02 pm

Can Anyone tell is there any way to check mask leakage without SD card directly on machine.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure high at start

Post by palerider » Tue Apr 23, 2019 12:33 pm

Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
There isn't one.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: Pressure high at start

Post by Dan_McD » Tue Apr 23, 2019 12:36 pm

Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure high at start

Post by palerider » Tue Apr 23, 2019 1:24 pm

Dan_McD wrote: ↑
Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Pressure high at start

Post by Pugsy » Tue Apr 23, 2019 1:46 pm

There is some sort of leak number that gets shown with the "Show AHI" turned on in the clinical setup menu area....so that people can get the limited data that the machine's LCD screen shows.
It's all explained in the clinical manual...that I guess someone hasn't bothered to get and/or read yet.
I forget off the top of my head which leak number it gives though....and no time to go look it up when people need to get their own manuals and do their own reading.
I provided the link...they have to do their own part of the work.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: Pressure high at start

Post by Dan_McD » Tue Apr 23, 2019 2:03 pm

palerider wrote: ↑
Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑
Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Pressure high at start

Post by Jas_williams » Tue Apr 23, 2019 2:28 pm

Dan_McD wrote: ↑
Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑
Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑
Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Sleepyhead

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Pressure high at start

Post by palerider » Tue Apr 23, 2019 2:30 pm

Dan_McD wrote: ↑
Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑
Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑
Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Looking at the data that's on the SD card.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: Pressure high at start

Post by Chris8243 » Tue Apr 23, 2019 2:44 pm

Pugsy wrote: ↑
Tue Apr 23, 2019 1:46 pm
There is some sort of leak number that gets shown with the "Show AHI" turned on in the clinical setup menu area....so that people can get the limited data that the machine's LCD screen shows.
It's all explained in the clinical manual...that I guess someone hasn't bothered to get and/or read yet.
I forget off the top of my head which leak number it gives though....and no time to go look it up when people need to get their own manuals and do their own reading.
I provided the link...they have to do their own part of the work.
The LCD screen shows "Mask Fit""as a percentage for one day, 7 days, 30 days. Not particularly helpful as it is after the fact. There ia a "Check Mask Fit" on the clinical menu which runs for 40 seconds. I don't use it, I guess because I can tell immediately if the mask is leaking as soon as I turn the machine on.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: Pressure high at start

Post by Dan_McD » Tue Apr 23, 2019 2:55 pm

palerider wrote: ↑
Tue Apr 23, 2019 2:30 pm
Dan_McD wrote: ↑
Tue Apr 23, 2019 2:03 pm
palerider wrote: ↑
Tue Apr 23, 2019 1:24 pm
Dan_McD wrote: ↑
Tue Apr 23, 2019 12:36 pm
Usman wrote: ↑
Tue Apr 23, 2019 12:02 pm
Can Anyone tell is there any way to check mask leakage without SD card directly on machine.
*
I'm not sure why you would want to do that, but if you downloaded the providers manual, you can use the "Check Mask Fit" setting.

"You can enable or disable the check mask fit setting. This feature allows the patient
to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit"
That won't tell you anything at all about your leaks during the night.
*
What will tell you "about your leaks during the night"?
Looking at the data that's on the SD card.
That's kinda what I thought, to check the usage, however long it is in use, on sleepyhead "Leak Rate". I thought the OP asked "is there any way to check mask leakage?" Dreamstation has that "Mask Fit" option. They quote "This feature allows the patient to check the fit of their mask prior to starting therapy. This is done by measuring the
amount of leak in the patient circuit".
You are right, at night if you knock your mask loose or pull the hose off, the "Mask Fit" feature is of no use.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up