CPAP IN ISRAEL

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Tammi
Posts: 2
Joined: Wed Apr 03, 2019 7:39 pm

CPAP IN ISRAEL

Post by Tammi » Wed Apr 03, 2019 7:46 pm

Help! This is the first time I’ve traveled out of the country and I’m going to Israel. I use a Resmed CPAP. The brick says 100-240V and 50-60Hz. Do I need a converter?

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: CPAP IN ISRAEL

Post by Jas_williams » Wed Apr 03, 2019 9:14 pm

Just a simple plug converter is required no transformer

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
ChicagoGranny
Posts: 15464
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: CPAP IN ISRAEL

Post by ChicagoGranny » Thu Apr 04, 2019 6:40 am

Tammi wrote:
Wed Apr 03, 2019 7:46 pm
Help! This is the first time I’ve traveled out of the country and I’m going to Israel. I use a Resmed CPAP. The brick says 100-240V and 50-60Hz. Do I need a converter?
The voltage is not a problem. You only need a type H plug adapter - https://www.amazon.com/OREI-Grounded-Un ... ef=sr_1_19

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: CPAP IN ISRAEL

Post by D.H. » Thu Apr 04, 2019 7:06 am

I've been to Israel twice with my CPAP and twice prior to being diagnosed (total four visits). The only issue I had there was that when I shut the light to go to bed, the power in the electric socket went out. That was solved my using a different socket and the extension cord I packed. As already stated, you'll need an adapter for the type "H" plugs. Unless you're staying in a very old accomodation, you will also be able to use a universal adapter.

Also, if you're staying in a "religious" accommodation, make sure that they won't have an issue with using the CPAP on the Sabbath (Friday night / Saturday morning), or on Jewish Holy Days. Since it is life-saving medical equipment, it should be exempt from this restriction (by religious law), but make sure that's understood.

Make sure that you also have a place to charge your phone (again, you won't need converter, only an adapter).

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Romalin
Posts: 6
Joined: Wed Apr 03, 2019 9:20 am

Re: CPAP IN ISRAEL

Post by Romalin » Thu Apr 04, 2019 7:27 am

Tammi,

If you could, when you get back from your trip please let us know about your CPAP experience, and any tips and tricks you may have for advice. This always helps others who may have similar concerns.

Thanks in advance! Enjoy your trip! :)

Rom

_________________
MachineMaskHumidifier

User avatar
metsfan302
Posts: 687
Joined: Fri Oct 09, 2015 9:44 pm
Location: Dewey Beach, Delaware

Re: CPAP IN ISRAEL

Post by metsfan302 » Thu Apr 04, 2019 10:38 am

Not sure if this is needed anymore

https://www.resmed.com/us/dam/documents ... r_eng3.pdf

If you want to use it in the plane or just snore :-)

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Prior - ResMed AirCurve ASV User
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a

freetimecreations
Posts: 145
Joined: Thu May 24, 2012 9:02 pm
Location: Utah, USA

Re: CPAP IN ISRAEL

Post by freetimecreations » Fri Apr 05, 2019 11:05 am

D.H. wrote:
Thu Apr 04, 2019 7:06 am
Also, if you're staying in a "religious" accommodation, make sure that they won't have an issue with using the CPAP on the Sabbath (Friday night / Saturday morning), or on Jewish Holy Days. Since it is life-saving medical equipment, it should be exempt from this restriction (by religious law), but make sure that's understood.
Here is a link from Apnea Board about Cpap machine on Shabbos. http://www.apneaboard.com/forums/Thread ... eat?page=2

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure 8.2cm-14.6cm, EPR 1, Humidity Auto, Contec CMS50F

yrnkrn
Posts: 172
Joined: Sat Feb 12, 2011 1:07 pm

Re: CPAP IN ISRAEL

Post by yrnkrn » Fri Apr 05, 2019 2:05 pm

The non-ground pins of the Israeli plug are identical to the EU plug. As the original plug is not grounded the simpler US to EU Plug will work just the same. These adapters are much cheaper and will also work across Europe.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame)
Additional Comments: OSCAR, CPAP=6, https://sites.google.com/view/palatal-prolapse/

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: CPAP IN ISRAEL

Post by Janknitz » Fri Apr 05, 2019 2:57 pm

[/quote]

Here is a link from Apnea Board about Cpap machine on Shabbos. http://www.apneaboard.com/forums/Thread ... eat?page=2
[/quote]


One of the many reasons I'm glad I was born to apikoros reformniks! :lol: (non-Orthodox)
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: CPAP IN ISRAEL

Post by D.H. » Fri Apr 05, 2019 8:13 pm

yrnkrn wrote:
Fri Apr 05, 2019 2:05 pm
The non-ground pins of the Israeli plug are identical to the EU plug. As the original plug is not grounded the simpler US to EU Plug will work just the same. These adapters are much cheaper and will also work across Europe.
This is true for the most part. The exception is that some older installations have flat slanty pins (not compatible with the Australia, New Zealand, Argentina type). The older slanty pins will also fit into newer installations.

Image

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

yrnkrn
Posts: 172
Joined: Sat Feb 12, 2011 1:07 pm

Re: CPAP IN ISRAEL

Post by yrnkrn » Sat Apr 06, 2019 1:28 am

D.H. wrote:
Fri Apr 05, 2019 8:13 pm
This is true for the most part. The exception is that some older installations have flat slanty pins (not compatible with the Australia, New Zealand, Argentina type). The older slanty pins will also fit into newer installations.
The three-pin round H adapter pictured above will not fit the "slanted" either.
I doubt this would present a real problem as
1) The standard was formally changed to round pins 30 years ago
2) All electronics appliances sold in Israel for the the last two (maybe three) decades came with round pins EU plug only
3) While CPAPs are not common, cell phone chargers with EU plug only are very common, this issue was already solved

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame)
Additional Comments: OSCAR, CPAP=6, https://sites.google.com/view/palatal-prolapse/

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: CPAP IN ISRAEL

Post by D.H. » Sat Apr 06, 2019 8:32 am

Most electric sockets in Israel are like the one below. They can accommodate the current standard (round pins, compatible with the ungrounded European plugs) and also the older slanty pins.



Image

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

Chris8243
Posts: 419
Joined: Sun Apr 08, 2018 10:42 pm
Location: Wyoming USA

Re: CPAP IN ISRAEL

Post by Chris8243 » Sat Apr 06, 2019 3:18 pm

freetimecreations wrote:
Fri Apr 05, 2019 11:05 am
D.H. wrote:
Thu Apr 04, 2019 7:06 am
Here is a link from Apnea Board about Cpap machine on Shabbos. http://www.apneaboard.com/forums/Thread ... eat?page=2
Now there is an interesting discussion on that link! From an explanation of Orthodox Jews not being able to turn on something that could create a spark on the Sabbath - to leaving the CPAP machine running the entire time to avoid needing to turn it on - to a suggestion of putting the mask on a manequin so the CPAP machine wouldn't turn off automatically - and that being a real possibility because his wife keeps wigs on a manequin head. Very complicated!

May I just add that I mean no disrespect, I'd like to learn about Orthodox Judaism since I realize that I know nothing about it.

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: CPAP IN ISRAEL

Post by D.H. » Sat Apr 06, 2019 3:44 pm

Chris8243 wrote:
Sat Apr 06, 2019 3:18 pm
freetimecreations wrote:
Fri Apr 05, 2019 11:05 am
D.H. wrote:
Thu Apr 04, 2019 7:06 am
Here is a link from Apnea Board about Cpap machine on Shabbos. http://www.apneaboard.com/forums/Thread ... eat?page=2
Now there is an interesting discussion on that link! From an explanation of Orthodox Jews not being able to turn on something that could create a spark on the Sabbath - to leaving the CPAP machine running the entire time to avoid needing to turn it on - to a suggestion of putting the mask on a mannequin so the CPAP machine wouldn't turn off automatically - and that being a real possibility because his wife keeps wigs on a manequin head. Very complicated!

May I just add that I mean no disrespect, I'd like to learn about Orthodox Judaism since I realize that I know nothing about it.

Note that Jewish religious laws almost always allow an exception for danger to life or limb. This even includes abortions (normally prohibited) when the mother's life or health are endangered.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

yrnkrn
Posts: 172
Joined: Sat Feb 12, 2011 1:07 pm

Re: CPAP IN ISRAEL

Post by yrnkrn » Sat Apr 06, 2019 11:52 pm

For the ultra orthodox BMC makes an especially-modified RESmart Auto CPAP device with "Shabbat mode" that makes the CPAP run in fixed pressure only, turns off the display and does not record data. Essentially makeing the CPAP behave like old non-digital CPAPs. To be ultra safe, one also installs a "sealing adapter" that actually seals all the CPAP sensors.

http://www.bhol.co.il/forums/topic.asp? ... rum_id=771
https://www.leami2000.com/

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame)
Additional Comments: OSCAR, CPAP=6, https://sites.google.com/view/palatal-prolapse/