BiPap back up.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
hev
Posts: 3
Joined: Fri Mar 29, 2019 12:18 am

BiPap back up.

Post by hev » Fri Mar 29, 2019 12:25 am

Hi, first time here.

I have read so many confusing info on back-ups for my BiPap, I would like to know if this would work.

I currently have a System one, I was thinking I could use a deep cycle marine battery and this inverter, since it has the clips. Would I need anything else?
image_25234.jpg
image_25234.jpg (120.08 KiB) Viewed 5481 times
Here's a link in case the pic can't be seen.

https://www.harborfreight.com/400-watt- ... ioEALw_wcB

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: BiPap back up.

Post by palerider » Fri Mar 29, 2019 10:59 am

Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jimbud
Posts: 1486
Joined: Sat Dec 01, 2018 8:19 pm

Re: BiPap back up.

Post by jimbud » Fri Mar 29, 2019 11:35 am

palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Why not a deep cycle MARINE battery. I have not heard that before. I believe you. I would just like to know what the reason is.
JPB

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: BiPap back up.

Post by Jas_williams » Fri Mar 29, 2019 12:52 pm

jimbud wrote: ↑
Fri Mar 29, 2019 11:35 am
palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Why not a deep cycle MARINE battery. I have not heard that before. I believe you. I would just like to know what the reason is.
JPB
More expensive for no benefit

User avatar
Dog Slobber
Posts: 4340
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: BiPap back up.

Post by Dog Slobber » Fri Mar 29, 2019 1:09 pm

jimbud wrote: ↑
Fri Mar 29, 2019 11:35 am
palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Why not a deep cycle MARINE battery. I have not heard that before. I believe you. I would just like to know what the reason is.
JPB
Deep cycle MARINE batteries are often not true deep cycle, instead they may be hybrid between a starting battery and a deep cycle battery.

You pay more and do not get as much of the same deep cycle charge life span.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: BiPap back up.

Post by palerider » Fri Mar 29, 2019 1:20 pm

jimbud wrote: ↑
Fri Mar 29, 2019 11:35 am
palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Why not a deep cycle MARINE battery. I have not heard that before. I believe you. I would just like to know what the reason is.
JPB
marine batteries are built more strongly in order to survive the pounding they receive on the waves, hence they usuallycost more (and are probably heavier for a given capacity).

There's no need to pay that premium.

Also, there are marine batteries that aren't even deep cycle... Further complicating the whole "marine" issue.

This is all covered much better in capnloki's excellent"choosing a battery" thread.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jimbud
Posts: 1486
Joined: Sat Dec 01, 2018 8:19 pm

Re: BiPap back up.

Post by jimbud » Fri Mar 29, 2019 2:53 pm

Thank you all for the answers. Glad to know no exploding cpap danger. I just bought a small battery powered boat last year and currently have a deep cycle battery and would not want to use it for cpap if there was some compelling reason I wasn't aware of.
In summary no reason not to use it, just no reason to go to the extra expense. And, man is it heavy.

JPB

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: BiPap back up.

Post by palerider » Fri Mar 29, 2019 3:30 pm

jimbud wrote: ↑
Fri Mar 29, 2019 2:53 pm
In summary no reason not to use it, just no reason to go to the extra expense. And, man is it heavy.

JPB
Correct!
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

hev
Posts: 3
Joined: Fri Mar 29, 2019 12:18 am

Re: BiPap back up.

Post by hev » Sun Mar 31, 2019 8:54 am

palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Thanks for the replies.

Do you mean one of these?
http://www.gocpap.com/philips-respironi ... for-47-00/

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: BiPap back up.

Post by palerider » Sun Mar 31, 2019 11:59 am

hev wrote: ↑
Sun Mar 31, 2019 8:54 am
palerider wrote: ↑
Fri Mar 29, 2019 10:59 am
Inverters are bad, they'll waste half of your battery.

Just get the DC cable that does with your machine.

Also, DO NOT GET a deep cycle "marine" battery.

Just get a deep cycle battery, like is used in mobility scooters.
Thanks for the replies.

Do you mean one of these?
http://www.gocpap.com/philips-respironi ... for-47-00/
IF you have a system one 60 series machine, if not, you need a different cable... the one for the system one 50 series. But, yes, broadly, that's the idea. Power the machine directly from the battery.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.