Does your cpap make you feel inflated?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Axlrose
Posts: 1
Joined: Thu Mar 21, 2019 11:52 pm

Does your cpap make you feel inflated?

Post by Axlrose » Thu Mar 21, 2019 11:59 pm

Wondering if anyone gets this feeling? My cpap makes me feel kind of like I need to burp out a bunch of air in the morning. Lungs feel kind of bloated. I know it's forcing air, so maybe that's a byproduct? Or maybe that means the air pressure isn't right?

Thanks for any thoughts.

User avatar
Pugsy
Posts: 65412
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Does your cpap make you feel inflated?

Post by Pugsy » Fri Mar 22, 2019 7:32 am

Welcome to the forum.

It's called aerophagia and fairly common. Sometimes mildly annoying and sometimes extremely uncomfortable and makes a person ill.

wiki/index.php/Aerophagia

There are some things that can be done to reduce it. Adding in exhale relief is the first one I usually suggest if it's available.

What machine brand and model are you using? What are your current settings? If we had more information we could maybe offer more specific ideas to help reduce the problem if it is more than mildly annoying.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11494
Joined: Fri Jun 19, 2015 4:36 pm

Re: Does your cpap make you feel inflated?

Post by zonker » Fri Mar 22, 2019 5:35 pm

Axlrose wrote: ↑
Thu Mar 21, 2019 11:59 pm
Wondering if anyone gets this feeling? My cpap makes me feel kind of like I need to burp out a bunch of air in the morning. Lungs feel kind of bloated. I know it's forcing air, so maybe that's a byproduct? Or maybe that means the air pressure isn't right?

Thanks for any thoughts.
welcome aboard. pugsy has already said it, so no need for me to repeat it. i do want to emphasize that if you do need to show us some data so that the experts can help you more, please click on the first link in my sig to find out how to go aout it.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

drumma
Posts: 7
Joined: Thu Mar 21, 2019 12:48 pm

Re: Does your cpap make you feel inflated?

Post by drumma » Sat Mar 23, 2019 5:48 am

My auto set was @ 18 and I got that feeling> i had it lowered to 17, not a noticeable change. Asked the doctor for 15 and all seems better now. YMMV

_________________
Mask
Additional Comments: also . F&P Simplus . Smply Go portable oxygen concentrator adding 1 litre to CPAP

User avatar
zorki1c
Posts: 253
Joined: Mon Jun 10, 2013 5:24 pm
Location: Intermountain West

Re: Does your cpap make you feel inflated?

Post by zorki1c » Sat Mar 23, 2019 2:57 pm

I've got mine set to inflate to 35 psi -- same as my tires. :D

_________________
MachineMask
Additional Comments: Use the P30i in the cooler months and the P10 the rest of the time.

User avatar
zonker
Posts: 11494
Joined: Fri Jun 19, 2015 4:36 pm

Re: Does your cpap make you feel inflated?

Post by zonker » Sat Mar 23, 2019 3:02 pm

zorki1c wrote: ↑
Sat Mar 23, 2019 2:57 pm
I've got mine set to inflate to 35 psi -- same as my tires. :D
so, you look like this?

michelinman.gif

:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

iamsleepynow
Posts: 7
Joined: Sat Mar 23, 2019 6:41 pm

Re: Does your cpap make you feel inflated?

Post by iamsleepynow » Sat Mar 23, 2019 7:54 pm

For years I had this problem. I noticed it my first night in the sleep lab. Sometimes Bilevel therapy helps. I've worked with my Dr to adjust to settings where I did not experience this discomfort. Sometimes I needed to reduce my pressure below what I needed but I could not tolerate it higher. As I get new xPAPs now and then, I can't say exactly what changed. xPAPS use different algorithms. And I'm different as I age.

iamsleepynow
Posts: 7
Joined: Sat Mar 23, 2019 6:41 pm

Re: Does your cpap make you feel inflated?

Post by iamsleepynow » Sat Mar 23, 2019 8:06 pm

My first advanced ovarian cancer symptom was, in fact, a severe case of aerophagia from my auto bilevel xPAP. I thought of calling 911. Then it happened again one month later and I wanted to call 911 again but did not. My belly was so full of air from the CPAP I thought my intestines would explode. Really, I was in that much pain. What I learned later was going on was that I had a belly full of fluid and it made it so I did not have sufficient room to breathe properly. This was an extreme case, a rare case. I mention it here because ovarian cancer is so darn difficult to diagnose. If one woman is helped by this post to be mindful of abdominal bloating then my discomfort with posting is worthwhile.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Does your cpap make you feel inflated?

Post by Miss Emerita » Sat Mar 23, 2019 9:04 pm

Thank you, iamsleepynow. It’s generous of you to share this important information. Much appreciated.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Steve of Cornubia
Posts: 73
Joined: Tue Mar 12, 2013 10:19 pm
Location: Brisbane, Australia

Re: Does your cpap make you feel inflated?

Post by Steve of Cornubia » Sat Mar 23, 2019 9:49 pm

Aerophagia. We all have to suffer it to some degree. Stands to reason when we're being inflated all night. APAP can help reduce it but, other than that, all you can do is make sure you're using no more pressure than you have to.

The relevant sphincter(s) can presumably stay closed until a critical pressure is reached, at which point I'm guessing it is forced open. I can sometimes feel this happening as my pressure rises during the ramp period.

It is important to deal with this however, especially if you're getting on in years. As we all know, once you pass 60, you can never completely trust a fart :oops:

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Does your cpap make you feel inflated?

Post by palerider » Sat Mar 23, 2019 11:19 pm

Steve of Cornubia wrote: ↑
Sat Mar 23, 2019 9:49 pm
Aerophagia. We all have to suffer it to some degree.
Not at all true, most don't suffer from it at all. Unfortunate for the minority that do suffer from it, but a minority nonetheless.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Firefox
Posts: 12
Joined: Wed Jul 27, 2016 9:05 am
Location: South Africa

Re: Does your cpap make you feel inflated?

Post by Firefox » Mon Mar 25, 2019 10:11 am

Yes sometimes. Let your Dr /Sleep Specialist look at your Rise time air flow on your machine. I struggled alot with it.Turns out my Rise time was set for an overweight guy that I wasn't(it was at 500). Sometimes when you pick up weight or loose weight it can change. Once my Sleep technician changed mine (after I made them aware of it) down I felt alot better. There are no reasons to suffer. Make them aware of it
Resmed Astral 150,Powerstation II, Resperonics Comfortfull 2,Resperonics ComfortGel Full, Resmed F30

zoomzoom
Posts: 142
Joined: Tue Mar 05, 2019 7:39 pm

Re: Does your cpap make you feel inflated?

Post by zoomzoom » Mon Mar 25, 2019 11:32 pm

Firefox wrote: ↑
Mon Mar 25, 2019 10:11 am
Yes sometimes. Let your Dr /Sleep Specialist look at your Rise time air flow on your machine. I struggled alot with it.Turns out my Rise time was set for an overweight guy that I wasn't(it was at 500). Sometimes when you pick up weight or loose weight it can change. Once my Sleep technician changed mine (after I made them aware of it) down I felt alot better. There are no reasons to suffer. Make them aware of it
Rise time is something I haven't come across yet. What is that exactly and how can I check mine?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Does your cpap make you feel inflated?

Post by palerider » Mon Mar 25, 2019 11:34 pm

zoomzoom wrote: ↑
Mon Mar 25, 2019 11:32 pm
Firefox wrote: ↑
Mon Mar 25, 2019 10:11 am
Yes sometimes. Let your Dr /Sleep Specialist look at your Rise time air flow on your machine. I struggled alot with it.Turns out my Rise time was set for an overweight guy that I wasn't(it was at 500). Sometimes when you pick up weight or loose weight it can change. Once my Sleep technician changed mine (after I made them aware of it) down I felt alot better. There are no reasons to suffer. Make them aware of it
Rise time is something I haven't come across yet. What is that exactly and how can I check mine?
Your machine doesn't have it, and it doesn't really have any affect on aerophagia.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

zoomzoom
Posts: 142
Joined: Tue Mar 05, 2019 7:39 pm

Re: Does your cpap make you feel inflated?

Post by zoomzoom » Mon Mar 25, 2019 11:43 pm

palerider wrote: ↑
Mon Mar 25, 2019 11:34 pm
zoomzoom wrote: ↑
Mon Mar 25, 2019 11:32 pm
Firefox wrote: ↑
Mon Mar 25, 2019 10:11 am
Yes sometimes. Let your Dr /Sleep Specialist look at your Rise time air flow on your machine. I struggled alot with it.Turns out my Rise time was set for an overweight guy that I wasn't(it was at 500). Sometimes when you pick up weight or loose weight it can change. Once my Sleep technician changed mine (after I made them aware of it) down I felt alot better. There are no reasons to suffer. Make them aware of it
Rise time is something I haven't come across yet. What is that exactly and how can I check mine?
Your machine doesn't have it, and it doesn't really have any affect on aerophagia.
Oh OK, thanks for letting me know!