Distilled Water or Filtered?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Distilled Water or Filtered?

Post by raisedfist » Fri Mar 15, 2019 11:45 am

It doesn't matter as long as you're using water that is drinkable

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Dog Slobber
Posts: 4330
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Distilled Water or Filtered?

Post by Dog Slobber » Fri Mar 15, 2019 12:07 pm

KnightSleeper wrote:
Fri Mar 15, 2019 11:52 am
Question for Palerider and those in the know... When regular water leaves deposits in the tub, might it also leave deposits somewhere else in the machine... on the way out??? :?
The biggest reason for deposits is evaporation. All the water evaporates leaving deposits in the tank. The minerals don't evaporate, and are left in the tank.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65337
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Distilled Water or Filtered?

Post by Pugsy » Fri Mar 15, 2019 12:26 pm

KnightSleeper wrote:
Fri Mar 15, 2019 12:15 pm
Thanks for your reply... Can any moisture form and/or evaporate in other parts of the machine... or is this impossible? - - And are there any other reasons that could cause deposits being left in the machine??? Thanks!
Only possible way would be water spilled backwards into the machine and that's darn near impossible now with the way the water chambers are designed now. Takes some weird positions for the water to flow backwards into the blower.
Now the blowers back when I first started therapy on those old machines were fairly easy to get water into the blower if you jostled or turned the machine over.
If you did that though....deposits were the least of your worries. Fried electronics was the usual result.
For deposits to happen the area first has to get water on it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4330
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Distilled Water or Filtered?

Post by Dog Slobber » Fri Mar 15, 2019 12:28 pm

Not in the device itself.

The tank is full of water, and the water might contain sediment.

The water is evaporated, and the vapour does NOT contain any sediment, leaving sediment in the tank. The vapour is blown into the hose and mask and may condense back into liquid within the tube and mask, but no longer has particles in it to leave.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 12:29 pm

KnightSleeper wrote:
Fri Mar 15, 2019 11:52 am
Question for Palerider and those in the know... When regular water leaves deposits in the tub, might it also leave deposits somewhere else in the machine... on the way out??? :?
No, the deposits are left in the tank because the water vapor evaporates, it can't carry the minerals, or bacteria, or viruses, and certainly not brain eating amoeba.

Evaporation of water is the first step in distillation, the second is condensation of that water vapor back to water... In a still, this distilled water.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 12:30 pm

KnightSleeper wrote:
Fri Mar 15, 2019 12:15 pm
Dog Slobber wrote:
Fri Mar 15, 2019 12:07 pm
KnightSleeper wrote:
Fri Mar 15, 2019 11:52 am
Question for Palerider and those in the know... When regular water leaves deposits in the tub, might it also leave deposits somewhere else in the machine... on the way out??? :?
The biggest reason for deposits is evaporation. All the water evaporates leaving deposits in the tank. The minerals don't evaporate, and are left in the tank.
Thanks for your reply... Can any moisture form and/or evaporate in other parts of the machine... or is this impossible? - - And are there any other reasons that could cause deposits being left in the machine??? Thanks!
Under normal operation, no. The humidifier is the last part of the CPAP.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 12:32 pm

Dog Slobber wrote:
Fri Mar 15, 2019 12:28 pm
Not in the device itself.

The tank is full of water, and the water might contain sediment.

The water is evaporated, and the vapour does NOT contain any sediment, leaving sediment in the tank. The vapour is blown into the hose and mask and may condense back into liquid within the tube and mask, but no longer has particles in it to leave.
The water that condenses in the hose is, in effect, distilled water, even if you started with tap water.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Dog Slobber
Posts: 4330
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Distilled Water or Filtered?

Post by Dog Slobber » Fri Mar 15, 2019 12:49 pm

palerider wrote:
Fri Mar 15, 2019 12:32 pm
Dog Slobber wrote:
Fri Mar 15, 2019 12:28 pm
Not in the device itself.

The tank is full of water, and the water might contain sediment.

The water is evaporated, and the vapour does NOT contain any sediment, leaving sediment in the tank. The vapour is blown into the hose and mask and may condense back into liquid within the tube and mask, but no longer has particles in it to leave.
The water that condenses in the hose is, in effect, distilled water, even if you started with tap water.
I hope you're not suggesting double distilled isn't twice as good as single distilled.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 1:34 pm

KnightSleeper wrote:
Fri Mar 15, 2019 1:18 pm
palerider wrote:
Fri Mar 15, 2019 12:32 pm
Dog Slobber wrote:
Fri Mar 15, 2019 12:28 pm
Not in the device itself.

The tank is full of water, and the water might contain sediment.

The water is evaporated, and the vapour does NOT contain any sediment, leaving sediment in the tank. The vapour is blown into the hose and mask and may condense back into liquid within the tube and mask, but no longer has particles in it to leave.
The water that condenses in the hose is, in effect, distilled water, even if you started with tap water.
Thanks Palerider. Do any of the chemicals in the tap water, ie, chlorine, fluiride, etc... have any impact on anyone? I realize we sometimes drink those chemicals in our water, but in this application, not that I would know anything at all about this, I'm talking about the extra effect that Zonker (not our Zonker!!!) might get using a bong....??? :|
If any, then less than taking a shower and breathing in the steam from the hot water, and the aerosolized water droplets from the spray.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 1:36 pm

Dog Slobber wrote:
Fri Mar 15, 2019 12:49 pm
palerider wrote:
Fri Mar 15, 2019 12:32 pm
Dog Slobber wrote:
Fri Mar 15, 2019 12:28 pm
Not in the device itself.

The tank is full of water, and the water might contain sediment.

The water is evaporated, and the vapour does NOT contain any sediment, leaving sediment in the tank. The vapour is blown into the hose and mask and may condense back into liquid within the tube and mask, but no longer has particles in it to leave.
The water that condenses in the hose is, in effect, distilled water, even if you started with tap water.
I hope you're not suggesting double distilled isn't twice as good as single distilled.
"Shakes fist" we're not talking vodka here.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11456
Joined: Fri Jun 19, 2015 4:36 pm

Re: Distilled Water or Filtered?

Post by zonker » Fri Mar 15, 2019 3:07 pm

after reading this thread, i've decided to quit buying ANY sort of water and use dog slobber.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

nicholasjh1
Posts: 517
Joined: Wed Jun 14, 2017 9:07 am

Re: Distilled Water or Filtered?

Post by nicholasjh1 » Fri Mar 15, 2019 3:11 pm

palerider wrote:
Fri Mar 15, 2019 11:29 am
nicholasjh1 wrote:
Fri Mar 15, 2019 9:37 am
depends on how hard your water is. Mine has a ton of chlorine, and even filtered starts leaving deposits after only 1 night, so I always try to use distilled and will only use filtered as backup if I run out of distilled. otherwise I have to constantly rinse it with vinegar.
Chlorine, or more likely, chloramine, doesn't leave deposits, it evaporates. (chlorine is a gas, after all). And, vinegar would not remove it.
point is there is something hard in the water... thanks for the clarification
Instead of Sleep apnea it should be called "Sleep deprivation, starving of oxygen, being poisoned by high CO2 levels, damaging the body and brain while it's supposed to be healing so that you constantly get worse and can never get healthy Apnea"

nicholasjh1
Posts: 517
Joined: Wed Jun 14, 2017 9:07 am

Re: Distilled Water or Filtered?

Post by nicholasjh1 » Fri Mar 15, 2019 3:12 pm

palerider wrote:
Fri Mar 15, 2019 11:31 am
nicholasjh1 wrote:
Fri Mar 15, 2019 9:43 am
Also I'll back up that distilled water is in fact NOT sterile. when I use sprouts distilled water I always get this weird pink mold, so I've stopped ever using it for that reason. I use king soopers instead.
The pink stuff isn't any form of mold, it's an airborne bacteria: http://edenprairieweblogs.org/scottneal/post/1338/
I realize that. Didn't think it was worth distinguishing.
Instead of Sleep apnea it should be called "Sleep deprivation, starving of oxygen, being poisoned by high CO2 levels, damaging the body and brain while it's supposed to be healing so that you constantly get worse and can never get healthy Apnea"

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Distilled Water or Filtered?

Post by palerider » Fri Mar 15, 2019 3:17 pm

nicholasjh1 wrote:
Fri Mar 15, 2019 3:12 pm
palerider wrote:
Fri Mar 15, 2019 11:31 am
nicholasjh1 wrote:
Fri Mar 15, 2019 9:43 am
Also I'll back up that distilled water is in fact NOT sterile. when I use sprouts distilled water I always get this weird pink mold, so I've stopped ever using it for that reason. I use king soopers instead.
The pink stuff isn't any form of mold, it's an airborne bacteria: http://edenprairieweblogs.org/scottneal/post/1338/
I realize that. Didn't think it was worth distinguishing.
Might as well call it blood, or a tree, or dog slobber if it's not "worth distinguishing".

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11456
Joined: Fri Jun 19, 2015 4:36 pm

Re: Distilled Water or Filtered?

Post by zonker » Fri Mar 15, 2019 4:06 pm

palerider wrote:
Fri Mar 15, 2019 3:17 pm


Might as well call it blood, or a tree, or dog slobber if it's not "worth distinguishing".
is this another vote in favor of my new plan to use dog slobber in my humidifier?