SH chart - what is happening here?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Spot123
Posts: 80
Joined: Mon Jan 14, 2019 8:16 pm

SH chart - what is happening here?

Post by Spot123 » Wed Mar 13, 2019 7:59 pm

Hi
I was wondering what is happening with this shape curve (first screenshot)-- persisted about 40 minutes last night. The second screenshot is "normal".

Thanks
screenshot-20190313-215438.png
screenshot-20190313-215438.png (152.18 KiB) Viewed 1623 times
screenshot-20190313-215428.png
screenshot-20190313-215428.png (158.01 KiB) Viewed 1622 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SH chart - what is happening here?

Post by palerider » Wed Mar 13, 2019 8:35 pm

Spot123 wrote:
Wed Mar 13, 2019 7:59 pm
Hi
I was wondering what is happening with this shape curve (first screenshot)-- persisted about 40 minutes last night. The second screenshot is "normal".
Beat's me, I can't see squat with those Y ranges. (stupid auto ranging).

Right click the 'flow rate', choose 'y axis', then 'override' and put in something sane, like -60 to 60. then pick 'dotted lines' and turn on the zero line, and repost the pics.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Spot123
Posts: 80
Joined: Mon Jan 14, 2019 8:16 pm

Re: SH chart - what is happening here?

Post by Spot123 » Thu Mar 14, 2019 4:30 pm

OK, reposting with better Y-axis range.
first one is normal. second is funny-looking
screenshot-20190314-182731.png
screenshot-20190314-182731.png (154.6 KiB) Viewed 1572 times
screenshot-20190314-182738.png
screenshot-20190314-182738.png (151.11 KiB) Viewed 1572 times
Thanks for any ideas.

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: SH chart - what is happening here?

Post by katestyles » Thu Mar 14, 2019 4:38 pm

It looks like flow limitation to me - but I am not the expert.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: SH chart - what is happening here?

Post by Jas_williams » Thu Mar 14, 2019 4:49 pm

Spot123 wrote:
Thu Mar 14, 2019 4:30 pm
OK, reposting with better Y-axis range.
first one is normal. second is funny-looking

screenshot-20190314-182731.png
screenshot-20190314-182738.png

Thanks for any ideas.
Second one looks like mouth breathing to me


Breath in through the nose and out through the mouth arround your mask somehow as you have little to no expiratory flow

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SH chart - what is happening here?

Post by palerider » Thu Mar 14, 2019 8:21 pm

Spot123 wrote:
Thu Mar 14, 2019 4:30 pm
OK, reposting with better Y-axis range.
first one is normal. second is funny-looking.
Now, those, I can see... thanks.

I'm with Jas, it looks like some of it is mouth exhaling, you're inhaling through the mask, but most of the air seems to be just 'disappearing'... probably out your mouth, instead of back into the mask.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SH chart - what is happening here?

Post by palerider » Thu Mar 14, 2019 8:24 pm

katestyles wrote:
Thu Mar 14, 2019 4:38 pm
It looks like flow limitation to me - but I am not the expert.
While expiratory flow limitation is possible, it's not often seen... usually it's inspiratory flow limitation (and that's what causes the machine to react.

for inspiratory FL, you have to look at the shape of the curves above the dotted line, are they flattened, or chair shaped? (or, basically, not very rounded), of that's the case, then they're FL breaths, and the FL chart should be kicking up.

Those weird breaths are pretty rounded, and the FL chart is flat.

Like I said above, I think it's mouth leaking on exhale. (but not certain).

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: SH chart - what is happening here?

Post by Jay Aitchsee » Thu Mar 14, 2019 8:40 pm

Yes, exhalation through the mouth rather than through the nose.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: SH chart - what is happening here?

Post by katestyles » Thu Mar 14, 2019 8:42 pm

palerider wrote:
Thu Mar 14, 2019 8:24 pm
katestyles wrote:
Thu Mar 14, 2019 4:38 pm
It looks like flow limitation to me - but I am not the expert.
While expiratory flow limitation is possible, it's not often seen... usually it's inspiratory flow limitation (and that's what causes the machine to react.

for inspiratory FL, you have to look at the shape of the curves above the dotted line, are they flattened, or chair shaped? (or, basically, not very rounded), of that's the case, then they're FL breaths, and the FL chart should be kicking up.

Those weird breaths are pretty rounded, and the FL chart is flat.

Like I said above, I think it's mouth leaking on exhale. (but not certain).
Wouldn't that show up as a leak on the leak chart? Her is some of my flow. Is it the same? or different? I seek to be educated here. This flow had no flow restrict or leak show up with it either.
educateme.png
educateme.png (18.88 KiB) Viewed 1533 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
Pugsy
Posts: 65197
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SH chart - what is happening here?

Post by Pugsy » Thu Mar 14, 2019 8:48 pm

katestyles wrote:
Thu Mar 14, 2019 8:42 pm
Wouldn't that show up as a leak on the leak chart?
Not necessarily. There's breathing out the mouth gently and calmly and then there's a mass exodus of air out the mouth with the tornado sounds.

Here's an example of known mouth breathing and how it looks on the leak graph. I happened to wake up one morning and I was mouth breathing. Nice gentle even calm mouth breathing with no mass exodus of air out my mouth. I decided to lay there and continue breathing like that so I could see for sure what it looks like on the leak graph. I circled the known awake and mouth breathing time at the end of the night.
This is why I tell people who think they are mouth breathing that they need to actually look at the leak reports/graphs before they run out and buy a full face mask that they don't want. The little bit of mouth breathing I did...well within the machine's ability to compensate. Now don't ask me what happened earlier in the night...I was asleep, didn't wake me up and I don't know and don't really care. :lol: I suspect mask movement. Since my time with the Bleep I have come to realize that I wasn't mouth breathing nearly as much as I thought I was.

Image

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SH chart - what is happening here?

Post by palerider » Thu Mar 14, 2019 8:53 pm

katestyles wrote:
Thu Mar 14, 2019 8:42 pm
palerider wrote:
Thu Mar 14, 2019 8:24 pm
katestyles wrote:
Thu Mar 14, 2019 4:38 pm
It looks like flow limitation to me - but I am not the expert.
While expiratory flow limitation is possible, it's not often seen... usually it's inspiratory flow limitation (and that's what causes the machine to react.

for inspiratory FL, you have to look at the shape of the curves above the dotted line, are they flattened, or chair shaped? (or, basically, not very rounded), of that's the case, then they're FL breaths, and the FL chart should be kicking up.

Those weird breaths are pretty rounded, and the FL chart is flat.

Like I said above, I think it's mouth leaking on exhale. (but not certain).
Wouldn't that show up as a leak on the leak chart? Her is some of my flow. Is it the same? or different? I seek to be educated here. This flow had no flow restrict or leak show up with it either.
Like Pugsy says, 'not necessarily'... all depends on what you're doing with your nasopharynx.

I can put on my nasal mask, open my mouth, stick out my tongue, and no air comes out. I can breathe that way through the cpap, or I can do whatever to make things move around in the back of my mouth, block off my nasal cavity, and breathe through my mouth, and have the cpap think I'm having an obstructive apnea... or, i can hit the middle position, and air goes in my nose, around the corner, and right back out my mouth. It all depends on how you 'flip the valve(s)'.

So, you can breathe in, block that off, and exhale out your mouth... it doesn't show up as a leak at that point... since there's no air actually flowing (except out the vent). If it stays that way too long, you'll see an obstructive get flagged.

Make sense?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: SH chart - what is happening here?

Post by katestyles » Thu Mar 14, 2019 9:04 pm

palerider wrote:
Thu Mar 14, 2019 8:53 pm

Make sense?
errrr......no.

Don't give up on me though - I'm trying.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SH chart - what is happening here?

Post by palerider » Thu Mar 14, 2019 9:06 pm

katestyles wrote:
Thu Mar 14, 2019 9:04 pm
palerider wrote:
Thu Mar 14, 2019 8:53 pm

Make sense?
errrr......no.

Don't give up on me though - I'm trying.
What part didn't make sense, ask questions... that way I can just try rewording that bit :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: SH chart - what is happening here?

Post by katestyles » Thu Mar 14, 2019 9:08 pm

palerider wrote:
Thu Mar 14, 2019 9:06 pm
katestyles wrote:
Thu Mar 14, 2019 9:04 pm
palerider wrote:
Thu Mar 14, 2019 8:53 pm

Make sense?
errrr......no.

Don't give up on me though - I'm trying.
What part didn't make sense, ask questions... that way I can just try rewording that bit :)
I can't ask questions - I have to go put on my mask, stick out my tongue and do all those experiments......

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
Pugsy
Posts: 65197
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: SH chart - what is happening here?

Post by Pugsy » Thu Mar 14, 2019 9:13 pm

katestyles wrote:
Thu Mar 14, 2019 9:08 pm
I can't ask questions - I have to go put on my mask, stick out my tongue and do all those experiments......
I can open my mouth...stick out my tongue...blow raspberries....cough....talk plainly...and breathe nice and evenly with my mask and machine on and zero air goes out through my open mouth.
It wasn't always that way....took months and months and then one night I tried to talk and I could without air tornadoing out my mouth.
I can't tell you how I do it but I am not the only one who can do it....lots of people can do it.
Open mouth doesn't necessarily mean massive leaks...sometimes doesn't mean any leaks at all.

_________________