New user - advice from cpap gods on sleepyhead data

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Tue Mar 12, 2019 12:33 am

Hello Everyone,

My main question is about a weird experience I had last night with my machine, but let me start with my basic info.

Age 35, male. AHI 19, oxy min 93, max 95. Machine settings: pressure min/max 4-15, erp 1, ramp on, comfort soft.

In my first three nights I saw steady progress with the final night peaking at 8 hours of use and very pretty ahi numbers. Then last night I ran into an issue where as I was dozing off the pressure would suddenly increase and I would be awoken. It happened two or three times and I got frustrated and quit. Was wondering if anyone could tell from my sleepyhead data if anything looks off?

Thanks in advance!

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
screenshot-20190312-143220.png
screenshot-20190312-143220.png (127.26 KiB) Viewed 1035 times

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: New user - advice from cpap gods on sleepyhead data

Post by D.H. » Tue Mar 12, 2019 10:01 am

This shows that you had a Central Apnea at 12:57, despite that you were not using the machine from about 12:41 to 1:06! I don’t know what to make of that.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New user - advice from cpap gods on sleepyhead data

Post by Pugsy » Tue Mar 12, 2019 10:18 am

D.H. wrote:
Tue Mar 12, 2019 10:01 am
This shows that you had a Central Apnea at 12:57, despite that you were not using the machine from about 12:41 to 1:06! I don’t know what to make of that.
The machine wasn't off during that time period between 12:41 to 1:06...it was simply at the minimum pressure of 4.
It was only off for maybe a minute at 12:41..very small break in therapy.
The pressure line was just stuck on 4 and didn't go up so you can't see it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Dreaming1
Posts: 166
Joined: Fri Aug 31, 2018 9:55 am

Re: New user - advice from cpap gods on sleepyhead data

Post by Dreaming1 » Tue Mar 12, 2019 10:43 am

What’s the lowest pressure you can handle awake? Your auto is taking a long time to get you up to 12. Perhaps you should start at a higher pressure than 4. Maybe consider 7 or 8 to start.

_________________
MachineMaskHumidifier
Additional Comments: Love my cpap life!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Tue Mar 12, 2019 12:53 pm

Pugsy wrote:
Tue Mar 12, 2019 10:18 am
D.H. wrote:
Tue Mar 12, 2019 10:01 am
This shows that you had a Central Apnea at 12:57, despite that you were not using the machine from about 12:41 to 1:06! I don’t know what to make of that.
The machine wasn't off during that time period between 12:41 to 1:06...it was simply at the minimum pressure of 4.
It was only off for maybe a minute at 12:41..very small break in therapy.
The pressure line was just stuck on 4 and didn't go up so you can't see it.
And this is why DH is best put on the foe list, and ignored.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Tue Mar 12, 2019 12:57 pm

yearning4zs wrote:
Tue Mar 12, 2019 12:33 am
Hello Everyone,

My main question is about a weird experience I had last night with my machine, but let me start with my basic info.

Age 35, male. AHI 19, oxy min 93, max 95. Machine settings: pressure min/max 4-15, erp 1, ramp on, comfort soft.

In my first three nights I saw steady progress with the final night peaking at 8 hours of use and very pretty ahi numbers. Then last night I ran into an issue where as I was dozing off the pressure would suddenly increase and I would be awoken. It happened two or three times and I got frustrated and quit. Was wondering if anyone could tell from my sleepyhead data if anything looks off?

Thanks in advance!
First, please fix your chart as shown in my sig.

Second, you're likely not being waked up by pressure, you're being waked up by the events you're having, because the starting pressure is way too low.

Bump the minimum up to 6 or 7 for starters.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Wed Mar 13, 2019 2:24 am

Thanks for the replies!

Yesterday I used the machine for an hour during the day to desensitize myself. It seems to have been helpful. Still had some wake-ups during the night, but I'm thinking that was more do to my girlfriend constantly coughing with the flu. I have included my data from last night below. Based on this data, does it seem like I would still need to bump up my starting pressure? I do like having it on 4 and basically feeling like there is no pressure at the start.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
screenshot-20190313-162137.png
screenshot-20190313-162137.png (128.63 KiB) Viewed 967 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Wed Mar 13, 2019 3:18 am

yearning4zs wrote:
Wed Mar 13, 2019 2:24 am
Thanks for the replies!

Yesterday I used the machine for an hour during the day to desensitize myself. It seems to have been helpful. Still had some wake-ups during the night, but I'm thinking that was more do to my girlfriend constantly coughing with the flu. I have included my data from last night below. Based on this data, does it seem like I would still need to bump up my starting pressure? I do like having it on 4 and basically feeling like there is no pressure at the start.
I'd suggest a min pressure of more like 9... if you want the machine to start at 4 and build up to the 9, then that's what ramp is for :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Wed Mar 13, 2019 3:19 am

Hello, sorry about that. I thought that the mask pressure and pressure graphs were interchangeable. Here is the updated version.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable
Attachments
screenshot-20190313-171739.png
screenshot-20190313-171739.png (132.22 KiB) Viewed 959 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Wed Mar 13, 2019 3:19 am

KnightSleeper wrote:
Wed Mar 13, 2019 2:59 am
yearning4zs wrote:
Wed Mar 13, 2019 2:24 am
Thanks for the replies!

Yesterday I used the machine for an hour during the day to desensitize myself. It seems to have been helpful. Still had some wake-ups during the night, but I'm thinking that was more do to my girlfriend constantly coughing with the flu. I have included my data from last night below. Based on this data, does it seem like I would still need to bump up my starting pressure? I do like having it on 4 and basically feeling like there is no pressure at the start.
yearning4zs-

Welcome to the forum! Just as a tip, you'll want to organize your Sleepyhead Charts like palerider mentioned above. Presently, you are not even showing your pressure chart, so there is absolutely no way the guys can give you advice. So, here's the link from palerider's signature...go to it and follow the instructions there on which charts to upload. Then they will be able to give you some good advice. - Good luck on getting more restful z's!!!

wiki/index.php/Sleepyhead:Organize
He's got 'mask pressure' instead of 'pressure'.. it's a little harder to read, but good enough. What's annoying is when they have both mask pressure AND pressure, since they're redundant, and just waste space. :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Wed Mar 13, 2019 3:21 am

yearning4zs wrote:
Wed Mar 13, 2019 3:19 am
Hello, sorry about that. I thought that the mask pressure and pressure graphs were interchangeable. Here is the updated version.
That's ok, ... the charts are pretty much interchangeable... though the pressure one is easier to read when looking at the whole night. Mask pressure really shines when you're zoomed in far enough to see individual breaths, then you can see how the pressure increases are timing themselves with the breaths. sometimes that matters, (but that's more often on fancier machines).

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Wed Mar 13, 2019 3:22 am

In response to the comment about ramp and starting at 7or 8, there is one thing I still don't entirely understand. With resmed autosense 10, lets say I set a minimum pressure of 7-8. It will immediately be on that pressure, or will it wait until I'm asleep and then ramp up to that pressure?

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New user - advice from cpap gods on sleepyhead data

Post by palerider » Wed Mar 13, 2019 3:25 am

yearning4zs wrote:
Wed Mar 13, 2019 3:22 am
In response to the comment about ramp and starting at 7or 8, there is one thing I still don't entirely understand. With resmed autosense 10, lets say I set a minimum pressure of 7-8. It will immediately be on that pressure, or will it wait until I'm asleep and then ramp up to that pressure?
Huh? I suggest a minimum pressure of 9.

It's the Resmed AirSense 10 AutoSet

You If you set the min pressure to 9, it'll start at that pressure.

If you want it to start at a lower pressure, then set a ramp start pressure of whatever (4?) and a ramp time, or use the autoramp feature, which starts ramping up when it thinks you're asleep.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Wed Mar 13, 2019 3:34 am

Ok, thanks for the advice. I'll slowly move my way toward a minimum of 9. Maybe start with 6 or 7 tonight.

_________________
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Additional Comments: humidifier is a Humidair Cleanable

yearning4zs
Posts: 49
Joined: Tue Mar 12, 2019 12:01 am

Re: New user - advice from cpap gods on sleepyhead data

Post by yearning4zs » Wed Mar 13, 2019 3:57 am

KnightSleeper wrote:
Wed Mar 13, 2019 3:45 am
yearning4zs wrote:
Wed Mar 13, 2019 3:34 am
Ok, thanks for the advice. I'll slowly move my way toward a minimum of 9. Maybe start with 6 or 7 tonight.
If you are concerned about your comfort at the higher levels of pressure, you could experiment with your EPR (Expiratory Pressure Relief) setting. Right now you have it set at 1cmw; If you move it to 2cm, it makes it easier/more comfortable on your exhalations. The pressure on your exhalations will be the "given pressure" less the EPR; therefore at a 9cm minimum pressure which is what palerider is recommending, you presently have a 1 EPR, so your exhalation pressure would be 9-1=8; if you set it to 2, you would have 9-2=7 on your exhalations. Some people go as high as 3. I'm still experimenting and have mine set to 2 right now.

I hope this helps. Good luck!
Thanks so much, I'll see how it goes with slowly cranking up the pressure and keep in mind the advice about ERP, thanks!