My Story and Advice Needed

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: My Story and Advice Needed

Post by palerider » Tue Mar 05, 2019 3:48 pm

HoseCrusher wrote:
Tue Mar 05, 2019 1:47 pm
The average of 90 - 93 is a little low so this is something that needs to be addressed.
No it doesn't. That's well within 'ok' for someone who's asleep.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

seeker_here_in
Posts: 24
Joined: Mon Mar 04, 2019 1:56 am

Re: My Story and Advice Needed

Post by seeker_here_in » Tue Mar 05, 2019 7:18 pm

I tried the suggested settings (Mode : S, IPAP : 18, EPAP 12) and the AHI has actually increased to 9.5.

Attaching the graph

6March.png
6March.png (182.91 KiB) Viewed 483 times

Happy to learn from your advice and work at this to get this down.

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My Story and Advice Needed

Post by Pugsy » Tue Mar 05, 2019 7:22 pm

I think I said 14 EPAP....not 12

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

seeker_here_in
Posts: 24
Joined: Mon Mar 04, 2019 1:56 am

Re: My Story and Advice Needed

Post by seeker_here_in » Tue Mar 05, 2019 7:26 pm

Pugsy wrote:
Tue Mar 05, 2019 7:22 pm
I think I said 14 EPAP....not 12
My mistake! Let me try that and get back with tomorrow's results.

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My Story and Advice Needed

Post by Pugsy » Tue Mar 05, 2019 7:28 pm

I am hoping that reducing that 6 cm difference down to 4 cm will also reduce those centrals....crossing my fingers here.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: My Story and Advice Needed

Post by raisedfist » Tue Mar 05, 2019 7:30 pm

I don't think he turned off AVAPS either while on S mode. When you look at his pressure line for the IPAP it is not static as it should be in S or S/T mode with AVAPS disabled. If it is off, you may need to lower the rise time setting by one notch because it's not able to generate the full prescribed pressure at the given flow rate of air. Respironics is dumb like that.

_________________
Mask
Last edited by raisedfist on Tue Mar 05, 2019 7:33 pm, edited 1 time in total.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

seeker_here_in
Posts: 24
Joined: Mon Mar 04, 2019 1:56 am

Re: My Story and Advice Needed

Post by seeker_here_in » Tue Mar 05, 2019 7:32 pm

raisedfist wrote:
Tue Mar 05, 2019 7:30 pm
I don't think he turned off AVAPS either while on S mode. When you look at his pressure line for the IPAP it is not static as it should be in S or S/T mode with AVAPS disabled.
No, AVAPS is switched off

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: My Story and Advice Needed

Post by raisedfist » Tue Mar 05, 2019 7:34 pm

seeker_here_in wrote:
Tue Mar 05, 2019 7:32 pm
raisedfist wrote:
Tue Mar 05, 2019 7:30 pm
I don't think he turned off AVAPS either while on S mode. When you look at his pressure line for the IPAP it is not static as it should be in S or S/T mode with AVAPS disabled.
No, AVAPS is switched off
Gotcha. Try lowering the rise time setting by 1 notch and see if that evens out your pressures. It shouldn't be off by 1.5cm at times, so it's probably a flow rate of air issue.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

seeker_here_in
Posts: 24
Joined: Mon Mar 04, 2019 1:56 am

Re: My Story and Advice Needed

Post by seeker_here_in » Tue Mar 05, 2019 7:37 pm

raisedfist wrote:
Tue Mar 05, 2019 7:34 pm
seeker_here_in wrote:
Tue Mar 05, 2019 7:32 pm
raisedfist wrote:
Tue Mar 05, 2019 7:30 pm
I don't think he turned off AVAPS either while on S mode. When you look at his pressure line for the IPAP it is not static as it should be in S or S/T mode with AVAPS disabled.
No, AVAPS is switched off
Gotcha. Try lowering the rise time setting by 1 notch and see if that evens out your pressures. It shouldn't be off by 1.5cm at times, so it's probably a flow rate of air issue.
Rise Time Lock is off, and Rise Time is 3. Do you suggest changing rise time to 2?

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My Story and Advice Needed

Post by Pugsy » Tue Mar 05, 2019 7:40 pm

Is Flex on by chance? If it is that would explain the little dips in pressure I think.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

seeker_here_in
Posts: 24
Joined: Mon Mar 04, 2019 1:56 am

Re: My Story and Advice Needed

Post by seeker_here_in » Tue Mar 05, 2019 7:43 pm

Pugsy wrote:
Tue Mar 05, 2019 7:40 pm
Is Flex on by chance? If it is that would explain the little dips in pressure I think.
Nope. Both Flex Lock and Bi-flex are off.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: My Story and Advice Needed

Post by raisedfist » Tue Mar 05, 2019 7:46 pm

seeker_here_in wrote:
Tue Mar 05, 2019 7:37 pm
raisedfist wrote:
Tue Mar 05, 2019 7:34 pm
seeker_here_in wrote:
Tue Mar 05, 2019 7:32 pm
raisedfist wrote:
Tue Mar 05, 2019 7:30 pm
I don't think he turned off AVAPS either while on S mode. When you look at his pressure line for the IPAP it is not static as it should be in S or S/T mode with AVAPS disabled.
No, AVAPS is switched off
Gotcha. Try lowering the rise time setting by 1 notch and see if that evens out your pressures. It shouldn't be off by 1.5cm at times, so it's probably a flow rate of air issue.
Rise Time Lock is off, and Rise Time is 3. Do you suggest changing rise time to 2?
You could try it, as long as it's comfortable. That worked for me I have the same machine. I'd increase the EPAP to reduce your pressure support like was suggested though.

I'm a bit confused because all the settings they set up point to them believing you have some type of lung or neuromuscular issue, but you say you don't. They went to the trouble of enabling AVAPS and prescribing an IPAP min and max. Very odd. I have to wonder how you'd do on an APAP.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Pugsy
Posts: 65109
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My Story and Advice Needed

Post by Pugsy » Tue Mar 05, 2019 7:47 pm

Hmmmm....
When I used the S/T I had in S mode my pressure lines were fixed...not bumpy...but then it was a ResMed S/T and not the Respironics.
I am scratching my head over the bumpiness.

PS....pressure support is the difference between inhale and exhale.
Let's see what happens tonight with EPAP 14 and IPAP 18 and that makes PS 4.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: My Story and Advice Needed

Post by raisedfist » Tue Mar 05, 2019 7:50 pm

Pugsy wrote:
Tue Mar 05, 2019 7:47 pm
Hmmmm....
When I used the S/T I had in S mode my pressure lines were fixed...not bumpy...but then it was a ResMed S/T and not the Respironics.
I am scratching my head over the bumpiness.

PS....pressure support is the difference between inhale and exhale.
Let's see what happens tonight with EPAP 14 and IPAP 18 and that makes PS 4.
It's a PR thing with the bumpy pressure line. It's only solved by decreasing the rise time, which increases the peak flow rate of air, or you can go really crazy and enable pressure control (pc) mode and set a fixed inspiratory time which forces you to inhale a certain amount of time each breath so usually it extends the inhale. I've had 2 PR machines do it and zero ResMed's.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: My Story and Advice Needed

Post by HoseCrusher » Tue Mar 05, 2019 9:22 pm

palerider wrote:
Tue Mar 05, 2019 3:48 pm
HoseCrusher wrote:
Tue Mar 05, 2019 1:47 pm
The average of 90 - 93 is a little low so this is something that needs to be addressed.
No it doesn't. That's well within 'ok' for someone who's asleep.
You are absolutely correct...

For someone at an elevation of 5000 feet an average oxygen level of 90% during sleep is perfectly normal...







At sea level 94% seems to be more therapeutic.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...