gone back to full mask

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

gone back to full mask

Post by jono_1uk » Fri Mar 01, 2019 12:03 am

so decided to try the full face mask again... here are last nights results..no leaks ..can someone tell me why my AHI is high?
am guessing air pressure needs to go up?
screenshot-20190301-060431.png
screenshot-20190301-060431.png (114.25 KiB) Viewed 6688 times

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: gone back to full mask

Post by zoocrewphoto » Fri Mar 01, 2019 12:12 am

Your pressure is too low. See how the pressure goes up and down? The minimum is too low to prevent a lot of your events. But the machine is programmed to wait until it detects flow limitations and snores. Then it goes up slowly. Once it does well, it goes back down again. Your machine is set too low, so it spends too much time going up to where it needs to be. It even stops at the top a few times, unable to go higher, but needing to. Your minimum is set at 7.5. I would bump it up to 10 and see how it does. It probably needs more, but sometimes a better start will keep it from needing to climb as much. I would also bump the max unless you are having trouble with higher pressures.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: gone back to full mask

Post by palerider » Fri Mar 01, 2019 2:47 am

jono_1uk wrote:
Fri Mar 01, 2019 12:03 am
.can someone tell me why my AHI is high?
am guessing air pressure needs to go up?
FFMs often push the lower jaw back, narrowing the airway, and thus needing more pressure to open things up.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: gone back to full mask

Post by LSAT » Fri Mar 01, 2019 6:09 am

Max pressure is too low also...set pressures at 10/20

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: gone back to full mask

Post by Janknitz » Fri Mar 01, 2019 2:57 pm

Your leaks are high too, need to get that under control as that will impair the effectiveness of the machine to manage your apneas.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Dog Slobber
Posts: 4327
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: gone back to full mask

Post by Dog Slobber » Fri Mar 01, 2019 3:06 pm

Janknitz wrote:
Fri Mar 01, 2019 2:57 pm
Your leaks are high too, need to get that under control as that will impair the effectiveness of the machine to manage your apneas.
Those Leak rate numbers are fine for Respironics.

Numbers like that on a ResMed and I'd agree.
Battery Backup: EcoFlow Delta 2

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: gone back to full mask

Post by jono_1uk » Sat Mar 02, 2019 3:11 am

ok so did 10/20 settings as suggested .. would like AHI to be a bit lower .. coped with pressure ok ..does it need to go up?
screenshot-20190302-090513.png
screenshot-20190302-090513.png (111.94 KiB) Viewed 6573 times

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: gone back to full mask

Post by zoocrewphoto » Sat Mar 02, 2019 3:18 am

It will probably need to be a minimum of 11 or 12. If you feel comfortable jumping to 12, go ahead and try it. But if not, just go up slowly and give it some time.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
Pugsy
Posts: 65325
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: gone back to full mask

Post by Pugsy » Sat Mar 02, 2019 5:50 am

I see at least 2 times where you woke up and turned the machine off and then back on again.
Were there more times where you woke up and didn't turn the machine off and then lay there for a period of time with mask and machine on?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: gone back to full mask

Post by jono_1uk » Sat Mar 02, 2019 5:56 am

Hi Pugsy

No they were all bathroom visits and machine was off and back on again straight way mask remained on and i used the quick release.

User avatar
Pugsy
Posts: 65325
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: gone back to full mask

Post by Pugsy » Sat Mar 02, 2019 6:29 am

Why 2 bathroom breaks? Are you male of an age where the prostate causes problems?

It is common for some people to need more pressure when they use a full face mask than when they use a nasal interface mask.
Sometimes a little more and sometimes substantially more minimum pressure when using auto adjusting mode.
I think it is because of what the lower straps, needed to secure the lower section of the mask for seal, does to the airway position.
They all seem to tuck the chin and jaw back a little (sometimes a lot) and this causes the airway to become less open and then when it does start to collapse once asleep it's just harder to keep the airway open so more pressure is needed.

If you want to continue to use a full face mask you will likely have to use a minimum pressure of around 11 or 12 (maybe a little more) to get the same good results you got with a lower minimum. You might cut some of that difference if you can use a cervical collar to help hold the jaw upwards a bit so that you can't tuck that jaw backwards...but then that's another thing to wear on the face.
You may or may not find it all that comfortable. There's always a trade off it seems in just about anything we do....we get to choose the option that is least annoying. :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: gone back to full mask

Post by jono_1uk » Sat Mar 02, 2019 7:54 am

Thanks Pugsy

Age wise am 53 ..prostate all good , not diabetic but keep needing a pee at night ..doctor investigating this issue put me on
to the sleep apnea think because it is known to trigger this.. my apnea was diagnosed a severe up to 70 incidents per hour.
hoping if i get my AHI down then the bathroom trips may stop ..

User avatar
Dog Slobber
Posts: 4327
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: gone back to full mask

Post by Dog Slobber » Sat Mar 02, 2019 8:02 am

jono_1uk wrote:
Sat Mar 02, 2019 7:54 am
Thanks Pugsy

Age wise am 53 ..prostate all good , not diabetic but keep needing a pee at night ..doctor investigating this issue put me on
to the sleep apnea think because it is known to trigger this.. my apnea was diagnosed a severe up to 70 incidents per hour.
hoping if i get my AHI down then the bathroom trips may stop ..
They did for me.

Prior to CPAP, and during the beginning of therapy I used to visit the bathroom a couple times a night.

Now that things have settled in, I can't remember my last nightly trip.
Battery Backup: EcoFlow Delta 2

User avatar
Pugsy
Posts: 65325
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: gone back to full mask

Post by Pugsy » Sat Mar 02, 2019 8:06 am

Yes...for some people once the sleep apnea issue is better managed the nocturia issues will also resolve.
So keep trying...but also bear in mind that sleep apnea isn't the only cause for nocturia and also there's always the possibility that people have more than one problem at the same time.

It appears you are going to have a substantial difference in pressure needs when using a full face mask. Some people just draw the short straw in that regard. How much of a difference is unknown and at the end of the day you need what you need and you have to weigh the pros and cons with whatever you are wanting to do.

My nocturia totally went away once I got my therapy dialed in optimally. That alone was reason enough for me to continue with cpap therapy even if nothing else improved.
But I am female and had no other potential medical issues to blame the nocturia on.....so mine was fairly simple.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: gone back to full mask

Post by jono_1uk » Sun Mar 03, 2019 1:35 am

well ..last night seemed really good .. i put the low pressure up again slightly ..only one bathroom visit ... and another stoppage to get to an itch under my mask ..
screenshot-20190303-072949.png
screenshot-20190303-072949.png (111.92 KiB) Viewed 6504 times