Are my leaks causing any harm?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Duroth
Posts: 2
Joined: Sat Feb 02, 2019 2:12 am

Are my leaks causing any harm?

Post by Duroth » Sat Feb 02, 2019 2:38 am

Hey all :D

I received my diagnostic in August 2018, but that was just a formality, i had apnea since my early '20, i just didn't bother to treat it (life endangering stupid, i know). The sleep test returned with horrific results, AHI 99.6, medium SPO2 87% and min SPO2 63%. After a month or so of accommodating with different masks, my choice was the Dreamwear. Now the thing is i have full beard. My guess is that the choice in masks alleviates the beard problem a bit, but i don't know to what extent. I'll attach the SleepyHead screenshot of my last night sleep. I know 5h is a bit short, but that's how long i usually sleep, maybe due to having bad sleep nights for such a long time.
Attachments
screenshot-20190202-101556.png
screenshot-20190202-101556.png (75.94 KiB) Viewed 171 times
Always tired bearded bald fatso

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Are my leaks causing any harm?

Post by Jas_williams » Sat Feb 02, 2019 9:01 am

Your leaks are fine


You are using a Philips machine and the large leak setting in sleepy head is set at 24L/M this is a Resmed thing. You can go into preferences and turn it off as you have a Philips machine or raise it to a more sensible value of 75L/m

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
Pugsy
Posts: 65122
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Are my leaks causing any harm?

Post by Pugsy » Sat Feb 02, 2019 9:09 am

Unless any minor excess leaks are disturbing your sleep....you aren't having any leaks anywhere near large enough to impact the therapy effectiveness.

See the LL line on the events graph....that's large leak and you aren't having any and even if you did the machine can compensate for quite a bit. The machine's ability to compensate really isn't impacted until you reach over 100 L/min for total leak (that top leak line on the graph).

Actually a very decent leak line....the minor variations are mainly just associated with pressure increases. As the pressure increases you can expect that top leak line to move upwards....more pressure means more vent rate/intentional leak and it's to be expected.
That top leak line on the graph is the mask's vent rate plus any excess leak.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Duroth
Posts: 2
Joined: Sat Feb 02, 2019 2:12 am

Re: Are my leaks causing any harm?

Post by Duroth » Sat Feb 02, 2019 10:03 am

Thank you for the wonderful and reassuring responses.
Always tired bearded bald fatso