I think it’s a DSX500T11C. See attached pic
Water runs out
Re: Water runs out
I think it’s a DSX500T11C. See attached pic
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
Re: Water runs out
The model DSX500 is the Auto CPAP model and it is indeed compatible with SleepyHead.
We have people posting screen shots from that model all over here.
Now why your Mac had a problem...I have no idea but SH does work with your machine and there is data on that SD card.
Irrelevant though for the water consumption issue....it would have taken massive large leaks to eat up the water and most assuredly would have been flagged by the machine as large leak and would show up on DreamMapper as well.
We have people posting screen shots from that model all over here.
Now why your Mac had a problem...I have no idea but SH does work with your machine and there is data on that SD card.
Irrelevant though for the water consumption issue....it would have taken massive large leaks to eat up the water and most assuredly would have been flagged by the machine as large leak and would show up on DreamMapper as well.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
dreamingofdreaming
- Posts: 198
- Joined: Wed Apr 18, 2018 12:07 pm
- Location: Chicago-ish
Re: Water runs out
I woke up this morning with no water left in the tank. DreamMapper says 0 large leaks, but SleepyHead says there were leaks.
Re: Water runs out
I did get SleepyHead to work. I had to insert/eject my SD card a few times until the Mac recognized it. Probably dust in the slot since I rarely use the SD Drive.
Sleepy head stated that I head periods of significant mask leaks. Can those be when I got up to go pee and left the machine running? If you'd like me to zip up and email my data I can do that too. Just PM me.
I attached a screenshot of the SleepyHead report. If someone could take a look at it for I'd be grateful. What are vibratory snores? Are they bad?
Sleepy head stated that I head periods of significant mask leaks. Can those be when I got up to go pee and left the machine running? If you'd like me to zip up and email my data I can do that too. Just PM me.
I attached a screenshot of the SleepyHead report. If someone could take a look at it for I'd be grateful. What are vibratory snores? Are they bad?
- Attachments
-
- Screen Shot 2019-01-21 at 4.14.14 PM.png (94.34 KiB) Viewed 1955 times
Last edited by higgledy on Mon Jan 21, 2019 3:19 pm, edited 2 times in total.
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
Re: Water runs out
Same here. Do you leave your machine running, as I do, when you get up to pee in the night?dreamingofdreaming wrote: ↑Mon Jan 21, 2019 3:11 pmI woke up this morning with no water left in the tank. DreamMapper says 0 large leaks, but SleepyHead says there were leaks.
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
-
dreamingofdreaming
- Posts: 198
- Joined: Wed Apr 18, 2018 12:07 pm
- Location: Chicago-ish
Re: Water runs out
I usually sleep all night and don't have to get up and disconnect. I had nasal surgery a month ago and am not able to tighten my mask as much as I need to, to prevent my mask from moving around, causing it to leak. Pugsy will be able to help you with your problem.
Re: Water runs out
You Respironics users with the machine saying no leaks and SleepyHead saying there were large leaks...
Your machine trumps SH.
The problem is SleepyHead defaults to the red line for ResMed machines...and that's where that % of time over red line statistic comes from...not from your machine.
ResMed machines report leaks different...excess leak only and the red line is 24 L/min.
Respironics doesn't have a fixed red line and if they did with would be a lot higher than 24 L/min.
Change your red line threshold from the default 24 L/min to something more suitable for a Respironics machine...since it varies so much with pressures and masks used...impossible to give you an exact fixed number but something around 70 L/min would be a nice conservative number.
Go to Preferences...CPAP tab...and change the 24 red line threshold to 70 L/min or just turn it off and rely on the machine to flag any large leaks in the Events graph....see the LL line on the Events graph??? That's large leak...no flags there means no large leaks despite what SH says in the statistics column for % of time over red line.
Your machine trumps SH.
The problem is SleepyHead defaults to the red line for ResMed machines...and that's where that % of time over red line statistic comes from...not from your machine.
ResMed machines report leaks different...excess leak only and the red line is 24 L/min.
Respironics doesn't have a fixed red line and if they did with would be a lot higher than 24 L/min.
Change your red line threshold from the default 24 L/min to something more suitable for a Respironics machine...since it varies so much with pressures and masks used...impossible to give you an exact fixed number but something around 70 L/min would be a nice conservative number.
Go to Preferences...CPAP tab...and change the 24 red line threshold to 70 L/min or just turn it off and rely on the machine to flag any large leaks in the Events graph....see the LL line on the Events graph??? That's large leak...no flags there means no large leaks despite what SH says in the statistics column for % of time over red line.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Water runs out
Thanks. I made this change in settings.Pugsy wrote: ↑Mon Jan 21, 2019 3:22 pmYou Respironics users with the machine saying no leaks and SleepyHead saying there were large leaks...
Your machine trumps SH.
The problem is SleepyHead defaults to the red line for ResMed machines...and that's where that % of time over red line statistic comes from...not from your machine.
ResMed machines report leaks different...excess leak only and the red line is 24 L/min.
Respironics doesn't have a fixed red line and if they did with would be a lot higher than 24 L/min.
Change your red line threshold from the default 24 L/min to something more suitable for a Respironics machine...since it varies so much with pressures and masks used...impossible to give you an exact fixed number but something around 70 L/min would be a nice conservative number.
Go to Preferences...CPAP tab...and change the 24 red line threshold to 70 L/min or just turn it off and rely on the machine to flag any large leaks in the Events graph....see the LL line on the Events graph??? That's large leak...no flags there means no large leaks despite what SH says in the statistics column for % of time over red line.
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
Re: Water runs out
I never ran out of water with my old Respironics machine, always had plenty left.
Fom Day 1 of using the DreamStation I am down to only a few drops of water remaining in the tank each morning.
Nothing else changed except the machine.
Can't help with a solution - but would like one. Just wanted to commiserate!
Fom Day 1 of using the DreamStation I am down to only a few drops of water remaining in the tank each morning.
Nothing else changed except the machine.
Can't help with a solution - but would like one. Just wanted to commiserate!
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: Water runs out
I followed my medical supply company’s recommendations to use the non-heated tube and put the machine humidifier into non-adaptive mode. I still wake up in the middle of the night with painfully dry nostrils, mouth, and throat. I am really beginning to hate my new Respironics Dreamstation. So much so, that I’m returning the entire unit to my medical supply company. How can they sell a $1000 piece of junk like this and sleep at night? The company rep. even admitted to me over the phone that they know there is a design flaw with the humidifier i.e. running out of water. So F them and F Respironics. I’m done.
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
Re: Water runs out
For what it's worth....with my Dreamstation....I use a heated tube, set to 2 and my humidifier is set to 4. It's only run out of water once in over a year. Maybe I'm lucky? And I live in the desert in Wyoming, no humidifier in the house.
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Re: Water runs out
I noticed that someone said that other people are not having your problem. Please notice the thread called New Machine. I am having the same problem. I received a new machine a week ago. I never had a problem with my old Respironics machine. My setting is on a 2. On a 1, I get no moisture, and on a 3, I get about 4 hours sleep. Sorry you are having problems.
Re: Water runs out
Last night I had the humidifier set at 3 and it ran dry. Plus, I have a room humidifier set to 45.
ResMed AirFit Mask
Respironics DreamStation
Respironics DreamStation
Re: Water runs out
Hi. I'm new here. Just got my first cpap 2 weeks ago. I changed to a PR Dream station a week ago. I've tried the humidifier at various levels all week. I didn't change the water and didn't top it up all week. There's still water in there. I thought I had a problem because it wasn't using water.
Im guessing your problem is that it is not adapting to the environment. Surely it should not be set to be nonadaptive.
Im guessing your problem is that it is not adapting to the environment. Surely it should not be set to be nonadaptive.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Wisp Nasal CPAP Mask with Headgear - Fit Pack |
| Humidifier: DreamStation Heated Humidifier |
We're all out of cornflakes. FU.
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Water runs out
Four things determine the amount of water you use:
Relative humidity of your bedroom
Temperature of the air going through the machine
Humidity level you choose
Your treatment pressure
Relative humidity of your bedroom
Temperature of the air going through the machine
Humidity level you choose
Your treatment pressure
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.