The 50D+ does not use a USB cable, it uses a cable that *looks* like USB, but has active circuitry in the large end.metsfan302 wrote: ↑Tue Feb 05, 2019 7:14 amugg, Seems my CMS-50D won’t work / transfer data without using the orig cable it came with. I tried using a diff mini usb to usb and my comp did not see the pulse ox. Going to have to dig deeper or buy another cable.
Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Metsfan.... I know this is still an ongoing battle with you, but if you get to the point where the old machine is redundant, I might be interested in it if the price is reasonable. I have the S9 ASV. The model I have does not have auto ASV (if that is right term). I think the 10's all do. I've always wanted to try it.
My machine is 6.5 years old but still works fine....
My machine is 6.5 years old but still works fine....
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Tom.... I was def thinking of putting it up in a post for listing. With em being so expensive, I am just unsure what to price it at to be completely honest - Was letting the wife use it, but all's she needs is a auto cpap so i guess I could sell the ASV to get her an Auto Cpap..... I will let you know / give you first right of refusal once I do list it for sale.
Here is a refurb one I found - its $1,699
https://www.1800cpap.com/aircurve-10-as ... med-1.aspx
Here is a refurb one I found - its $1,699
https://www.1800cpap.com/aircurve-10-as ... med-1.aspx
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
Last edited by metsfan302 on Wed Feb 06, 2019 5:45 am, edited 2 times in total.
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Well last night was def Better! Going to leave it on this setting for now to see how it continues
Epap - 13
Min PS - 3.0
Max PS - 12.0 (Highest it would let me set it too)
Epap highest it would go adjust wise to in the ASV Mode is still 15
This first screen is my longest stretch of sleep from like midnight till 4ish - Taking just that section and not the entire time period
My AHI shows as only 3.54
The few hr stretch toward the end of the night

The Entire "day"

Epap - 13
Min PS - 3.0
Max PS - 12.0 (Highest it would let me set it too)
Epap highest it would go adjust wise to in the ASV Mode is still 15
This first screen is my longest stretch of sleep from like midnight till 4ish - Taking just that section and not the entire time period
My AHI shows as only 3.54
The few hr stretch toward the end of the night

The Entire "day"

_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
My sleep dr appt got moved up to tomorrow! Guess we will go over the reports, talk about schedule sleep study and then my upcoming ENT appt.
Hopefully will get this all figured out.
Does a deviated septum effect apnea? How is that surgery?
Hopefully will get this all figured out.
Does a deviated septum effect apnea? How is that surgery?
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
I'd try even less MinPS...
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
When I had my septoplasty several years ago, it wasn't bad at all. They used the splints instead of packing my nose with gause. I was actually breathing better with the splints in place than I did before the surgery. Once they took the splints out, I almost hyperventolated. The first time I had ever been able to get that much nose through my nose at one time. I don't know if it had much effect on my apnea, but it sure made it easier to breathe.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Saw Sleep Doc Today, She was pissy I was changing the pressures but anywho -- She bumped up the Epap to 14, leaving it on ASV mode. returning in 2 weeks, review reports then either ASV Tritration to see if other settings would be more optimal.
She did mention if ever needed more Epap it would be thru the Trilogy Ventilator but its super expensive, but insurance would have to allow if its medically needed. If I fail on the Trilogy then yea its over lol...... I guess the Trilogy is more for well my pressures needs ? and end stage COPD - I do have COPD but not end stage. Just found out today there are stages lol.
https://www.usa.philips.com/healthcare/ ... ifications
Some of its Specs / Settings
Specifications
IPAP - 50 cm H₂O
CPAP - 4 - 20 cmH20 (passive leak port circuits) cm H₂O
EPAP/PEEP (a) - 0 - 25 cmH20 (active valve circuits) cm H₂O
EPAP/PEEP (b) - 4 - 25 cmH20 (passive leak port circuits) cm H₂O
Pressure support - 0 – 30 cmH2O cm H₂O
Tidal volume - 50 - 2000 ml
Breath rate(a) - 0 - 60 BPM (AC mode) beats per minute
Breath rate(b) - 1 - 60 BPM (all other modes) beats per minute
Inspiratory time - 0.3 - 5.0 -s Synchrony features
Rise time - 1 - 6 (relative scale)
Ramp start pressure (a) - 0 - 25 cmH20 (active circuits) cm H₂O
Ramp start pressure (b) - 4 - 25 cmH20 (passive circuits) cm H₂O
Ramp start pressure (c) - 4 - 19 cmH20 (CPAP mode) cm H₂O
Ramp time - 5 - 45 min
C-Flex - 1 - 3 (relative scale)
Flow trigger sensitivity - 1 - 9 l/min
She did mention if ever needed more Epap it would be thru the Trilogy Ventilator but its super expensive, but insurance would have to allow if its medically needed. If I fail on the Trilogy then yea its over lol...... I guess the Trilogy is more for well my pressures needs ? and end stage COPD - I do have COPD but not end stage. Just found out today there are stages lol.
https://www.usa.philips.com/healthcare/ ... ifications
Some of its Specs / Settings
Specifications
IPAP - 50 cm H₂O
CPAP - 4 - 20 cmH20 (passive leak port circuits) cm H₂O
EPAP/PEEP (a) - 0 - 25 cmH20 (active valve circuits) cm H₂O
EPAP/PEEP (b) - 4 - 25 cmH20 (passive leak port circuits) cm H₂O
Pressure support - 0 – 30 cmH2O cm H₂O
Tidal volume - 50 - 2000 ml
Breath rate(a) - 0 - 60 BPM (AC mode) beats per minute
Breath rate(b) - 1 - 60 BPM (all other modes) beats per minute
Inspiratory time - 0.3 - 5.0 -s Synchrony features
Rise time - 1 - 6 (relative scale)
Ramp start pressure (a) - 0 - 25 cmH20 (active circuits) cm H₂O
Ramp start pressure (b) - 4 - 25 cmH20 (passive circuits) cm H₂O
Ramp start pressure (c) - 4 - 19 cmH20 (CPAP mode) cm H₂O
Ramp time - 5 - 45 min
C-Flex - 1 - 3 (relative scale)
Flow trigger sensitivity - 1 - 9 l/min
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Well not much change as far as AHI goes with the Adjustment she made to the EPAP in making it 14.
So the Settings are
ASV Mode - non-Auto
EPAP - 14
Min PS - 3.0
MAx PS - 11.0
Think I am getting mask leaks now all of a sudden showing large leak events
She wants me to not touch the settings and see her again in like 10 days.... Ugg -- She said if this did not work she would send me in for the ASV Tritration. Not sure how that is going to help but maybe it will show us some helpful data, who knows. This is getting so frustrating.
I do have an appt with a really good ENT on the 13th - maybe she will be able to see whats up as far as the "flow limitation/ obstruction etc" While I have used the P-10 Pillow Mask for a few years now, I do think I have air flow issues with my nose. I think when I exhale it only feels like air is coming out of one nostril. Maybe its the hole in my septum or i dunno - the sleep doc seems to think that it does not matter I have that as the ASV should take care of that?./ Unsure if switching back to a full face mask I have in the bin would be helpful in all this or if masks dont matter.
I did request a referral to a local pulmonologist, my family dr sent it over along with my records. I am thinking the pulmonologist might be able to see also whats up with my inhale/exhale issues, and also since he I think also deals with sleep medical I will have a second person over seeing my ASV / reports etc. (Might be why my current non-pulmonologist sleep doc has not recommended me going to a pulmonologist).
An Overview of Last Night

My AHI Shoots up fast as I sleep?

Large Leaks and event summary

So the Settings are
ASV Mode - non-Auto
EPAP - 14
Min PS - 3.0
MAx PS - 11.0
Think I am getting mask leaks now all of a sudden showing large leak events
She wants me to not touch the settings and see her again in like 10 days.... Ugg -- She said if this did not work she would send me in for the ASV Tritration. Not sure how that is going to help but maybe it will show us some helpful data, who knows. This is getting so frustrating.
I do have an appt with a really good ENT on the 13th - maybe she will be able to see whats up as far as the "flow limitation/ obstruction etc" While I have used the P-10 Pillow Mask for a few years now, I do think I have air flow issues with my nose. I think when I exhale it only feels like air is coming out of one nostril. Maybe its the hole in my septum or i dunno - the sleep doc seems to think that it does not matter I have that as the ASV should take care of that?./ Unsure if switching back to a full face mask I have in the bin would be helpful in all this or if masks dont matter.
I did request a referral to a local pulmonologist, my family dr sent it over along with my records. I am thinking the pulmonologist might be able to see also whats up with my inhale/exhale issues, and also since he I think also deals with sleep medical I will have a second person over seeing my ASV / reports etc. (Might be why my current non-pulmonologist sleep doc has not recommended me going to a pulmonologist).
An Overview of Last Night

My AHI Shoots up fast as I sleep?

Large Leaks and event summary

_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
- raisedfist
- Posts: 1176
- Joined: Wed Jun 15, 2016 7:21 am
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Have you tried using a full face mask and seeing what happens to those UA events? If you have one laying around, maybe worth a shot.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
My wife just reminded me that back in October when I was in the hospital for 2 nights after my Robotic Assisted Vertical Sleeve Gastrectomy, They required me to use my ASV and they ran Oxygen thru it.
I just looked at my sleep reports from those two night, man they are my lowest AHI's Ever! The 2nd night is really fragmented as they kept waking me every hour damn near for one reason or another - taking blood, checking whatever etc
I hope Oxygen is not the missing link with my Therapy......
If different screens are helpful, let me know!
Night 1

Night 2

I just looked at my sleep reports from those two night, man they are my lowest AHI's Ever! The 2nd night is really fragmented as they kept waking me every hour damn near for one reason or another - taking blood, checking whatever etc
I hope Oxygen is not the missing link with my Therapy......
If different screens are helpful, let me know!
Night 1

Night 2

_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
I think more advanced units were mentioned in this thread.metsfan302 wrote: ↑Thu Feb 07, 2019 2:42 pmSaw Sleep Doc Today, She was pissy I was changing the pressures but anywho -- She bumped up the Epap to 14, leaving it on ASV mode. returning in 2 weeks, review reports then either ASV Tritration to see if other settings would be more optimal.
She did mention if ever needed more Epap it would be thru the Trilogy Ventilator but its super expensive, but insurance would have to allow if its medically needed. If I fail on the Trilogy then yea its over lol...... I guess the Trilogy is more for well my pressures needs ? and end stage COPD - I do have COPD but not end stage. Just found out today there are stages lol.
https://www.usa.philips.com/healthcare/ ... ifications
Some of its Specs / Settings
Specifications
IPAP - 50 cm H₂O
CPAP - 4 - 20 cmH20 (passive leak port circuits) cm H₂O
EPAP/PEEP (a) - 0 - 25 cmH20 (active valve circuits) cm H₂O
EPAP/PEEP (b) - 4 - 25 cmH20 (passive leak port circuits) cm H₂O
Pressure support - 0 – 30 cmH2O cm H₂O
Tidal volume - 50 - 2000 ml
Breath rate(a) - 0 - 60 BPM (AC mode) beats per minute
Breath rate(b) - 1 - 60 BPM (all other modes) beats per minute
Inspiratory time - 0.3 - 5.0 -s Synchrony features
Rise time - 1 - 6 (relative scale)
Ramp start pressure (a) - 0 - 25 cmH20 (active circuits) cm H₂O
Ramp start pressure (b) - 4 - 25 cmH20 (passive circuits) cm H₂O
Ramp start pressure (c) - 4 - 19 cmH20 (CPAP mode) cm H₂O
Ramp time - 5 - 45 min
C-Flex - 1 - 3 (relative scale)
Flow trigger sensitivity - 1 - 9 l/min
_________________
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: S9 ST-A iVAPS and adapt ASV |
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Why?metsfan302 wrote: ↑Mon Feb 11, 2019 5:50 am
I hope Oxygen is not the missing link with my Therapy......
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- metsfan302
- Posts: 687
- Joined: Fri Oct 09, 2015 9:44 pm
- Location: Dewey Beach, Delaware
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
Just don't know if it would be covered by insurance etc..... kind of weird needing it at only 42 years old
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Prior - ResMed AirCurve ASV User |
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a
Re: Central & Obstructive - AHI always around 10 with ASV. Tweak Settings?>?
O2 is covered, but It would have shown up in your previous titrations, if you needed O2. There are too many variables on the hospital admission to draw conclusions. They may have just got your mask fit right is one thing.metsfan302 wrote: ↑Mon Feb 11, 2019 1:47 pmJust don't know if it would be covered by insurance etc..... kind of weird needing it at only 42 years old
_________________
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: S9 ST-A iVAPS and adapt ASV |
Last edited by ajack on Mon Feb 11, 2019 3:19 pm, edited 1 time in total.