Problems with airflow for Resmed P10

Post by Mr Nosey » Sat Dec 29, 2018 9:01 am

I have successfully used CPAP for 15 years. I currently own an Auto Intellipap and a Resmed Airsense 10 and a couple of pillow masks that I alternate if having any nose soreness. My "go to" masks include the Aloha, the Resmed Swift FX, Pilaro Q (not my favorite), Swift LT and now the p10. The Aloha, Swift LT and p10 are my favorites. The p10 particularly impressive as it has less leak issues than the others (frequently no leaks/night).

Using AUTO CPAP machines, I find that my pressure is usually 9-10cm/night. I have set up my machines so that my minimum pressure is 8cm, high pressure =20cm with a 5 minute ramp. Although the p10 has worked great for a few months, I now find that I'm having problems getting enough air, especially noticeable after stopping CPAP briefly for a nighttime bathroom break. I'm able to resume CPAP without problems using one of my other masks. However, the p10 is problematic and no machine-dependent (have the same problem with the Airsense and the Intellipap).

I'd like to continue to include the p10 as one of my treatment options. The fit is good and there are little/no leaks. Maybe the air diffuser has gotten clogged (though it does behave the same re the amount of air expelled from the front of the mask). Turning off the ramp so that the machine starts at 8cm has not helped. I just don't get enough air to comfortably fall back asleep..and need to switch masks. The odd thing is that I can usually (but not always) start off with the p10 when I first go to bed.

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine
Last edited by Mr Nosey on Wed Jan 02, 2019 5:16 am, edited 1 time in total.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed p20

Post by palerider » Sat Dec 29, 2018 3:01 pm

Mr Nosey wrote:
Sat Dec 29, 2018 9:01 am
Maybe the air diffuser has gotten clogged (though it does behave the same re the amount of air expelled from the front of the mask).
Give it a good scrubbing with some hot water, dishwashing soap, and a toothbrush.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Mr Nosey
Posts: 44
Joined: Sat Dec 29, 2018 8:35 am

Re: Problems with airflow for Resmed p10

Post by Mr Nosey » Mon Dec 31, 2018 7:54 am

Palerider,

MUCH thanks for your suggestion. I scrubbed the diffuser with a toothbrush, soap and warm water and set it aside for one day to dry. Tried the mask last night..didn't have any issues. I was not struggling to get air. Seems like the problem is solved and will post again soon if I still have the issue.

I did think the problem strange as I'd used the mask for 4 months without a problem. Was quite useful to know the problem was mask-specific and not machine-specific. One thing I should note is that one difference the past month (when the problem first arose) was the use of a new, more efficient humidifier in the bedroom (we now keep the humidity at 55%). Regardless of my hose and machine humidifier settings, I had found some water in the tube that on occasion was such that I could hear a gurgling noise coming from the hose due to water accumulation. Maybe there was some residue build up in the diffuser from so much water (though I do use distilled water)? Overall, I don't know the issue but have a tentative solution..and hope this post useful for anyone else who might some day encounter the issue. Really sucks to put on a mask that does not provide enough air and leaves you gasping for breath!
Best,
Richard

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine
Last edited by Mr Nosey on Wed Jan 02, 2019 5:16 am, edited 1 time in total.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed p20

Post by palerider » Mon Dec 31, 2018 2:42 pm

Mr Nosey wrote:
Mon Dec 31, 2018 7:54 am
Palerider,

MUCH thanks for your suggestion. I scrubbed the diffuser with a toothbrush, soap and warm water and set it aside for one day to dry. Tried the mask last night..didn't have any issues. I was not struggling to get air. Seems like the problem is solved and will post again soon if I still have the issue.

I did think the problem strange as I'd used the mask for 4 months without a problem. Was quite useful to know the problem was mask-specific and not machine-specific. One thing I should note is that one difference the past month (when the problem first arose) was the use of a new, more efficient humidifier in the bedroom (we now keep the humidity at 55%). Regardless of my hose and machine humidifier settings, I had found some water in the tube that on occasion was such that I could hear a gurgling noise coming from the hose due to water accumulation. Maybe there was some residue build up in the diffuser from so much water (though I do use distilled water)? Overall, I don't know the issue but have a tentative solution..and hope this post useful for anyone else who might some day encounter the issue. Really sucks to put on a mask that does not provide enough air and leaves you gasping for breath!
Best,
Richard
Make it a regular habit to give it a scrub every week or so, and you shouldn't have problems...

I think most of the crud that piles up there comes out of us, not the water. Whether you use distilled water or not, all that comes off the humidifier is pure water vapor. Basically, the first step in condensation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Mr Nosey
Posts: 44
Joined: Sat Dec 29, 2018 8:35 am

Re: Problems with airflow for Resmed p10

Post by Mr Nosey » Mon Dec 31, 2018 10:07 pm

palerider wrote:
Mon Dec 31, 2018 2:42 pm
Mr Nosey wrote:
Mon Dec 31, 2018 7:54 am
Palerider,

MUCH thanks for your suggestion. I scrubbed the diffuser with a toothbrush, soap and warm water and set it aside for one day to dry. Tried the mask last night..didn't have any issues. I was not struggling to get air. Seems like the problem is solved and will post again soon if I still have the issue.

I did think the problem strange as I'd used the mask for 4 months without a problem. Was quite useful to know the problem was mask-specific and not machine-specific. One thing I should note is that one difference the past month (when the problem first arose) was the use of a new, more efficient humidifier in the bedroom (we now keep the humidity at 55%). Regardless of my hose and machine humidifier settings, I had found some water in the tube that on occasion was such that I could hear a gurgling noise coming from the hose due to water accumulation. Maybe there was some residue build up in the diffuser from so much water (though I do use distilled water)? Overall, I don't know the issue but have a tentative solution..and hope this post useful for anyone else who might some day encounter the issue. Really sucks to put on a mask that does not provide enough air and leaves you gasping for breath!
Best,
Richard
Make it a regular habit to give it a scrub every week or so, and you shouldn't have problems...

I think most of the crud that piles up there comes out of us, not the water. Whether you use distilled water or not, all that comes off the humidifier is pure water vapor. Basically, the first step in condensation.
I do reasonably frequently soak my masks in water containing an antimicrobial agent but had never scrubbed the P10 with a brush. Go figure. Somewhat counterintuitive that a clogged diffuser would have such an impact. The good news is that the P10 seals better than any of my other pillow masks. Great to have it back in rotation.

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine
Last edited by Mr Nosey on Wed Jan 02, 2019 5:22 am, edited 1 time in total.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed p20

Post by palerider » Mon Dec 31, 2018 10:23 pm

Mr Nosey wrote:
Mon Dec 31, 2018 10:07 pm
I do reasonably frequently soak my masks in water containing an antimicrobial agent but had never scrubbed the P10 with a brush. Go figure.
Scrubbing, .... necessary, antimicrobial soak? not necessary. It's only your own germs that's on there, if it makes you feel better, go ahead, but soap and water is all that's needed.

Just like the recommendation has swung against having triclosan in everything, and going back to soap and water, or plain toothpaste.
Mr Nosey wrote:
Mon Dec 31, 2018 10:07 pm
Somewhat counterintuitive that a clogged diffuser would have such an impact.
Well, you were having a buildup of co2 because the clogged vent wasn't venting enough to flush it all out.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Mr Nosey
Posts: 44
Joined: Sat Dec 29, 2018 8:35 am

Re: Problems with airflow for Resmed p10

Post by Mr Nosey » Tue Jan 01, 2019 3:21 pm

palerider wrote:
Mon Dec 31, 2018 10:23 pm
Mr Nosey wrote:
Mon Dec 31, 2018 10:07 pm
I do reasonably frequently soak my masks in water containing an antimicrobial agent but had never scrubbed the P10 with a brush. Go figure.
Scrubbing, .... necessary, antimicrobial soak? not necessary. It's only your own germs that's on there, if it makes you feel better, go ahead, but soap and water is all that's needed.

Just like the recommendation has swung against having triclosan in everything, and going back to soap and water, or plain toothpaste.
Mr Nosey wrote:
Mon Dec 31, 2018 10:07 pm
Somewhat counterintuitive that a clogged diffuser would have such an impact.
Well, you were having a buildup of co2 because the clogged vent wasn't venting enough to flush it all out.
Well, as I have a strong backgroung in microbiology and virology, Im most comfortable and sticking to occasional anti-microbial soaks.

Re the diffuser and your comment about CO2 build up, does Resmed highlight this as a potential problem with the P20 mask (maybe a trouble shooting section in the literature provided with the mask). Will admit that I no longer read those booklets..
Cheers and happy New Year.

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine
Last edited by Mr Nosey on Wed Jan 02, 2019 5:22 am, edited 1 time in total.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed p20

Post by palerider » Tue Jan 01, 2019 5:26 pm

Mr Nosey wrote:
Tue Jan 01, 2019 3:21 pm
Resmed highlight this as a potential problem with the P20 mask
You'll do better looking for the proper name... P10

cleaning the mesh is mentioned as part of the normal cleaning:
https://www.youtube.com/watch?v=MylY6a6KXMY

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Mr Nosey
Posts: 44
Joined: Sat Dec 29, 2018 8:35 am

Re: Problems with airflow for Resmed p10

Post by Mr Nosey » Wed Jan 02, 2019 5:21 am

palerider wrote:
Tue Jan 01, 2019 5:26 pm
Mr Nosey wrote:
Tue Jan 01, 2019 3:21 pm
Resmed highlight this as a potential problem with the P20 mask
You'll do better looking for the proper name... P10

cleaning the mesh is mentioned as part of the normal cleaning:
https://www.youtube.com/watch?v=MylY6a6KXMY
Thanks for correcting me re the mask being P10, not P20...had it correct in the body of the post but not the subject title (which I've corrected so it might hopefully come up in searches).

The mesh cleaning mention in a Youtube video is a joke..is it standard to use Youtube for official product info, especially given the issue? Resmed should explicitly state the issue so it's prominent and a known, potentially serious issue.

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Problems with airflow for Resmed p10

Post by Dog Slobber » Wed Jan 02, 2019 9:22 am

Mr Nosey wrote:
Wed Jan 02, 2019 5:21 am
is it standard to use Youtube for official product info, especially given the issue?
Yes it is. A lot of companies use YouTube for official, promotional and information videos.
Battery Backup: EcoFlow Delta 2

User avatar
Mr Nosey
Posts: 44
Joined: Sat Dec 29, 2018 8:35 am

Re: Problems with airflow for Resmed p10

Post by Mr Nosey » Wed Jan 02, 2019 9:40 am

Dog Slobber wrote:
Wed Jan 02, 2019 9:22 am
Mr Nosey wrote:
Wed Jan 02, 2019 5:21 am
is it standard to use Youtube for official product info, especially given the issue?
Yes it is. A lot of companies use YouTube for official, promotional and information videos.
I am not questioning whether companies use Youtube for promotion or providing information. The question is whether a company (in this case, Resmed) would use it as the SOLE method of providing information. I strongly assume you agree that not everyone is internet savvy andbofvthose that are, only a subset views Youtube?

On additional bit of information. I checked with my supplies provider. The P10 mask does not come with printed information highlighting the potential for there being clogged mask filter issues (though note there is an insert that mentions the importance of cleaning their machine filters). I am only suggesting something proactive on Resmed's part (provided with their product) that undoubtedly may have value to some.

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Also use Aloha, Nuance, Swift LT and Swift FX nasal pillow masks. Intellipap = backup machine

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: Problems with airflow for Resmed P10

Post by jnk... » Wed Jan 02, 2019 10:09 am

All it takes is a little condensation in the mask for the venting to become worthless for that night. So I agree that there is a real problem for some when they can't figure out why they suddenly seem starved for air in the middle of the night and aren't knowledgeable enough to recognize it as CO2 buildup caused by water in that mask.

I agree that any mask in which simple moisture renders the vent completely ineffective is a mask that needs a bold warning somewhere for legal and ethical reasons.

I do not recommend that anyone else do what I do to that mask, but I actually punch extra holes in the vents, myself. It makes the mask much louder and increases the intentional leak beyond what the mask was designed to provide, but waking up from CO2 buildup is not a pleasant experience for me, and I don't always get my humidification perfect for preventing condensation in the mask.
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed p10

Post by palerider » Wed Jan 02, 2019 12:44 pm

Mr Nosey wrote:
Wed Jan 02, 2019 5:21 am
The mesh cleaning mention in a Youtube video is a joke..is it standard to use Youtube for official product info, especially given the issue? Resmed should explicitly state the issue so it's prominent and a known, potentially serious issue.
You know that manual, the one you
Mr Nosey wrote:
Tue Jan 01, 2019 3:21 pm
Resmed highlight this as a potential problem with the P20 mask (maybe a trouble shooting section in the literature provided with the mask). Will admit that I no longer read those booklets..
Didn't bother to read? It's in there.

What do you want them to do, put a plastic flag on the mask itself? like you find on electrical cords?

When was the last time you read one of those? The last Christmas lights I bought had *THREE* flags... I didn't read those either.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

realshelby
Posts: 337
Joined: Thu Jul 26, 2018 9:36 am

Re: Problems with airflow for Resmed P10

Post by realshelby » Wed Jan 02, 2019 1:25 pm

I have a P10. What I don't understand is the problem with airflow? Block the vent....and you get the same or higher actual airflow. This isn't a Carbon Dioxide issue, putting it on after a bathroom break won't allow enough build up to matter for a while at least. I too seem to need to take a few deep breaths after getting up for a bathroom break in the middle of the night to get back to a normal breath intake. But that applies to all the masks I use. I have been using a Nuance Pro lately and it has a very open vent....same breathing with it.

I am thinking this may be more of a nasal passage restriction of some sort that is beginning to effect airflow?
Min 7 Max 20 (cmH2O)

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Problems with airflow for Resmed P10

Post by palerider » Wed Jan 02, 2019 2:15 pm

realshelby wrote:
Wed Jan 02, 2019 1:25 pm
I have a P10. What I don't understand is the problem with airflow? Block the vent....and you get the same or higher actual airflow. This isn't a Carbon Dioxide issue, putting it on after a bathroom break won't allow enough build up to matter for a while at least. I too seem to need to take a few deep breaths after getting up for a bathroom break in the middle of the night to get back to a normal breath intake. But that applies to all the masks I use. I have been using a Nuance Pro lately and it has a very open vent....same breathing with it.

I am thinking this may be more of a nasal passage restriction of some sort that is beginning to effect airflow?
Your thinking is flawed.

Block the vent, and you get *no* airflow (unless you've got gills or something else weird that gives the air sowhere to go).

Block the vent on any mask and co2 will immediately start to build up.