Trouble with Mouth Leaks

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Trouble with Mouth Leaks

Post by BretH » Fri Dec 28, 2018 7:09 pm

I think I have tried everything to stop mouth leaks without any luck.
I have tried tape: Drool ruins the adhesive and the tape falls off.
Full face Dreamwear mask: Drool causes loss of seal and the mask starts making strange sounds that wake me up. I also tried a mask liner hoping it would absorb the drool and stop the leaks. Did not help.
Chin straps: Tried several. They seem to help a little with leaks but I sleep worse and am more tired the next day even with less leakage.
I tried a neck collar. This helped stop the leaks some nights and didn't do anything other nights. It was hit or miss and not very comfortable.
I'm pretty sure its mouth leaks because I wake up with a very dry mouth several times a night.
When I use the full face mask, my AHI goes way up. With all the leaks and only the nasal mask, my AHI is very low. I am wondering if the low AHI is real or if it just can't measure the apnea's because of all the leaks?
I have attached 2 SleepyHead reports.

This one dated 12/27 is typical with the nasal mask.
screenshot-20181228-170210.png
Nasal Dreamwear Mask
screenshot-20181228-170210.png (319.18 KiB) Viewed 2123 times
This one from 12/2 was with the full face mask.
screenshot-20181228-170149.png
Full Face Dreamwear Mask
screenshot-20181228-170149.png (309.84 KiB) Viewed 2123 times
I sleep with a night guard and I think this may cause me to drool more than most people.

Any help would be greatly appreciated!

User avatar
zonker
Posts: 11358
Joined: Fri Jun 19, 2015 4:36 pm

Re: Trouble with Mouth Leaks

Post by zonker » Fri Dec 28, 2018 7:39 pm

BretH wrote:
Fri Dec 28, 2018 7:09 pm
I think I have tried everything to stop mouth leaks without any luck.
<snip>
Any help would be greatly appreciated!
may i suggest this?

viewtopic/p1089718/viewtopic.php?f=1&t= ... s#p1086296

worked like a champ for me. maybe it could work for you?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Trouble with Mouth Leaks

Post by djams » Fri Dec 28, 2018 7:47 pm

When switching masks , are you changing the mask type in the clinical menu?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
accrn
Posts: 76
Joined: Sat Dec 15, 2018 11:03 pm

Re: Trouble with Mouth Leaks

Post by accrn » Fri Dec 28, 2018 8:15 pm

djams wrote:
Fri Dec 28, 2018 7:47 pm
When switching masks , are you changing the mask type in the clinical menu?
Is that universal for all machines, or just the one the OP is using? After reading your comment I went and looked through my menu (because I have changed masks multiple times and never done this) and I have a little mask symbol with numbers and I can't figure out what they mean... have downloaded the user guide and I still can't figure out how the numbers correspond to mask type. I'm sorry I'm so clueless.

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Alternate mask is Resmed P10 nasal pillows

BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Re: Trouble with Mouth Leaks

Post by BretH » Fri Dec 28, 2018 8:32 pm

djams wrote:
Fri Dec 28, 2018 7:47 pm
When switching masks , are you changing the mask type in the clinical menu?
Yes, I changed the mask type.

BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Re: Trouble with Mouth Leaks

Post by BretH » Fri Dec 28, 2018 8:37 pm

zonker wrote:
Fri Dec 28, 2018 7:39 pm
BretH wrote:
Fri Dec 28, 2018 7:09 pm
I think I have tried everything to stop mouth leaks without any luck.
<snip>
Any help would be greatly appreciated!
may i suggest this?

viewtopic/p1089718/viewtopic.php?f=1&t= ... s#p1086296

worked like a champ for me. maybe it could work for you?
Thanks, just ordered some. They will come Sunday and I'll give it a try.

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Trouble with Mouth Leaks

Post by djams » Fri Dec 28, 2018 8:58 pm

accrn wrote:
Fri Dec 28, 2018 8:15 pm
djams wrote:
Fri Dec 28, 2018 7:47 pm
When switching masks , are you changing the mask type in the clinical menu?
Is that universal for all machines, or just the one the OP is using? After reading your comment I went and looked through my menu (because I have changed masks multiple times and never done this) and I have a little mask symbol with numbers and I can't figure out what they mean... have downloaded the user guide and I still can't figure out how the numbers correspond to mask type. I'm sorry I'm so clueless.
I don't know the answer to this. Is your manual the user guide or clinical guide? I'd think the clinical guide would tell what the numbers mean.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

User avatar
zonker
Posts: 11358
Joined: Fri Jun 19, 2015 4:36 pm

Re: Trouble with Mouth Leaks

Post by zonker » Fri Dec 28, 2018 9:04 pm

BretH wrote:
Fri Dec 28, 2018 8:37 pm

Thanks, just ordered some. They will come Sunday and I'll give it a try.
oh! most excellent. please let me know if it works for you. i've done a few "mods" on what jay 1st proposed there.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
accrn
Posts: 76
Joined: Sat Dec 15, 2018 11:03 pm

Re: Trouble with Mouth Leaks

Post by accrn » Fri Dec 28, 2018 9:20 pm

djams wrote:
Fri Dec 28, 2018 8:58 pm
accrn wrote:
Fri Dec 28, 2018 8:15 pm
djams wrote:
Fri Dec 28, 2018 7:47 pm
When switching masks , are you changing the mask type in the clinical menu?
Is that universal for all machines, or just the one the OP is using? After reading your comment I went and looked through my menu (because I have changed masks multiple times and never done this) and I have a little mask symbol with numbers and I can't figure out what they mean... have downloaded the user guide and I still can't figure out how the numbers correspond to mask type. I'm sorry I'm so clueless.
I don't know the answer to this. Is your manual the user guide or clinical guide? I'd think the clinical guide would tell what the numbers mean.
I found a thread that addresses this. Apparently it really doesn't matter too much on my machine since I'm using a ResMed mask.
viewtopic/t113696/Dreamstation--Mask-Type-Setting.html

_________________
Machine: DreamStation BiPAP® Auto Machine
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Alternate mask is Resmed P10 nasal pillows

BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Re: Trouble with Mouth Leaks

Post by BretH » Tue Jan 01, 2019 4:40 pm

zonker wrote:
Fri Dec 28, 2018 9:04 pm
BretH wrote:
Fri Dec 28, 2018 8:37 pm

Thanks, just ordered some. They will come Sunday and I'll give it a try.
oh! most excellent. please let me know if it works for you. i've done a few "mods" on what jay 1st proposed there.

good luck!
So I have been working on some sort of mouth leak stopper device for a while. I was close but couldn’t find a good way to keep it in the right place until you suggested the Scunci. Then, everything kind of fell into place.

First, I did a 3D scan of my face with my mask on. I also taped an object of known size to my forehead so I could scale the scan to the correct size. Then I eliminated all the area where I didn’t want the mask and converted the scan into a 5mm wide mask that covers and fits perfectly around my mouth area.
ScanFaceAndMask.jpg
ScanFaceAndMask.jpg (82.17 KiB) Viewed 2014 times
I then printed the mask on my 3D printer in nylon material so it would be smooth and somewhat soft. The mask goes slightly under my chin to help keep my mouth closed. I cut a hole in the scunci and inserted the mask. I decided not to use the earholes since I woke up at 2:30AM with sore ears. I think it works better without the earholes. I placed the strap just above my ears. I just put the top of the scunci more towards the top of my head. That way it pulls more up and is better able to keep my jaw from opening. I may add some type of strap to keep the top of the scunci in place if needed.
Mask.jpg
Mask.jpg (290.17 KiB) Viewed 2014 times
I have used it for only one night but it looks good so far. This is my report form last night. No major leaks, yeah!
screenshot-20190101-065257.png
screenshot-20190101-065257.png (304.47 KiB) Viewed 2014 times

User avatar
zonker
Posts: 11358
Joined: Fri Jun 19, 2015 4:36 pm

Re: Trouble with Mouth Leaks

Post by zonker » Tue Jan 01, 2019 5:30 pm

BretH wrote:
Tue Jan 01, 2019 4:40 pm
<snip>
So I have been working on some sort of mouth leak stopper device for a while. I was close but couldn’t find a good way to keep it in the right place until you suggested the Scunci. Then, everything kind of fell into place.
<snip>
holeee catttss! that is astonishing!! i can't believe the amount of work you put into that.

have you considered doing this as a business?

seriously, i think you are on to something here.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Re: Trouble with Mouth Leaks

Post by BretH » Tue Jan 01, 2019 5:48 pm

zonker wrote:
Tue Jan 01, 2019 5:30 pm
BretH wrote:
Tue Jan 01, 2019 4:40 pm
<snip>
So I have been working on some sort of mouth leak stopper device for a while. I was close but couldn’t find a good way to keep it in the right place until you suggested the Scunci. Then, everything kind of fell into place.
<snip>
holeee catttss! that is astonishing!! i can't believe the amount of work you put into that.

have you considered doing this as a business?

seriously, i think you are on to something here.
The process is actually not too much work once figured out. I do a lot of 3D printing anyway. It would be fun to do for others, but, scanning people all over the US would be difficult. I suppose there might be quite a few people in the LA area.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Trouble with Mouth Leaks

Post by Jay Aitchsee » Tue Jan 01, 2019 6:33 pm

Perfect!!
Great Job!


For those of us without printers, do you think there is a material that could be cut to size and molded to the face? I'm thinking of something like a sheet of silicone type material like that used for boil and bite mouth guards.


And speaking of 3D printers, there was an outfit in Florida that purported to make cpap face masks from scans. IIRC they charged quite a bit. A Google search might turn them up.


Perhaps you should talk with Karen from Pad a Cheek. Maybe you could partner with her to make your device marketable.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
djams
Posts: 927
Joined: Mon Jun 18, 2018 4:37 pm
Location: DFW area

Re: Trouble with Mouth Leaks

Post by djams » Tue Jan 01, 2019 6:40 pm

Jay Aitchsee wrote:
Tue Jan 01, 2019 6:33 pm
Perhaps you should talk with Karen from Pad a Cheek. Maybe you could partner with her to make your device marketable.
Or at least research a patent on "custom fitted cpap masks by 3d scanning". So when the big players start offering this, BrettH gets paid. :)

Very impressed. Nicely done BrettH!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins

BretH
Posts: 8
Joined: Thu May 24, 2018 5:49 pm

Re: Trouble with Mouth Leaks

Post by BretH » Wed Jan 02, 2019 12:28 pm

Jay Aitchsee wrote:
Tue Jan 01, 2019 6:33 pm
Perfect!!

For those of us without printers, do you think there is a material that could be cut to size and molded to the face? I'm thinking of something like a sheet of silicone type material like that used for boil and bite mouth guards.
You could definitely do this with molds, but that would require quite a bit of work and knowledge of materials. If you have ever seen Face-off on the Syfy channel, they would be able to create a silicone prosthetic but would need to create a positive first and then mold the mask on the positive face and then create a negative mold of both. Not too practical.
All the heat moldable material I’m aware of would be too hard to shape around soft face tissue, but I’m not really a materials expert. You could just use a soft silicone material or gel and let it conform. Maybe a thin dense memory foam would worK? It wouldn’t apply much force to keep the face in a specific location but it would conform and seal the leaks.