S9 and Airsense 10 --How does humidifier work?
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
S9 and Airsense 10 --How does humidifier work?
I have both machines and I am wondering if either humidifier is working. I've always kept humidity setting low. Recently I turned the heat setting up to 86 degrees and in the morning when I turned the machine off the metal plate under the humifier was completely cold. It happened with both machines. Is that right? I realized I didn't know how it was supposed to be working.
Similarly, how can you tell if the Climateline hose is working? Does that always give an error message if it is faulty?
Thanks,
E25
Similarly, how can you tell if the Climateline hose is working? Does that always give an error message if it is faulty?
Thanks,
E25
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
Re: S9 and Airsense 10 --How does humidifier work?
Hose air temperature isn’t governed by the heater plate at all.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
freetimecreations
- Posts: 145
- Joined: Thu May 24, 2012 9:02 pm
- Location: Utah, USA
Re: S9 and Airsense 10 --How does humidifier work?
I've sometimes doubted that my climate hose was working. To test it, I changed the hose temperature to maximum, on my machine it's 86. With a machine that had been off for 30 minutes I could feel the tube heat up in my hand. If you can't tell put the hose in a colder place than your bedroom for 30 minutes and then try again.elizabeth25 wrote: ↑Sun Dec 23, 2018 11:35 amSimilarly, how can you tell if the Climateline hose is working? Does that always give an error message if it is faulty?
_________________
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Additional Comments: Pressure 8.2cm-14.6cm, EPR 1, Humidity Auto, Contec CMS50F |
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
Re: S9 and Airsense 10 --How does humidifier work?
If I have humidity set to 1 and temp turned up to 86 degrees should the heater plate be warm?Pugsy wrote: ↑Sun Dec 23, 2018 12:04 pmHose air temperature isn’t governed by the heater plate at all.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
Re: S9 and Airsense 10 --How does humidifier work?
The 'temp' is the hose *ONLY*, it has absolutely nothing at all to do with the heater plate.elizabeth25 wrote: ↑Sun Dec 23, 2018 5:14 pmIf I have humidity set to 1 and temp turned up to 86 degrees should the heater plate be warm?Pugsy wrote: ↑Sun Dec 23, 2018 12:04 pmHose air temperature isn’t governed by the heater plate at all.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
All that matters is the '1' for humidity, and that's low, so it only heats up the water a little, as needed.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
Re: S9 and Airsense 10 --How does humidifier work?
Oh! So if I turn the humidifier up to maximum temp the metal plate would be warm?palerider wrote: ↑Sun Dec 23, 2018 5:49 pmThe 'temp' is the hose *ONLY*, it has absolutely nothing at all to do with the heater plate.elizabeth25 wrote: ↑Sun Dec 23, 2018 5:14 pmIf I have humidity set to 1 and temp turned up to 86 degrees should the heater plate be warm?Pugsy wrote: ↑Sun Dec 23, 2018 12:04 pmHose air temperature isn’t governed by the heater plate at all.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
All that matters is the '1' for humidity, and that's low, so it only heats up the water a little, as needed.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
- Okie bipap
- Posts: 3567
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: S9 and Airsense 10 --How does humidifier work?
I am not familiar with the S9, but I have the Air Curve 10 which has a preheat option. Of your machine has this option, put a small amount of water in the water chamber, start the preheat option, then after 10 minutes see if the metal plate is warm. This will confirm if it is working or not.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Evora Full Face Mask - Fitpack |
| Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: S9 and Airsense 10 --How does humidifier work?
Maybe, it would depend on the humidity in your room.elizabeth25 wrote: ↑Sun Dec 23, 2018 7:52 pmOh! So if I turn the humidifier up to maximum temp the metal plate would be warm?palerider wrote: ↑Sun Dec 23, 2018 5:49 pmThe 'temp' is the hose *ONLY*, it has absolutely nothing at all to do with the heater plate.elizabeth25 wrote: ↑Sun Dec 23, 2018 5:14 pmIf I have humidity set to 1 and temp turned up to 86 degrees should the heater plate be warm?Pugsy wrote: ↑Sun Dec 23, 2018 12:04 pmHose air temperature isn’t governed by the heater plate at all.
Two totally separate functions. Hose air temp for the air in the hose and the heater plate under the water chamber for regulating humidity per whatever you have selected combined with the ambient humidity in the bedroom.
All that matters is the '1' for humidity, and that's low, so it only heats up the water a little, as needed.
Using the pre-heat function would be the best test.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
Re: S9 and Airsense 10 --How does humidifier work?
I'm in Seattle where we have no lack of humidity during the winter
I'll try the preheat tonight and see what happens.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
Re: S9 and Airsense 10 --How does humidifier work?
I've been playing with the S9 and I can't for the life of me find where you can do the Warm Up.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
Re: S9 and Airsense 10 --How does humidifier work?
With the S9 you have to hold down a button (or maybe 2 buttons) to start the preheat option but it’s been so long Since I used the S9 that I forget which buttons and how long.
It’s all explained in the clinical/provider manual though.
Head over to apneaboard.com to the page that lets you request the manual via email and get the manual and read it.
Sorry don’t have the link handy on the iPad I am using now.
It’s all explained in the clinical/provider manual though.
Head over to apneaboard.com to the page that lets you request the manual via email and get the manual and read it.
Sorry don’t have the link handy on the iPad I am using now.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: S9 and Airsense 10 --How does humidifier work?
It's explained on the display... actually, when you go to the humidifier setting, there's a graphic at the bottom of the display... hold down the knob.Pugsy wrote: ↑Sun Dec 23, 2018 11:14 pmWith the S9 you have to hold down a button (or maybe 2 buttons) to start the preheat option but it’s been so long Since I used the S9 that I forget which buttons and how long.
It’s all explained in the clinical/provider manual though.
Head over to apneaboard.com to the page that lets you request the manual via email and get the manual and read it.
Sorry don’t have the link handy on the iPad I am using now.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: S9 and Airsense 10 --How does humidifier work?
What is causing you to be concerned about the plate warming up?elizabeth25 wrote: ↑Sun Dec 23, 2018 11:07 pmI've been playing with the S9 and I can't for the life of me find where you can do the Warm Up.
As far as I can tell by the *VERY* infrequent reported failures of Resmed humidifiers, they do a self test diagnostic, and report errors.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
elizabeth25
- Posts: 44
- Joined: Wed Jan 28, 2015 6:47 pm
Re: S9 and Airsense 10 --How does humidifier work?
What is causing you to be concerned about the plate warming up?
As far as I can tell by the *VERY* infrequent reported failures of Resmed humidifiers, they do a self test diagnostic, and report errors.
[/quote]
My S9 totally conked out last summer after (only) four years. The DME where I got the machine gave me an old S9 to use until I can replace it. My insurance changes to Medicare in January and I'm hoping to start the process of getting a new machine. I wanted to understand if the humidifier is working so that I can accurately describe what shape the machine is in if I need to.
I got the Warm Up to work last night.
As far as I can tell by the *VERY* infrequent reported failures of Resmed humidifiers, they do a self test diagnostic, and report errors.
[/quote]
My S9 totally conked out last summer after (only) four years. The DME where I got the machine gave me an old S9 to use until I can replace it. My insurance changes to Medicare in January and I'm hoping to start the process of getting a new machine. I wanted to understand if the humidifier is working so that I can accurately describe what shape the machine is in if I need to.
I got the Warm Up to work last night.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Thank goodness for Sleepyhead. |
Elizabeth25
- ChicagoGranny
- Posts: 15513
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: S9 and Airsense 10 --How does humidifier work?
All you will need for Medicare coverage of a new CPAP machine is a statement from your doctor that you have had a sleep study and you have been compliant with CPAP. Your old machine can still be functioning perfectly and Medicare will cover a new one.elizabeth25 wrote: ↑Mon Dec 24, 2018 11:52 amMy insurance changes to Medicare in January and I'm hoping to start the process of getting a new machine. I wanted to understand if the humidifier is working so that I can accurately describe what shape the machine is in if I need to.
Here's a tip about humidifiers. Compare the level of water in the morning to the level that was there when you went to bed.