I have noticed something strange lately
I have noticed something strange lately
Or I should say it seems strange to me, and that is why I am posing a question to those with far more experience than I. Starting about six weeks ago by looking at my SH charts, by AHI and pressures have changed for certain. I know that a few nights do not make a trend, but this is weeks and weeks now.
Keeping this as brief as possible, my AHI was typically split about 50/50 between a mix of obstructive apneas/hypopneas and centrals. I keep my machine in APAP mode with settings of 7/12, EPR of 2, and humidifier at 4 or 5. Typically my pressures will bump up to the nine plus range at least two times a night to kill an OA, and often a tad higher.
Looking at these recent charts though, my AHI consists of virtually ALL central events. I'm not bringing this up because I'm worried about centrals, just saying that my OAs have all disappeared, and I would have a 0.00 AHI otherwise. My median pressures are running low sevens consistently, and showing a max less than eight now with very little variation during the night.
I guess I'm confused as to the distinct change, and have even wondered if something is wrong with my machine in terms of its data reporting. Could my mild/moderate OSA have changed somehow, or is this a normal occurrence that you folks see often in your own charts?
I'm considering taking my machine over to AC Biomed to be tested since I have a location close by, or I am considering switching to straight CPAP with a pressure of 4 for a few nights to see if any OAs appear in the data again.
Any thoughts on this are appreciated as usual!
Here is last nights chart which is typical of late,............
Keeping this as brief as possible, my AHI was typically split about 50/50 between a mix of obstructive apneas/hypopneas and centrals. I keep my machine in APAP mode with settings of 7/12, EPR of 2, and humidifier at 4 or 5. Typically my pressures will bump up to the nine plus range at least two times a night to kill an OA, and often a tad higher.
Looking at these recent charts though, my AHI consists of virtually ALL central events. I'm not bringing this up because I'm worried about centrals, just saying that my OAs have all disappeared, and I would have a 0.00 AHI otherwise. My median pressures are running low sevens consistently, and showing a max less than eight now with very little variation during the night.
I guess I'm confused as to the distinct change, and have even wondered if something is wrong with my machine in terms of its data reporting. Could my mild/moderate OSA have changed somehow, or is this a normal occurrence that you folks see often in your own charts?
I'm considering taking my machine over to AC Biomed to be tested since I have a location close by, or I am considering switching to straight CPAP with a pressure of 4 for a few nights to see if any OAs appear in the data again.
Any thoughts on this are appreciated as usual!
Here is last nights chart which is typical of late,............
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead / ResScan / AirStart 10 Backup / Min6-Max12 APAP Mode, EPR 2 |
Re: I have noticed something strange lately
Why not first try changing the minimum to 4 and see what happens?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: I have noticed something strange lately
Maybe your machine is having issues try unplugging it and plug it back in
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
-
- Posts: 548
- Joined: Thu Nov 05, 2015 5:43 am
Re: I have noticed something strange lately
Hi Mark,
An OT question: do you generally sleep for only just over 4 hours per night, or was last night a bit unusual?
An OT question: do you generally sleep for only just over 4 hours per night, or was last night a bit unusual?
Re: I have noticed something strange lately
People have reported issues with sh not reporting flags properly with the airsense 10 machines in rare occasions. Zoom in in the flow rate till you can see individual breaths, and hen loon for any 10+ second pauses that aren't flagged.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: I have noticed something strange lately
I just scanned through the FR for all of last night, and saw nothing abnormal short of the sections where the centrals occur. Also, my AHI that is being shown on the Autoset is virtually the same as SH,....give or take a tenth or so, and sometimes right on.
I already set my machine like Pugsy mentioned. Just left it in APAP with a 4 minimum, and turned off the ramp. It will be interesting to see what that report has to say. If it's clear of any OAs,....I am either cured, or something is wrong with my machine.
Arlene,.....I typically sleep anywhere from 5.5 to 7 hours, with 6 being my average during the work week. Last night I was propped up in bed watching some old sitcom DVDs before I masked up and went to sleep. Needless to say I dropped off, and woke up later with no mask!

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead / ResScan / AirStart 10 Backup / Min6-Max12 APAP Mode, EPR 2 |
Re: I have noticed something strange lately
Other option...if it never moves off 4 cm...maybe 4 cm is enough to hold the airway open.
I know people coming out of a titration sleep study with a RX for 4 cm..because that's all they needed but without the 4 cm...definitely not cured.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: I have noticed something strange lately
I was using the word 'cured' kind of sarcastically.
I know I'm not cured since nothing has changed in my life from meds to weight, or stress.
I would be surprised to see a flat 4cm all night considering that I had a titration study that resulted in my initial scrip for a constant 8cm with an EPR of 3. I'm really more of the thinking that something might be wrong with my Autoset, although I can't sense anything not working properly, or differently.
It's simply that my OAs disappeared, and my pressure in APAP is not even wandering a full point of pressure during the night anymore.

I would be surprised to see a flat 4cm all night considering that I had a titration study that resulted in my initial scrip for a constant 8cm with an EPR of 3. I'm really more of the thinking that something might be wrong with my Autoset, although I can't sense anything not working properly, or differently.
It's simply that my OAs disappeared, and my pressure in APAP is not even wandering a full point of pressure during the night anymore.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead / ResScan / AirStart 10 Backup / Min6-Max12 APAP Mode, EPR 2 |
Re: I have noticed something strange lately
Any chance that there is a strong positional component to your OSA and you are just avoiding that position or at least keeping it to a minimum?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: I have noticed something strange lately
I am worse on my back, although that is how I sleep most of the night due to back/neck pain (which I know you are very familiar with also).
I have been spending more time side sleeping, but that has only been for the past week or so. The changes in my SH reports started months ago. It's really a bit of a mystery at this point.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead / ResScan / AirStart 10 Backup / Min6-Max12 APAP Mode, EPR 2 |
Re: I have noticed something strange lately
Aliens then.
They do some weird crap.


They do some weird crap.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: I have noticed something strange lately
Well, not necessarily.
Even at 4, you're still getting some assistance holding your airway open.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: I have noticed something strange lately
I'm on a week+ string of AHI=0. It's not impossible,
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.
Never, never, never, never say never.
Re: I have noticed something strange lately
Any chance you've been working on keeping a seal between your tongue and the roof of your mouth and you've finally mastered it? That made my AHI drop a little even when back sleeping. If you happen to change positions now and then to help with the pain you mentioned, maybe you hold your breath (like me) or the machine is just doesn't know it's awake breathing.
_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions |
My SleepDancing Video link https://www.youtube.com/watch?v=jE7WA_5c73c
Re: I have noticed something strange lately
Well, here is last nights chart. I really feel like something weird is going on now. Again,...no OAs, a median pressure of 4.46 and a 95% pressure of 5.46 when it took 8cm during my titration to stop them. This makes no sense to me, and I could use any suggestions available to try and figure out what is going on.
I know that OSA just doesn't disappear for no reason, so I am only joking when I suggest I am 'cured' per se. I honestly am baffled by whatever has happened, and am finding it hard to trust the data from my machine at this point.
I know that OSA just doesn't disappear for no reason, so I am only joking when I suggest I am 'cured' per se. I honestly am baffled by whatever has happened, and am finding it hard to trust the data from my machine at this point.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
Additional Comments: SleepyHead / ResScan / AirStart 10 Backup / Min6-Max12 APAP Mode, EPR 2 |