Question on airsense 10 side cover

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mets123
Posts: 367
Joined: Thu Feb 08, 2018 3:19 pm

Question on airsense 10 side cover

Post by mets123 » Thu Dec 13, 2018 1:06 pm

Just a fyi question,I put the airsense 10 side cover on and when I look at my options it comes up with tube temperature as an option to adjust. Am I wrong but how can it have a temperature control to adjust if there is no humidifier attached that has water in it to heat up the hot plate?
Thanks

_________________
MachineMask

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question on airsense 10 side cover

Post by Pugsy » Thu Dec 13, 2018 1:11 pm

Because you can warm the air without using the humidifier at all.
Totally separate functions.
Heated hose just warms the air in the hose.
Humidifier heating plate just warms the water in the water chamber.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mets123
Posts: 367
Joined: Thu Feb 08, 2018 3:19 pm

Re: Question on airsense 10 side cover

Post by mets123 » Thu Dec 13, 2018 1:16 pm

Thanks for that really fast reply Pugsy. That makes sense but is that air warming only with the heated hose? If I were to use a slimline hose do I also have heated air by adjusting the temperature? Crazy question since a heated hose can heat the air but a slimline hose?

_________________
MachineMask

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question on airsense 10 side cover

Post by Pugsy » Thu Dec 13, 2018 1:20 pm

You won't/can't get a hose air temp choice available without the heated hose being attached.
Just won't happen. You can't warm the air in the hose if the hose doesn't have the little heater coils inside the hose.
If you stick a regular slimline hose on the blower that doesn't have electrical plug in stuff...you won't get a choice to adjust the temperature of the hose.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

mets123
Posts: 367
Joined: Thu Feb 08, 2018 3:19 pm

Re: Question on airsense 10 side cover

Post by mets123 » Thu Dec 13, 2018 1:23 pm

Makes sense Pugsy, Thanks AGAIN!

_________________
MachineMask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question on airsense 10 side cover

Post by palerider » Thu Dec 13, 2018 2:43 pm

mets123 wrote:
Thu Dec 13, 2018 1:06 pm
Just a fyi question,I put the airsense 10 side cover on and when I look at my options it comes up with tube temperature as an option to adjust. Am I wrong but how can it have a temperature control
Because tube temperature is not the same thing as humidification level.

Take off the heated tube and put on a plain hose, and that option disappears.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mets123
Posts: 367
Joined: Thu Feb 08, 2018 3:19 pm

Re: Question on airsense 10 side cover

Post by mets123 » Thu Dec 13, 2018 3:16 pm

Just tried the slimline hose and Palerider you are correct. Thanks

_________________
MachineMask

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question on airsense 10 side cover

Post by palerider » Thu Dec 13, 2018 4:51 pm

mets123 wrote:
Thu Dec 13, 2018 3:16 pm
Just tried the slimline hose and Palerider you are correct. Thanks
When not using the heated hose, you have to make sure you set the proper tube type (size) slimline or regular, or your pressures will be very off.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mets123
Posts: 367
Joined: Thu Feb 08, 2018 3:19 pm

Re: Question on airsense 10 side cover

Post by mets123 » Thu Dec 13, 2018 4:54 pm

Thanks for reminding me.

_________________
MachineMask

BrandonB
Posts: 79
Joined: Tue Oct 18, 2011 3:41 pm

Re: Question on airsense 10 side cover

Post by BrandonB » Thu Dec 13, 2018 11:28 pm

palerider wrote:
Thu Dec 13, 2018 4:51 pm
When not using the heated hose, you have to make sure you set the proper tube type (size) slimline or regular, or your pressures will be very off.
As a related question, on my S9 whenever I wasn't going to use humidification (basically, on an airplane), I would use my ClimateLine but not "click" it into place for the machine to recognize its a ClimateLine. I then had humification turned off. When doing something like that, would you tell the machine the hose is a "standard" or "slimline"?
Life before my CPAP was really no life at all!

User avatar
Pugsy
Posts: 65377
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Question on airsense 10 side cover

Post by Pugsy » Thu Dec 13, 2018 11:34 pm

BrandonB wrote:
Thu Dec 13, 2018 11:28 pm
I would use my ClimateLine but not "click" it into place for the machine to recognize its a ClimateLine. I then had humification turned off. When doing something like that, would you tell the machine the hose is a "standard" or "slimline"?
The ClimateLine heated hose is a slimline hose no matter if it's got power to heat it or not.
The settings are for diameter of the hose....and just because you don't power the heated hose doesn't change the diameter.

Standard and slimline refer to hose diameter.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Question on airsense 10 side cover

Post by palerider » Fri Dec 14, 2018 3:14 am

BrandonB wrote:
Thu Dec 13, 2018 11:28 pm
I would use my ClimateLine but not "click" it into place for the machine to recognize its a ClimateLine.
Why not? Just so you don't have to change the setting to 'off'?