Thoughts & Analysis wanted

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
comfortsteve
Posts: 6
Joined: Thu Nov 15, 2018 6:47 pm
Location: Canada

Thoughts & Analysis wanted

Post by comfortsteve » Thu Nov 15, 2018 7:19 pm

Hi, I'm looking for some feedback on my results (attached screenshot). This is also my first time using sleepyhead.

I feel like some nights my sleep is better than others & last night felt particularly good. Without a cpap I snore like monster. I've been using my cpap for about two years now. Thanks in advance!

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: RemStar Auto with A-Flex (561CA)
Attachments
screenshot-20181115-211051.png
screenshot-20181115-211051.png (211.51 KiB) Viewed 3116 times

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Thoughts & Analysis wanted

Post by Okie bipap » Thu Nov 15, 2018 7:43 pm

I would recommend you raise your minimum pressure to at least 8, or maybe even 9. If you don't want to make that much of a jump at one time, try going up one cm every two or three nights.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32306
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Thoughts & Analysis wanted

Post by palerider » Thu Nov 15, 2018 10:15 pm

Okie bipap wrote:
Thu Nov 15, 2018 7:43 pm
I would recommend you raise your minimum pressure to at least 8, or maybe even 9. If you don't want to make that much of a jump at one time, try going up one cm every two or three nights.
Looks about like what I'd suggest.

Basically, you want to smooth out the ups and downs of the pressure line somewhat, so that you don't have as many events while waiting for the machine to work it's way up to pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

comfortsteve
Posts: 6
Joined: Thu Nov 15, 2018 6:47 pm
Location: Canada

Re: Thoughts & Analysis wanted

Post by comfortsteve » Fri Nov 16, 2018 6:51 am

What are you looking at in that report that leads you to those suggestions?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: RemStar Auto with A-Flex (561CA)

User avatar
Pugsy
Posts: 65373
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Thoughts & Analysis wanted

Post by Pugsy » Fri Nov 16, 2018 7:08 am

Suggestions made to increase the minimum pressure because of the clustering of events that are signs that the airway is trying to collapse.
Things like Flow Limitations and snores and RERAs which aren't part of the AHI but still indicators of airway collapsing a bit. Maybe not enough to worsen to the point of earning a flag but enough to perhaps disturb sleep.
Give the machine a better head start on where it needs to go because sometimes it just can't get there fast enough to prevent the airway from collapsing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

comfortsteve
Posts: 6
Joined: Thu Nov 15, 2018 6:47 pm
Location: Canada

Re: Thoughts & Analysis wanted

Post by comfortsteve » Fri Nov 16, 2018 3:22 pm

Thanks everyone. I'm gonna jack that minimum up tonight.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: RemStar Auto with A-Flex (561CA)

comfortsteve
Posts: 6
Joined: Thu Nov 15, 2018 6:47 pm
Location: Canada

Re: Thoughts & Analysis wanted

Post by comfortsteve » Mon Nov 19, 2018 10:10 am

I've attached a recent day's results after adjusting the min pressure from 5 to 9. I looks like I should increase some more. Thoughts?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: RemStar Auto with A-Flex (561CA)
Attachments
screenshot-20181119-120726.png
screenshot-20181119-120726.png (210.17 KiB) Viewed 2988 times

User avatar
kteague
Posts: 7788
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: Thoughts & Analysis wanted

Post by kteague » Mon Nov 19, 2018 11:06 am

Does anyone observe you sleeping and confirms there is still snoring? Just want to make sure you don't have the hose draped over a headboard or across a nightstand in a way that is causing vibrations being interpreted as snores.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
palerider
Posts: 32306
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Thoughts & Analysis wanted

Post by palerider » Mon Nov 19, 2018 11:44 am

comfortsteve wrote:
Mon Nov 19, 2018 10:10 am
I've attached a recent day's results after adjusting the min pressure from 5 to 9. I looks like I should increase some more. Thoughts?
Try 10.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

comfortsteve
Posts: 6
Joined: Thu Nov 15, 2018 6:47 pm
Location: Canada

Re: Thoughts & Analysis wanted

Post by comfortsteve » Mon Nov 19, 2018 4:42 pm

kteague wrote:
Mon Nov 19, 2018 11:06 am
Does anyone observe you sleeping and confirms there is still snoring? Just want to make sure you don't have the hose draped over a headboard or across a nightstand in a way that is causing vibrations being interpreted as snores.
No, but the hose does goes straight from a side table to under my pillow. It should be good.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: RemStar Auto with A-Flex (561CA)

helencpap
Posts: 4
Joined: Thu Oct 25, 2018 2:52 am

Re: Thoughts & Analysis wanted

Post by helencpap » Mon Nov 19, 2018 8:41 pm

Maybe 10

_________________
MachineMask