Post by klm49 » Wed Nov 14, 2018 10:29 am

I have been on CPAP sine 4/2016 with relatively few problems. Seems like the last 5 or 6 mos. My AHI has creeped up quite a bit. I had attributed it to stress but that has subsided yet the AHI's are still high. When I changed Sleep Dr.'s nearly a year ago she mentioned that I might need a different machine (BiPAP?) but since I hadn't been on CPAP for long that Medicare wouldn't pay for a new one. I had been attributing the high AHI's to leaks but now I am beginning to think the pressure just cant go high enough. I'm thinking of raising pressure to 18 possibly. Here's a shot from last night, appreciate any help. I go back to Sleep Dr. in a few months, maybe she will get me a new machine. If you think I need a BIPAP or whatever, what kind of used one would you recommend?
screenshot-20181114-100118.png
screenshot-20181114-100118.png (99.8 KiB) Viewed 3909 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
LSAT
Posts: 13413
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Opinions Please

Post by LSAT » Wed Nov 14, 2018 10:58 am

I think if the doctor wrote a prescription indicating that your current machine was not adequate for your therapy, Medicare would pay for a different machine. It looks like on several occasions your current machine maxed out at 20 even though it wanted to go higher.

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Opinions Please

Post by Pugsy » Wed Nov 14, 2018 11:03 am

Obviously the machine wants to go higher and can't.

I have actually had a chance to use both the ResMed auto adjusting bilevel as well as the Respironics auto adjusting bilevel.
I prefer the ResMed bilevel myself.

It's a bit less noisy in terms of both motor noise and the breathing noise (the darth vadar sounds) and the algorithm responds a bit faster which usually means less time at the higher pressures (when they are needed which in my case is REM related).
Plus the EasyBreathe thing is better than the BiFlex relief IMHO and there are a couple of other customizable features that can be useful for fine tuning the inhale/exhale timing thing.

They need to put you on the fast track for a bilevel ...and an auto adjusting one because you only need the higher pressures for parts of the night. No sense in using higher pressures if you don't need them.

There should be more than enough proof here to satisfy Medicare.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
esel
Posts: 272
Joined: Fri May 20, 2016 3:35 am
Location: switzerland

Re: Opinions Please

Post by esel » Wed Nov 14, 2018 11:13 am

You mainly have two short periods, one from mid night to quater past and one of 40 min after 4 am. Could it be that those are sleep wake junk events ? could you get some close up of the flow rate during these periods.

If you sleep OK at such high pressure, increase by 1 the min and max pressure.

Stress is a tough factor. It may need some time to recover.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Forma Full Face CPAP Mask with Headgear
Additional Comments: AirCurve 10 CS PaceWave, Full Mask, sleepyhead, ASV Min EPAP 5 Max IPAP 10.4 PS 0.4-5.4
Only ME... - :) - Some days sooo slow, some days just running off track ...

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Opinions Please

Post by Pugsy » Wed Nov 14, 2018 11:19 am

My best guess as for the clustering....either REM stage sleep or you were on your back.
With the timing of the clustering....I lean towards REM with maybe a secondary influence combination of being on your back too.
Either way I vote bilevel auto.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Opinions Please

Post by klm49 » Wed Nov 14, 2018 1:14 pm

Pugsy wrote: ↑
Wed Nov 14, 2018 11:19 am
My best guess as for the clustering....either REM stage sleep or you were on your back.
With the timing of the clustering....I lean towards REM with maybe a secondary influence combination of being on your back too.
Either way I vote bilevel auto.

Thanks for all the replies, I'll have to get the zoomed in screenshots later.

Because the Dr. mentioned a different machine last year I have held off on buying a spare machine. Which used Resmed model would you recommend?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Opinions Please

Post by Pugsy » Wed Nov 14, 2018 1:27 pm

For the regular auto adjusting bilevel (not the specialty models with back up rates and all that stuff) it depends on who is paying for it and is cost a factor.

First choice ResMed AirCurve 10 VAuto just because I have a penchant for the latest and greatest especially if insurance is paying the bulk of the cost.
Second choice if paying out of pocket would be the older big brother ResMed S9 VPAP Auto....essentially the same machine in terms of major therapy algorithms and function as the AirCurve 10 VAuto but with the cosmetic upgrade and a few minor "improvements".
Since it is now a discontinued model the price for this model has dropped in the secondary market substantially.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Opinions Please

Post by palerider » Wed Nov 14, 2018 1:55 pm

Second for the recommendation pugsy made.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Opinions Please

Post by klm49 » Wed Nov 14, 2018 2:06 pm

Pugsy wrote: ↑
Wed Nov 14, 2018 1:27 pm
For the regular auto adjusting bilevel (not the specialty models with back up rates and all that stuff) it depends on who is paying for it and is cost a factor.

First choice ResMed AirCurve 10 VAuto just because I have a penchant for the latest and greatest especially if insurance is paying the bulk of the cost.
Second choice if paying out of pocket would be the older big brother ResMed S9 VPAP Auto....essentially the same machine in terms of major therapy algorithms and function as the AirCurve 10 VAuto but with the cosmetic upgrade and a few minor "improvements".
Since it is now a discontinued model the price for this model has dropped in the secondary market substantially.
Thanks again.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: Opinions Please

Post by D.H. » Wed Nov 14, 2018 2:46 pm

A bi-level machine can go higher that 20. As far as I know, all can go to 25; some even higher.

If the change in pressure between inhaling and exhaling bothers you (it sure bothers me), I think you can use it as you would a regular APAP, except with a higher max pressure.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Opinions Please

Post by klm49 » Sat Dec 01, 2018 3:55 pm

Pugsy wrote: ↑
Wed Nov 14, 2018 11:03 am
Obviously the machine wants to go higher and can't.

I have actually had a chance to use both the ResMed auto adjusting bilevel as well as the Respironics auto adjusting bilevel.
I prefer the ResMed bilevel myself.

It's a bit less noisy in terms of both motor noise and the breathing noise (the darth vadar sounds) and the algorithm responds a bit faster which usually means less time at the higher pressures (when they are needed which in my case is REM related).
Plus the EasyBreathe thing is better than the BiFlex relief IMHO and there are a couple of other customizable features that can be useful for fine tuning the inhale/exhale timing thing.

They need to put you on the fast track for a bilevel ...and an auto adjusting one because you only need the higher pressures for parts of the night. No sense in using higher pressures if you don't need them.

There should be more than enough proof here to satisfy Medicare.
Pugsy, I just bought Lucyhere's S9 VPAP Auto to use as a secondary machine when I nap. Hopefully you will help me with the setup. I have the Clinician's so no problem changing other than I don't know what to set where. Any help would be appreciated.
Thanks

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Opinions Please

Post by Pugsy » Sat Dec 01, 2018 3:58 pm

Tell me all your current settings with your apap...and we can get real close using the S9 VPAP.
Easy peasy...I went from a Respironics apap to the same S9 VPap auto super easy.
And you might get lucky and get by with a little less minimum pressure than you need with the apap.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Opinions Please

Post by klm49 » Sat Dec 01, 2018 6:00 pm

Pugsy wrote: ↑
Sat Dec 01, 2018 3:58 pm
Tell me all your current settings with your apap...and we can get real close using the S9 VPAP.
Easy peasy...I went from a Respironics apap to the same S9 VPap auto super easy.
And you might get lucky and get by with a little less minimum pressure than you need with the apap.
Apap
min 17
max 20
c-flex
flex = 1
humidity = auto
2 or 3
Heated tube but off. I only use when very cold.
ramp is 14.5 but I don't use it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask
O2 - 2-4 lpm

User avatar
Pugsy
Posts: 65415
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Opinions Please

Post by Pugsy » Sat Dec 01, 2018 6:12 pm

The flex setting of 1 is minimal exhale relief. Exhale relief on the VAuto is accomplished with PS.
Let's pretend you get 1 cm drop during exhale with Flex at 1...you don't but it's close.

So the apap setting of minimum 17 means 17 on inhale and a little drop to 16 during exhale....max doesn't really matter if you never reach it and with the apap of course limited to the 20 anyway.

So with the VPAP...when you use PS you add to EPAP instead of subtract from IPAP (when apap is used) to get similar settings.
You would set the minimum EPAP to 16...PS to 1 and max IPAP of 20 or higher if you ever hit 20 with the APAP...I would set it to 25 and let the machine do its thing. It won't go anywhere without a good reason to go there.
This will give you 16 exhale and 17 inhale just like the apap is giving you... the timing of the reduction might be a little different and if you find it seems a bit off...let me know as there are some additional settings available to change the timing a bit to get it better in sync with your own respiration.

Now...at some point...do play with the PS a bit just to see how you like it...
Once we see some reports we will know if you can maybe drop minimum EPAP a little and maybe add more IPAP by using more PS or not.
I love PS of 4...I can't get that with my Autoset but I can get close by using EPR at 3...which is like PS of 4.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
klm49
Posts: 442
Joined: Sat May 28, 2016 4:24 pm
Location: Texas

Re: Opinions Please

Post by klm49 » Sat Dec 01, 2018 6:59 pm

Pugsy wrote: ↑
Sat Dec 01, 2018 6:12 pm
The flex setting of 1 is minimal exhale relief. Exhale relief on the VAuto is accomplished with PS.
Let's pretend you get 1 cm drop during exhale with Flex at 1...you don't but it's close.

So the apap setting of minimum 17 means 17 on inhale and a little drop to 16 during exhale....max doesn't really matter if you never reach it and with the apap of course limited to the 20 anyway.

So with the VPAP...when you use PS you add to EPAP instead of subtract from IPAP (when apap is used) to get similar settings.
You would set the minimum EPAP to 16...PS to 1 and max IPAP of 20 or higher if you ever hit 20 with the APAP...I would set it to 25 and let the machine do its thing. It won't go anywhere without a good reason to go there.
This will give you 16 exhale and 17 inhale just like the apap is giving you... the timing of the reduction might be a little different and if you find it seems a bit off...let me know as there are some additional settings available to change the timing a bit to get it better in sync with your own respiration.

Now...at some point...do play with the PS a bit just to see how you like it...
Once we see some reports we will know if you can maybe drop minimum EPAP a little and maybe add more IPAP by using more PS or not.
I love PS of 4...I can't get that with my Autoset but I can get close by using EPR at 3...which is like PS of 4.

Mode = Vauto
Max ipap 25
min epap 16
PS 1
I think I got it all. I'm leaving Ti max at default of 2 and Ti min to 0.3 default.
Trigger = Med.
Cycle = Med.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Amara View Full Face CPAP Mask with Headgear
Additional Comments: Setting: PS 4.0 over 14.0-25.0; Humidifer 4 & Heated Hose
Resmed AirCurve10 Vauto w/humidifier
Amara View mask