1 Month Newbie Check-in

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 1 Month Newbie Check-in

Post by palerider » Thu Nov 08, 2018 8:20 pm

mcguirejohnr wrote:
Thu Nov 08, 2018 6:56 pm
A good number of days now and I am having very good success with the higher minimum pressure setting...thanks for the help and suggestions.
Thanks for the followup. Hearing that you're having much better success is what makes this all worthwhile.

Are you feeling better? less headaches?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Fri Nov 09, 2018 11:47 am

Headaches BY 80-90% thanks to the help of you and this board. AHI consistently well under 1.0. Interestingly, the Sleep Doctor had no idea a person could figure out how to adjust their own therapy settings but was happy to accept my therapy efficacy findings and apply those to my official prescription.

Unrelated mask update. Really like the Simplus...it sees to be wider at the corners of the mouth. I really have no big issues with any of those that I now have in rotation. F20 AirFit and AirTouch, F30 AirFit, Dreamwear FF and now Simplus. I suppose I am lucky to not have many mask issues. Only thing that does work for me are the nasal options as I am apparently a knuckle dragging mouth breather.

_________________
MachineMaskHumidifier

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Sat Feb 23, 2019 2:14 pm

4 more months and all is doing very well. AHI is always < 1.0. Headaches gone. Only remaining issue is mask leaks in the early AM hours that seems to be correlated to pressure increases around that time. Snugged the mask down a little tighter, moved max pressure down just a tiny bit and removed the Cflex/Aflex which seems to have worked. I really think removing the Cflex/Aflex made the most difference for me since it appeared to allow lower max pressure needed to keep the airway open. Similar results on the primary Dreamstation machine and the Resmed Autoset I have at the lake. Also pretty good results with the AirMini when traveling although I am finding I prefer the F&P Simplus mask so I broke down and ordered an adapter to convert the AirMini port to a "standard" hose.

Thanks to the forum members.

_________________
MachineMaskHumidifier

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Tue Jan 21, 2020 6:23 pm

Almost a year since my last check-in. Using the Simplus mask mostly,,,F20s for traveling. A lot more leaks lately so tried a fresh cusion and same issue. Moved pressures from 9/14.5 to 9/12.5. Seems like the machine is ramping to higher pressures more quickly...perhaps jsut leak related. Anyone have thoughts on problems/issues in the charts?
SleepyHeadData 2019_01.png
SleepyHeadData 2019_01.png (485.43 KiB) Viewed 6001 times

_________________
MachineMaskHumidifier

gv280z
Posts: 137
Joined: Mon Apr 30, 2018 9:19 pm

Re: 1 Month Newbie Check-in

Post by gv280z » Wed Jan 22, 2020 9:27 am

Lowering your max pressure doesn't help anything. Infact, it restricts your machine in how much therapy it can provide for you, should you have a rough night. It happens man, mine is set at 14 / 20, on most nights I'm around 15 or so, occasionally I've seen it spike up to 18 or 19 and I think, damn man....wow, guess I had a bad night..you never know. The higher pressure limit changes nothing in how the machine delivers pressure to you, up until it actually needs to reach up and touch that ceiling. When I first began, after a month my doctor tells me he needs to raise my max from 15 up to 20 because I was banging against the limit every single night and suffering from apneas while getting pressure.

Aside from that, it sounds like your doing good.
Yay :D For no more diaphragm cramps and dozing off while driving! :lol:

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: 1 Month Newbie Check-in

Post by Jas_williams » Wed Jan 22, 2020 10:02 am

The things that jump out to me are leaks they can be high at times, also your having the machine reporting vibratory snores (VS) which indicates more minimum pressure needed prob 10

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Wed Jan 22, 2020 5:46 pm

The leaks seem to be when the pressures get higher. I think I will try 10.5/12.5 for a while and see if it helps. The AHI seems pretty darn good and not many hypopneas, just hate waking up to my mask hissing like crazy. I wnat to let the maximum go as high as needed but itrading that for the mask leaks does not seem worth it.

_________________
MachineMaskHumidifier

User avatar
Pugsy
Posts: 65350
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 1 Month Newbie Check-in

Post by Pugsy » Wed Jan 22, 2020 7:05 pm

You are having some large leaks even when the pressure isn't all that high.
Look at the time where the first large leak is flagged and look at the pressure line. It's essentially not increased at all.
The little tiny saw tooth looking increases in pressure are just 1.5 cm pressure probes....not high pressures at all.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 1 Month Newbie Check-in

Post by palerider » Wed Jan 22, 2020 9:18 pm

mcguirejohnr wrote:
Tue Jan 21, 2020 6:23 pm
Almost a year since my last check-in. Using the Simplus mask mostly,,,F20s for traveling. A lot more leaks lately so tried a fresh cusion and same issue. Moved pressures from 9/14.5 to 9/12.5. Seems like the machine is ramping to higher pressures more quickly...perhaps jsut leak related. Anyone have thoughts on problems/issues in the charts?

SleepyHeadData 2019_01.png
Set the max back to 20.

Pressure *NEVER* goes up in response to leaks, in fact, auto machines will *LOWER* the pressure if the leaks are too bad.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Thu Jan 23, 2020 8:32 pm

I will experiment and report back in a few weeks. i will on the Autoset with F20 Toucnh for the next few days then back to the Dreamstation with Simplus for a week or so after that.

_________________
MachineMaskHumidifier

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: 1 Month Newbie Check-in

Post by Jas_williams » Fri Jan 24, 2020 2:02 am

mcguirejohnr wrote:
Thu Jan 23, 2020 8:32 pm
I will experiment and report back in a few weeks. i will on the Autoset with F20 Toucnh for the next few days then back to the Dreamstation with Simplus for a week or so after that.
Any reason swapping machines and then back, many people find much better results with the Resmed

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Tue Jan 28, 2020 9:02 pm

Dreamsation is at home, Autoset is at lake hosue and AirMini is when I travel.

_________________
MachineMaskHumidifier

mcguirejohnr
Posts: 22
Joined: Mon Oct 08, 2018 6:24 pm

Re: 1 Month Newbie Check-in

Post by mcguirejohnr » Sat Feb 01, 2020 7:38 pm

I tried adding a C-Flex EPR setting of 2 and that seems to work sicne the mask leaks were aloways on exhale. I think this may be the winning ticket.

_________________
MachineMaskHumidifier